    Low yield after library prep

    Hello,

    I am using the Ion Shear Plus reagents kit to prepare libraries for the PGM. I input 200 ng of DNA (measured by Qubit) but am only getting 0.01-0.1 ng/ul at the end of the process. This is enough to continue to sequencing, but I'm wondering if there's anything I can do to improve the yield slightly.

    I am shearing the DNA for 7 minutes, which looks really good on the E-gel. It seems like maybe the amplification step is not working? I use zymo instead of the Ampure beads for cleanup and elute in 100 mM Tris.

    Thanks,

    Jessica
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Originally posted by jhalpin View Post
    I use zymo instead of the Ampure beads for cleanup and elute in 100 mM Tris.
    100 mM Tris is too concentrated and would inhibit many enzymes. If Ion protocol uses 1.8x bead for clean up you can replace it with column clean up. For this application I would use "Q" columns. If the bead ratio is less than 1.8x, it does a small fragment removal as well and column may not be a good replacement for bead.

    Comment

    • jhalpin
      Member
      • Jan 2015
      • 26

      #3
      Originally posted by nucacidhunter View Post
      100 mM Tris is too concentrated and would inhibit many enzymes. If Ion protocol uses 1.8x bead for clean up you can replace it with column clean up. For this application I would use "Q" columns. If the bead ratio is less than 1.8x, it does a small fragment removal as well and column may not be a good replacement for bead.

      Well huh. Looking back at my notes I wrote down 10 mM. Not sure when/how that turned into 100 mM. I guess I know why my libraries are weird now.

      I'll check out the Q columns, too. Thanks!

      Comment

      • jhalpin
        Member
        • Jan 2015
        • 26

        #4
        Allright, fixing the Tris concentration didn't fix my problem. I'm inputting 200 ng and getting 0.02 ng/ul at the end of library prep. It's technically enough to sequence but it's really low.. I'd really like to be getting 10 times that at least, and I used to get that much. Any other suggestions??

        Comment

        Latest Articles

        Collapse

        • GATTACAT
          Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by GATTACAT
          Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
          07-01-2026, 11:43 AM
        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 07-02-2026, 11:08 AM
        0 responses
        25 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-30-2026, 05:37 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        55 views
        0 reactions
        Last Post SEQadmin2  
        Working...