    Primer dimer problem

    Hello,

    I have just finished prepping multiplex samples for paired end Illumina sequencing using NEB products but after a bioanalyzer I have a lot of primer dimer at 75bp and around 120bp which seems even stronger than my bands at 300bp.
    The sequencing facility does not want to sequence them due to the primer dimer problem.

    Anybody else having the same problem?

    I was suggested to re-gel purify all the samples but I am afraid that I will end up with not enough DNA to sequence.

    Any other suggestion or the effect it could have if we sequence it like this?

    Attached is a picture of the caliper data.

    Thank you,
    Attached Files
  • pbluescript
    Senior Member
    • Nov 2009
    • 224

    #2
    If you sequence it like this, you will lose a lot of data due to the machine just sequencing primer dimers.
    You could size select using AMPure beads. That will give you better yield than gel extraction.
    When you do your ligations next time, use less adapter and be sure to add the ligase last to your reactions, not in a master mix.

    Comment

    • cybeline
      Member
      • Jul 2010
      • 15

      #3
      Originally posted by pbluescript View Post
      If you sequence it like this, you will lose a lot of data due to the machine just sequencing primer dimers.
      You could size select using AMPure beads. That will give you better yield than gel extraction.
      When you do your ligations next time, use less adapter and be sure to add the ligase last to your reactions, not in a master mix.
      Thank you very much for your answer.

      I will look into Ampure beads for the cleanup.
      But I believe the primer diner occurs during the pcr enrichment step since after the ligation I gel purify my samples.

      Comment

      • pbluescript
        Senior Member
        • Nov 2009
        • 224

        #4
        Sorry, I shouldn't have said primer dimers. They would be more accurately described as adapter-adapter ligations. The primers are nowhere near 120bp in length, so what you are actually amplifying are adapters that ligated to each other rather than to your sample. My suggestions for altering your ligation reaction should help with that.

        Comment

        • cascoamarillo
          Senior Member
          • Oct 2010
          • 164

          #5
          Hi,

          One stupid question; are you sure it's primer dimer?, or maybe it could be a double strain adapter-adapter construct which is amplified after the pcr.

          Comment

          • cybeline
            Member
            • Jul 2010
            • 15

            #6
            Originally posted by cascoamarillo View Post
            Hi,

            One stupid question; are you sure it's primer dimer?, or maybe it could be a double strain adapter-adapter construct which is amplified after the pcr.
            I assume that the 75 bp peak is unreacted primer and that the 150 bp peak is a ligation artifact of two primers ligating together. But I thought it was the enrichment primers and not the ligation adapters.

            The way I made my multiplex the barcodes were added with the adapters and not with the enrichment pcr.

            Please let me know if that helps and thank you.

            Comment

            • cybeline
              Member
              • Jul 2010
              • 15

              #7
              Originally posted by pbluescript View Post
              Sorry, I shouldn't have said primer dimers. They would be more accurately described as adapter-adapter ligations. The primers are nowhere near 120bp in length, so what you are actually amplifying are adapters that ligated to each other rather than to your sample. My suggestions for altering your ligation reaction should help with that.
              So if I understand correctly, I should use less enrichment primers?

              Comment

              • pbluescript
                Senior Member
                • Nov 2009
                • 224

                #8
                Originally posted by cybeline View Post
                So if I understand correctly, I should use less enrichment primers?
                You should use less of the adapters that you ligated to your sample, not less enrichment primers. A 10:1 molar ratio at most, but I usually use less. If you have too many of them during your ligation reaction, they will end up ligating to each other and then that will end up amplifying in your enrichment PCR step. Even if you size selected with an agarose gel, you can still get some, especially if you used a razor to cut out bands. Since they are smaller than your sample+adapter, the adapter-only ligation products can dominate the reaction even if they are at a lower concentration than your sample to begin with.

