  • pbluescript
    replied
    It depends on the sample and what you want to sequence, but in your case you could use 1.1X or 1.2X beads to remove the adapter-adapter ligations and retain your 300-400 bp library. This would retain more DNA than doing another gel extraction.

    Leave a comment:


  • cybeline
    replied
    Originally posted by pbluescript View Post
    I have attached an image showing the effect of altering the ratio of AMPure beads/sample. You can see that using a lower ratio will remove increasing sizes of DNA fragments. You can even use this for size selection by using two different concentrations. For example, if you wanted to get a band centered ~400 bp, you could use 0.7X beads and then take the supernatant. This would include all the DNA fragments that were smaller than ~500bp and didn't bind to the beads. Then mix the supernatant with 0.9X beads and everything larger than ~300bp would bind. Elute the bound DNA and you'll have a fairly broad, but good enough to sequence, library.
    If you try the beads, there can be a bit of lot-to-lot variability, so I would test them with a 100 bp library as I did in this gel.

    I haven't used the Truseq kit, but I have used the Truseq adapters with outsourced enzymes/buffers and the same problem can occur.
    This is great, thank you.

    So do you only use this once after the ligation step? or also in the very end to make sure the libraries are fully cleaned?

    Thanks again.

    Leave a comment:


  • pbluescript
    replied
    I have attached an image showing the effect of altering the ratio of AMPure beads/sample. You can see that using a lower ratio will remove increasing sizes of DNA fragments. You can even use this for size selection by using two different concentrations. For example, if you wanted to get a band centered ~400 bp, you could use 0.7X beads and then take the supernatant. This would include all the DNA fragments that were smaller than ~500bp and didn't bind to the beads. Then mix the supernatant with 0.9X beads and everything larger than ~300bp would bind. Elute the bound DNA and you'll have a fairly broad, but good enough to sequence, library.
    If you try the beads, there can be a bit of lot-to-lot variability, so I would test them with a 100 bp library as I did in this gel.

    I haven't used the Truseq kit, but I have used the Truseq adapters with outsourced enzymes/buffers and the same problem can occur.
    Attached Files

    Leave a comment:


  • Simone78
    replied
    2% SizeSelect E-gel from Invitrogen:



    The peak you get on the Bioanalyzer after size selection is very narrow (see attachment) and you can collect as many fractions as you want (I usually collect 4 or 5 fractions at 30 sec interval, which corresponds to about 50 bp difference).

    After AMPure XP you won't probably remove 100% of the band at 120 bp, but you'll make this fraction much smaller than after Qiagen columns (for example).
    Playing with the ratio DNA/beads might further improve the final results, but I haven't tried it yet.

    Sorry, no experience with the truseq kit. I'm using a modified version of the PE Illumina protocol for whole methylome seq.
    Last edited by Simone78; 08-24-2017, 05:42 AM.

    Leave a comment:


  • cybeline
    replied
    Those are great answers, thank you very much!

    Simone78, could you please link the E-gel? and how do you get rid of the 120bp fragments?

    Thanks again,

    Another question, does this happen as well while using the Illumina truseq kit?

    Leave a comment:


  • Simone78
    replied
    I had the same problem(s). To get rid of the adaptor-adaptor dimers I now do size selection using 2% E-gel (great because you can also select more than one single fraction, so you have a backup in case something goes wrong during the pre-amplification step). And you don't need an additional purification step, just pipette out the water (containing your DNA) from the well. To remove the primer dimers I use AMPure XP (1.8:1 ratio) which removes (almost) 100% of the fragments <100 bp.

    Leave a comment:


  • jgibbons1
    replied
    I've had the same issue but routinely use a second gel purification to fix this. Although, you absolutely lose some DNA with the second gel purification. I use the Qiagen Minelute kit to elute with a smaller volume (10-12ul) in order to increase the concentration. This seems to do the trick nicely. I haven't played around with the AMpure beads but will be doing so soon.

