    Data transfer can overheat SOLiD

    If you transfer result files from your Solid during a run, you might be overheating your instrument. Meaning that processes impacted by temps being out of range could suffer.

    Due to climate control issues in our lab, we keep a thermometer/hygrometer probe inside our Solid, just in front of the chiller. This morning our lab was unseasonably warm, but still within the temperature specifications of the Solid (20 - 24 oC). However, inside the instrument the temps were quite a bit higher -- approaching 28 oC!

    I quickly turned on our extra chiller and some of our fans. This helped, but what was the source of this heat? Turns out we were copying about 130 gigabytes of data off the head node to our downstream processing cluster over a 1 gigabit/second network connection.

    Unlike the larger RAID array blade, (MD1000), the head node is high in the rack, nearer to the instrument. Its drives apparently produce enough heat that either it conducts through the instrument to the "wet" part of the instrument above, or the exhaust from the head node is getting sucked into the instrument though the (vortex?) fan of the reagent chiller block.

    Either way, something you might want to consider if you do image processing on your instrument cluster then transfer the .csfasta and .qual files off the instrument. You might want to limit your transfer rates if the instrument is still running!

    --
    Phillip
  • KevinLam
    Senior Member
    • Nov 2009
    • 204

    #2
    Gosh that's one thing I would never have thought possible.
    http://kevin-gattaca.blogspot.com/

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
      by SEQadmin2



      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
      ...
      10-14-2026, 09:55 AM
    • SEQadmin2
      New Genomics Technologies Take Aim at Long-Standing Limits
      by SEQadmin2


      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
      ...
      09-28-2026, 10:25 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, 10-22-2026, 11:13 AM
    0 responses
    60 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-22-2026, 10:48 AM
    0 responses
    42 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-15-2026, 10:52 AM
    0 responses
    75 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-14-2026, 05:39 AM
    0 responses
    74 views
    0 reactions
    Last Post SEQadmin2  
    Working...