  • eoh001
    Member
    • Mar 2011
    • 19

    #1

    reset reagent

    Hello
    I observed a strange sequence result for two times.
    The run seemed OK until a certain point, and sequnce quality collapse and the low sequence quality retained until the end of the run.
    Two Times !!!
    I found that the sequnce quality drop down right after resetting new primer. and I found that there was a reset process right before applying new primer from the log file.
    I checked the reset reagent, and the reset reagent bottle was almost empty!!!
    I am not sure that the low sequence quality mainly from the shortage of the reset reagent, but I am kind of positive.

    Anyone has an experience of running out of the reset reagent?
    Any idea about the low sequence quality observed from the first cycle of new primer?
  • bioits
    Member
    • Aug 2010
    • 23

    #2
    Since there is no cleave step after last cycle of each primer, if you run out of reset buffer, any further ligation won'thappen and most likely the instrument will just repeat to scan the last cycle of the primer before reset. What you can do is to get a new bottle of reset buffer and rerun the primer with low quality as far as you haven't clear the run and slide is still sitting on the instrument.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      Originally posted by eoh001 View Post
      Hello
      I observed a strange sequence result for two times.
      The run seemed OK until a certain point, and sequnce quality collapse and the low sequence quality retained until the end of the run.
      Two Times !!!
      I found that the sequnce quality drop down right after resetting new primer. and I found that there was a reset process right before applying new primer from the log file.
      I checked the reset reagent, and the reset reagent bottle was almost empty!!!
      I am not sure that the low sequence quality mainly from the shortage of the reset reagent, but I am kind of positive.

      Anyone has an experience of running out of the reset reagent?
      Any idea about the low sequence quality observed from the first cycle of new primer?
      Deleting the comment. Did not see that the original comment referred to a SOLiD machine.
      Last edited by GenoMax; 07-07-2011, 10:21 AM. Reason: Wrong forum post

      Comment

      • SeqAA

        #4
        There is a list of required volumes for each buffer in the appendix of the instrument manual. You sometimes have to add more reset depending on the type of run.

        Comment

        • asaleh
          Member
          • Jun 2011
          • 16

          #5
          Originally posted by SeqAA View Post
          There is a list of required volumes for each buffer in the appendix of the instrument manual. You sometimes have to add more reset depending on the type of run.
          Agreed. I have noticed also that the reset buffer get sucked up a little faster than the manual states on my solid 4. However I think it is always a good idea to pause at a safe pause point and check your buffer levels if you think any of your reagents will come close to running out based on your calculations at the beginning of your run.

          Comment

          • sciencegnome
            Junior Member
            • Oct 2010
            • 4

            #6
            As a general rule we change the bottle of reset buffer when we change the tags. We have a huge surplus of that particular buffer so "when in doubt, change it out".

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            27 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            20 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            37 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            44 views
            0 reactions
            Last Post SEQadmin2  
            Working...