    Originally posted by pbluescript View Post
    Read post 16 in this thread.
    Thanks.
    I was just wondering if someone would have a little bit more percised answer.
    My problem is that my fragments of interest range from 150bp up to 400bp and I need to get rid of the 120-130bp fragments. They are just so close. I tried a 1.8 :1 ratio and was successful once but starting out with less DNA I could not reporduce that.

    Comment

    • pbluescript
      Senior Member
      • Nov 2009
      • 224

      #32
      Originally posted by dagmar View Post
      Thanks.
      I was just wondering if someone would have a little bit more percised answer.
      My problem is that my fragments of interest range from 150bp up to 400bp and I need to get rid of the 120-130bp fragments. They are just so close. I tried a 1.8 :1 ratio and was successful once but starting out with less DNA I could not reporduce that.
      Ah... that would be a problem. For getting separation resolution like that, the beads and even an agarose gel wouldn't be that great. You might want to try PAGE or, if you have the money, the Caliper XT.

      The Caliper is capable of some very tight size separation.

      Comment

      • cascoamarillo
        Senior Member
        • Oct 2010
        • 164

        #33
        Originally posted by dagmar View Post
        Thanks.
        I was just wondering if someone would have a little bit more percised answer.
        My problem is that my fragments of interest range from 150bp up to 400bp and I need to get rid of the 120-130bp fragments. They are just so close. I tried a 1.8 :1 ratio and was successful once but starting out with less DNA I could not reporduce that.
        yeah, the PAGE is very effective for that sizes (although more time consume); I did it with my small RNA libraries and apparently I get rid off all the adapter-adapter artifacts.

        Maybe the system Pippin Prep gives you another solution:
        The Pippin platform is a proprietary technology that automates preparative gel electrophoresis for life science: NGS library construction and DNA band capture.

        Comment

        • dagmar
          Member
          • Apr 2011
          • 10

          #34
          Originally posted by pbluescript View Post
          Ah... that would be a problem. For getting separation resolution like that, the beads and even an agarose gel wouldn't be that great. You might want to try PAGE or, if you have the money, the Caliper XT.

          The Caliper is capable of some very tight size separation.
          Originally posted by cascoamarillo View Post
          yeah, the PAGE is very effective for that sizes (although more time consume); I did it with my small RNA libraries and apparently I get rid off all the adapter-adapter artifacts.

          Maybe the system Pippin Prep gives you another solution:
          http://www.sagescience.com/
          Thanks. I guess you just saved me time and effort.
          I was planning on using the Pippin Prep next but the test runs were a little disappointing. I had problems loading samples because 2 out of 5 wells would refill with buffer and the yield from the ones that worked was very little.
          Do you have any experience with the Pippin Prep? I will not get the chance to try the Caliper XP. I guess I will give the Pippin Prep a second chance so any further advice would be great.

          Comment

          • cascoamarillo
            Senior Member
            • Oct 2010
            • 164

            #35
            Originally posted by dagmar View Post
            Thanks. I guess you just saved me time and effort.
            I was planning on using the Pippin Prep next but the test runs were a little disappointing. I had problems loading samples because 2 out of 5 wells would refill with buffer and the yield from the ones that worked was very little.
            Do you have any experience with the Pippin Prep? I will not get the chance to try the Caliper XP. I guess I will give the Pippin Prep a second chance so any further advice would be great.
            The Pippin Prep that we have is been used by other people in the department, not me. They also do RNA-seq, but I'm not sure what's the potencial separation-resolution of that machine (it's still a agarose separation system). For my stuff, where the adapters artifacts (75 bp) and the library (100 bp) are so closed, was not an option.
            Hope it helps.

            Comment

            • odile
              Member
              • Sep 2009
              • 30

              #36
              One alternative from the E-gel is the recovery gels from Lonza (flashgels). I tested them last week and they work great.

              Comment

              • niceday
                Member
                • Apr 2010
                • 68

                #37
                to dagmar. Some different ratios are 1:1 for getting rid of 150bp and smaller. That might be useful if you don't mind taking a potential hit on the 150bp area of your library range. Alternatively 1.5:1. I'm not precisely sure of the size but it might get rid of some or all of your 120-130bp fragments.

                Comment

                • dagmar
                  Member
                  • Apr 2011
                  • 10

                  #38
                  Originally posted by niceday View Post
                  to dagmar. Some different ratios are 1:1 for getting rid of 150bp and smaller. That might be useful if you don't mind taking a potential hit on the 150bp area of your library range. Alternatively 1.5:1. I'm not precisely sure of the size but it might get rid of some or all of your 120-130bp fragments.
                  Thanks. I tried it out now. 1.5 which did a good job. 1.3 and 1.0 somehow did not get rid of any of the artifacts around 120bp.

                  Comment

                  • dagmar
                    Member
                    • Apr 2011
                    • 10

                    #39
                    Originally posted by cascoamarillo View Post
                    The Pippin Prep that we have is been used by other people in the department, not me. They also do RNA-seq, but I'm not sure what's the potencial separation-resolution of that machine (it's still a agarose separation system). For my stuff, where the adapters artifacts (75 bp) and the library (100 bp) are so closed, was not an option.
                    Hope it helps.
                    I finally have some results from the Pippin Prep 2% and it does seperate the 120bp artifacts from my 150bp fragments so I could imagine a 3% cassette might work for you too, but the yield is not as good as I excpected so far.

