  • 5500xl run time -- two flowchips vs one

    Any experience regarding the relative run time for one flowchip or two flowchips in parallel on a 5500xl? ie is there a major runtime advantage in running two flowchips in parallel compared with a single flowchip on back to back runs?

  • #2
    Hi Jimmy,

    We started last week our first run with 2 flowchips in parallel on a 5500XL sequencer and the run is supposed to last 18 days for the conditions we setup. For the same conditions, one single chip lasts 11 days.

    Running 2 chips in parallel is better if you consider the time your reagents will be unfrozen (if you use "double volume" reagents) and so their conservation, compared to 2 runs of 1 flowchip.

    D.

    Comment

    Latest Articles

    Collapse

    • seqadmin
      Essential Discoveries and Tools in Epitranscriptomics
      by seqadmin




      The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
      Yesterday, 07:01 AM
    • seqadmin
      Current Approaches to Protein Sequencing
      by seqadmin


      Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
      04-04-2024, 04:25 PM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by seqadmin, 04-11-2024, 12:08 PM
    0 responses
    55 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 10:19 PM
    0 responses
    52 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 09:21 AM
    0 responses
    45 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-04-2024, 09:00 AM
    0 responses
    55 views
    0 likes
    Last Post seqadmin  
    Working...
    X