                Comment

                • cybeline
                  Member
                  • Jul 2010
                  • 15

                  #9
                  Originally posted by pbluescript View Post
                  You should use less of the adapters that you ligated to your sample, not less enrichment primers. A 10:1 molar ratio at most, but I usually use less. If you have too many of them during your ligation reaction, they will end up ligating to each other and then that will end up amplifying in your enrichment PCR step. Even if you size selected with an agarose gel, you can still get some, especially if you used a razor to cut out bands. Since they are smaller than your sample+adapter, the adapter-only ligation products can dominate the reaction even if they are at a lower concentration than your sample to begin with.
                  Thank you very much for your answer.

                  Could you please expand on the role of the razor to cut the gel? Thanks again

                  Comment

                  • pbluescript
                    Senior Member
                    • Nov 2009
                    • 224

                    #10
                    Originally posted by cybeline View Post
                    Thank you very much for your answer.

                    Could you please expand on the role of the razor to cut the gel? Thanks again
                    Sure. The razor is usually very long compared to the region of the gel you want to cut out. It will touch lots of other sizes of DNA in the gel, which can lead to intra-sample contamination with sizes you didn't originally intend to be in your sample. I got better results when I purchased some gel cutting tips or used AMpure beads for size selection.
                    Taking a look at your gel, did you cut out everything between 300-400bp for samples CB5-8?

                    Comment

                    • cybeline
                      Member
                      • Jul 2010
                      • 15

                      #11
                      Originally posted by pbluescript View Post
                      Sure. The razor is usually very long compared to the region of the gel you want to cut out. It will touch lots of other sizes of DNA in the gel, which can lead to intra-sample contamination with sizes you didn't originally intend to be in your sample. I got better results when I purchased some gel cutting tips or used AMpure beads for size selection.
                      Taking a look at your gel, did you cut out everything between 300-400bp for samples CB5-8?
                      Yes. That gel is after everything, just before submitting it to sequencing.
                      They suggested to re-gel purify all the samples between 300 to 400. So far I have done cb1-4 but I obtained final concentration of 1 to 4 ng/ul so around 4.5 to 17 nM final concentration. I lost a lot of dna from the 2nd purification.

                      Comment

                      • jgibbons1
                        Senior Member
                        • Oct 2009
                        • 135

                        #12
                        I've had the same issue but routinely use a second gel purification to fix this. Although, you absolutely lose some DNA with the second gel purification. I use the Qiagen Minelute kit to elute with a smaller volume (10-12ul) in order to increase the concentration. This seems to do the trick nicely. I haven't played around with the AMpure beads but will be doing so soon.

                        Comment

                        • Simone78
                          Senior Member
                          • Oct 2010
                          • 208

                          #13
                          I had the same problem(s). To get rid of the adaptor-adaptor dimers I now do size selection using 2% E-gel (great because you can also select more than one single fraction, so you have a backup in case something goes wrong during the pre-amplification step). And you don't need an additional purification step, just pipette out the water (containing your DNA) from the well. To remove the primer dimers I use AMPure XP (1.8:1 ratio) which removes (almost) 100% of the fragments <100 bp.

                          Comment

                          • cybeline
                            Member
                            • Jul 2010
                            • 15

                            #14
                            Those are great answers, thank you very much!

                            Simone78, could you please link the E-gel? and how do you get rid of the 120bp fragments?

                            Thanks again,

                            Another question, does this happen as well while using the Illumina truseq kit?

                            Comment

                            • Simone78
                              Senior Member
                              • Oct 2010
                              • 208

                              #15
                              2% SizeSelect E-gel from Invitrogen:



                              The peak you get on the Bioanalyzer after size selection is very narrow (see attachment) and you can collect as many fractions as you want (I usually collect 4 or 5 fractions at 30 sec interval, which corresponds to about 50 bp difference).

                              After AMPure XP you won't probably remove 100% of the band at 120 bp, but you'll make this fraction much smaller than after Qiagen columns (for example).
                              Playing with the ratio DNA/beads might further improve the final results, but I haven't tried it yet.

                              Sorry, no experience with the truseq kit. I'm using a modified version of the PE Illumina protocol for whole methylome seq.
                              Last edited by Simone78; 08-24-2017, 05:42 AM.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              49 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              37 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              46 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              46 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...