                    Comment

                    • dagmar
                      Member
                      • Apr 2011
                      • 10

                      #40
                      Originally posted by odile View Post
                      One alternative from the E-gel is the recovery gels from Lonza (flashgels). I tested them last week and they work great.
                      Never heard of it. Sounds great. It really might be an alternative.
                      Can you do tight seperations like 120bp from 150bp and collect library ranges up to 300bp with it? On the Pippin prep low single nanogramms of DNA can be loeaded what about the minimum load of DNA of the Flashgels.
                      Thanks!

                      Comment

                      • patisev
                        Junior Member
                        • Apr 2011
                        • 1

                        #41
                        adaptor-adaptor ligation problem with SOLiD small RNA library

                        Originally posted by pbluescript View Post
                        You should use less of the adapters that you ligated to your sample, not less enrichment primers. A 10:1 molar ratio at most, but I usually use less. If you have too many of them during your ligation reaction, they will end up ligating to each other and then that will end up amplifying in your enrichment PCR step. Even if you size selected with an agarose gel, you can still get some, especially if you used a razor to cut out bands. Since they are smaller than your sample+adapter, the adapter-only ligation products can dominate the reaction even if they are at a lower concentration than your sample to begin with.
                        I seem to have the same problem while constructing small RNA libraries for Solid sequencing, do you have any experience with this equipment? Since the fragments I should select are almost the same size as adaptor-adaptor fragments, we are having trouble collecting them from the gel. Do you think that using less adaptor could help me out too? Any experience with Invitrogen eGel for this kind of size selection? Thanks!

                        Comment

                        • captainentropy
                          Member
                          • Mar 2009
                          • 89

                          #42
                          If you have access to one, I recommend the Pippin Prep. The lab next door to me tested the Caliper XT recently but settled on and purchased the Pippin (which we both used in another lab before). Since the Pippin uses electroelution there is little loss of DNA in the range you are selecting for compared to the manual gel-excision method - you'll lose at least 50% of your library that way in addition to the possibility of cross-contamination.

                          However, after the elution step there is no need for purification of the DNA prior to the PCR, you will only lose DNA doing that. You can amplify directly from the eluted sample. What I do is I take the elution volume (I elute 200-600bp) and divide it into 36 uL aliquots and how ever many that is is how many PCR reactions I set up for that sample. Usually this is between 7-12. Doing this means you are able to maximize the amount of size-selected DNA in the PCR step resulting in more library. This also means you can get away with fewer amplification cycles thus reducing the number of potential PCR artifacts.

                          As for the OP question, the 124 bp peak is definitely amplified adapters. I've seen it many times. The Pippin (or CaliperXT) will reduce/eliminate this. After I switched to the Pippin I've only seen any adapter peak once, and it was small compared to my size-selected library. Also, reducing the adapter dilution ratio helps. I use 1:30. However, my lab is moving away from that and we are now going to try to optimize the dilution amount for each library. More work, but we'll have better libraries. Also, Illumina (I think it was them) now suggests using longer ligation times. I do overnight ligations anyway.

                          Comment

                          • dagmar
                            Member
                            • Apr 2011
                            • 10

                            #43
                            Hi Captainentropy,
                            I am glad to here you recommend the Pippin Prep. The few samples I just recently got to run were size selected very well but I still have the problem of loading my samples. Did you ever experience that the sample wells would refill with buffer before you could load your sample? Even if I would load my sample very quickly after emptying the sample well I would loose some of my sample do overflow. I do not want to take that risk but running only 2 to 3 samples on one cassette is very pricy too.
                            I ligate over night as well. I also tried using less adapters but this would always end up in almost no library at all. May I asked with how much DNA you start out with.
                            Thanks!!!

                            Comment

                            • cybeline
                              Member
                              • Jul 2010
                              • 15

                              #44
                              I was just looking at the TruSeq kit from Illumina. After ligation they do 2 AMpure beads purification and then a gel purification.
                              Seemed like overkill and I am a little concerned about loss of DNA.
                              Any thoughts?

                              Comment

                              • captainentropy
                                Member
                                • Mar 2009
                                • 89

                                #45
                                @dagmar. Yes, I've seen that happen with the well. Only once or twice maybe. It was probably just a malformed well allowing more buffer to flow into it since the buffer chamber is right next to it. If you're seeing that all the time then I'm not sure. Maybe poking a hole in the bottom of the well would cause that?

                                I try to use at least 100ng of ChIPed DNA if I can. If I have more then I use more. I've gone down to 1:50 dilution and got good library yield based on the Bioanalyzer trace. In fact for that one test it was slightly better than 1:30, but I'd say with low amounts of starting material 1:30 is fine. I have used 1 ug of starting DNA with 1:30 adapters and still got great library yield based on Bioanalyzer. We're going to try and optimize that step though, finding the optimal amount of adapters to add to 10, 50, 100, 500 ng, etc. of DNA.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                                  by SEQadmin2



                                  CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                                  Despite this, “CRISPR helped turn genome editing from a specialized technique into
                                  ...
                                  07-31-2026, 11:01 AM
                                • SEQadmin2
                                  Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                  by SEQadmin2


                                  Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                  The systematic characterization of the human proteome has
                                  ...
                                  07-20-2026, 11:48 AM
                                • SEQadmin2
                                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                  by SEQadmin2



                                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                  ...
                                  07-09-2026, 11:10 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 07-31-2026, 02:55 AM
                                0 responses
                                18 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-24-2026, 12:17 PM
                                0 responses
                                15 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-23-2026, 11:41 AM
                                0 responses
                                14 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-20-2026, 11:10 AM
                                0 responses
                                26 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...