  • epistatic
    Senior Member
    • Mar 2009
    • 129

    EZ Bead experiences

    I found out we are getting a SOLiD 5500. I have heard mixed reviews of the new EZ bead system, the sales guy even said the early customers had problems. I am not sure if he is setting low expectations for a product that does not work well or if it has improved. Any input from experienced users is greatly appreciated.
  • chenglei
    Junior Member
    • Nov 2010
    • 4

    #2
    Our lab has 3 sets of EZbead at Hongkong. Most problems mainly come from the Emulsifier, the oil master mix made by this machine is not stable for reaction and make the WFA result very poor.

    Comment

    • VanessaS
      Member
      • Nov 2009
      • 49

      #3
      We have two ezbead systems, love them and swear they are the best money ever spent. I've had zero problems with the machines.
      Make sure however, that until further notice, you are using the modified protocol for making the oil, where you use syringes and measure by volume not weight. My understanding is that the weighing protocol is highly dependent on your scale and many people had problems with the new protocol, so AB now has a protocol where you take the kit and make 6 tubes of the oil like we used to do before EZBead. I do that, vortex then combine and throw on the emuslifier for good measure.
      Since doing this we've never had a problem and get fabulous, consistent beads from the EZBead system.

      Comment

      • epistatic
        Senior Member
        • Mar 2009
        • 129

        #4
        That is great to hear, thanks!

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          I heard of an instance where 50 ml conical centrifuge tubes used to mix or hold some reagents for EZ Bead emulsion were cited as the culprit for poor results. (Possibly problematic "releasing agents" coating the interior of the tubes detrimental to the emulsion?)

          So, that said, maybe someone getting good results with the EZ Bead (VanessaS?) could tell us what type of syringes they are using to measure oil in the modified protocol?

          --
          Phillip

          Comment

          • VanessaS
            Member
            • Nov 2009
            • 49

            #6
            We use the Falcon 50 ml conical tubes and the BD slip tip syringes.

            Comment

            • VanessaS
              Member
              • Nov 2009
              • 49

              #7
              For those of you not getting good results, is your emulsion broken after ePCR?

              Comment

              • pmiguel
                Senior Member
                • Aug 2008
                • 2328

                #8
                Hi Vanessa,

                We have only done 2 runs on our EZ Beads -- both the same library. We have not had any EZ Bead training, so I'm not sure we would recognize broken emulsions. In plates you look for "freckling" at the bottom of the wells. Not sure what the equivalent would be for a giant bag of beads.

                --
                Phillip

                Comment

                • VanessaS
                  Member
                  • Nov 2009
                  • 49

                  #9
                  Its kind of the same but on a bigger scale. You would see a line of the beads at the bottom and a bit of a gradient going up from there. When its not broken, and is just freshly done, its a nice pale tan-peachy color and is uniform.
                  If you let it sit over night it does separate a bit but doesn't have the bead dropout.

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Originally posted by VanessaS View Post
                    For those of you not getting good results, is your emulsion broken after ePCR?
                    Hi VanessaS,
                    I spoke with someone yesterday who saw an instance where an EZ Bead was not giving good results. The emulsion was broken. It was tracked down to 50 ml conicals being used, as I mentioned above.

                    --
                    Phillip

                    Comment

                    • VanessaS
                      Member
                      • Nov 2009
                      • 49

                      #11
                      Do you know what brand, so everyone can steer clear?

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        I don't know, but probably best to only purchase the ones recommended by AB -- I believe those are the Falcon tubes that you use.

                        Comment

                        • PersianExcurzion
                          Junior Member
                          • Feb 2011
                          • 6

                          #13
                          But how long?

                          I see that AB claims about an hour of hands on time, but how long does the total run take, and is the hands on time all at once, or is it spread out between steps of each machine?

                          Thanks!

                          Comment

                          • epistatic
                            Senior Member
                            • Mar 2009
                            • 129

                            #14
                            I would calculate over an hour of hands on time for the emulsifier and enricher. The amplifier is just hanging a bag in a sideways t-cycler.

                            The quick reference guides will give you a good idea of actual effort required for each step. The EZ bead system is the use of the Covaris, Emulsifier, Amplifier, and Enricher. Lots of fun compared to the cBot.

                            EZ Bead Emulsifier
                            Thermo Fisher Scientific enables our customers to make the world healthier, cleaner and safer. Delivering technology, pharmaceutical and biotechnology services.


                            EZ Bead Amplifier
                            Thermo Fisher Scientific enables our customers to make the world healthier, cleaner and safer. Delivering technology, pharmaceutical and biotechnology services.


                            EZ Bead Enricher
                            Thermo Fisher Scientific enables our customers to make the world healthier, cleaner and safer. Delivering technology, pharmaceutical and biotechnology services.

                            Comment

                            • VanessaS
                              Member
                              • Nov 2009
                              • 49

                              #15
                              Hmmm. So you gotta make the oil, that takes 15 min or so, and have to let that degas for 30 min. You have to defrost your ePCR stuff, which can be going at the same time. You need to dilute your library, so however long that takes you. Prepping the bag takes about 20 min after the oil is ready. Making the emulsion is 15 min.
                              I'd say its a fair statement that it takes an hour of prep time to get things onto the amplifier. Thats like a 3.5-4 hour run, which can be done overnight. It takes about 10-15 min to set up the enricher including transferring the beads. The enricher takes another 4 ish hours.
                              Its all easy-peasy and makes life in the lab just a bit more rosy.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                                by SEQadmin2


                                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                                ...
                                06-02-2026, 10:05 AM
                              • SEQadmin2
                                Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                                by SEQadmin2


                                With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                                Introduction

                                Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                                05-22-2026, 06:42 AM
                              • SEQadmin2
                                Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                                by SEQadmin2

                                Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                                Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                                05-06-2026, 09:04 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 06-02-2026, 12:03 PM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 11:40 AM
                              0 responses
                              14 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 05-28-2026, 11:40 AM
                              0 responses
                              29 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 05-26-2026, 10:12 AM
                              0 responses
                              31 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...