    Leave a comment:


  • cybeline
    replied
    Originally posted by pbluescript View Post
    Sure. The razor is usually very long compared to the region of the gel you want to cut out. It will touch lots of other sizes of DNA in the gel, which can lead to intra-sample contamination with sizes you didn't originally intend to be in your sample. I got better results when I purchased some gel cutting tips or used AMpure beads for size selection.
    Taking a look at your gel, did you cut out everything between 300-400bp for samples CB5-8?
    Yes. That gel is after everything, just before submitting it to sequencing.
    They suggested to re-gel purify all the samples between 300 to 400. So far I have done cb1-4 but I obtained final concentration of 1 to 4 ng/ul so around 4.5 to 17 nM final concentration. I lost a lot of dna from the 2nd purification.

    Leave a comment:


  • pbluescript
    replied
    Originally posted by cybeline View Post
    Thank you very much for your answer.

    Could you please expand on the role of the razor to cut the gel? Thanks again
    Sure. The razor is usually very long compared to the region of the gel you want to cut out. It will touch lots of other sizes of DNA in the gel, which can lead to intra-sample contamination with sizes you didn't originally intend to be in your sample. I got better results when I purchased some gel cutting tips or used AMpure beads for size selection.
    Taking a look at your gel, did you cut out everything between 300-400bp for samples CB5-8?

    Leave a comment:


  • cybeline
    replied
    Originally posted by pbluescript View Post
    You should use less of the adapters that you ligated to your sample, not less enrichment primers. A 10:1 molar ratio at most, but I usually use less. If you have too many of them during your ligation reaction, they will end up ligating to each other and then that will end up amplifying in your enrichment PCR step. Even if you size selected with an agarose gel, you can still get some, especially if you used a razor to cut out bands. Since they are smaller than your sample+adapter, the adapter-only ligation products can dominate the reaction even if they are at a lower concentration than your sample to begin with.
    Thank you very much for your answer.

    Could you please expand on the role of the razor to cut the gel? Thanks again

    Leave a comment:


  • pbluescript
    replied
    Originally posted by cybeline View Post
    So if I understand correctly, I should use less enrichment primers?
    You should use less of the adapters that you ligated to your sample, not less enrichment primers. A 10:1 molar ratio at most, but I usually use less. If you have too many of them during your ligation reaction, they will end up ligating to each other and then that will end up amplifying in your enrichment PCR step. Even if you size selected with an agarose gel, you can still get some, especially if you used a razor to cut out bands. Since they are smaller than your sample+adapter, the adapter-only ligation products can dominate the reaction even if they are at a lower concentration than your sample to begin with.

    Leave a comment:


  • cybeline
    replied
    Originally posted by pbluescript View Post
    Sorry, I shouldn't have said primer dimers. They would be more accurately described as adapter-adapter ligations. The primers are nowhere near 120bp in length, so what you are actually amplifying are adapters that ligated to each other rather than to your sample. My suggestions for altering your ligation reaction should help with that.
    So if I understand correctly, I should use less enrichment primers?

    Leave a comment:


  • cybeline
    replied
    Originally posted by cascoamarillo View Post
    Hi,

    One stupid question; are you sure it's primer dimer?, or maybe it could be a double strain adapter-adapter construct which is amplified after the pcr.
    I assume that the 75 bp peak is unreacted primer and that the 150 bp peak is a ligation artifact of two primers ligating together. But I thought it was the enrichment primers and not the ligation adapters.

    The way I made my multiplex the barcodes were added with the adapters and not with the enrichment pcr.

    Please let me know if that helps and thank you.

    Leave a comment:


  • cascoamarillo
    replied
    Hi,

    One stupid question; are you sure it's primer dimer?, or maybe it could be a double strain adapter-adapter construct which is amplified after the pcr.

    Leave a comment:


  • pbluescript
    replied
    Sorry, I shouldn't have said primer dimers. They would be more accurately described as adapter-adapter ligations. The primers are nowhere near 120bp in length, so what you are actually amplifying are adapters that ligated to each other rather than to your sample. My suggestions for altering your ligation reaction should help with that.

    Leave a comment:

Latest Articles

Collapse

  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM
  • SEQadmin2
    New Genomics Technologies Take Aim at Long-Standing Limits
    by SEQadmin2


    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
    ...
    09-28-2026, 10:25 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
38 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
28 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-15-2026, 10:52 AM
0 responses
41 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-14-2026, 05:39 AM
0 responses
45 views
0 reactions
Last Post SEQadmin2  
Working...