Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • BBoy
    Member
    • Oct 2010
    • 50

    #31
    Originally posted by clivey View Post
    Thank you for spotting that, it's nice to meet someone with insight.

    C.
    Yes, that is an important but fairly trivial point which does not seem to have precluded people from skewering previous single-molecule technologies, nor investors for suing the respective companies. The averaging over hundreds of thousands of strands at each site is the very same thing that is liming read lengths in the cluster technologies.

    However, the original question was whether this is acceptable for clinical applications. And this is where one has to wonder - what good is a 10kbp read if you only care about 200 bp? If you have to read the 10kbp 50x to get the right accuracy on that 200bp, then the lack of massive parallelization with ONT would seem to start working against them. This would not seem to be ONT's strong point. Am I wrong here?

    Finally, on the question of deconvoluting over 3 bases/64 levels. In the limited time I have played with deconvolving signals this is about on the hairy edge of what is doable with a signal with a few percent noise in it and probably explains why ONT has, for the moment, stayed away from 5+ bases. Without seeing an actual "heartbeat trace" it is difficult to judge whether the sated 95% accuracy is typical or best case. Sooner or later the data will be in the wild and we will know. For the moment it pays to remember that there are lies, damn lies, and statistics... and then there are conference papers, particularly at conferences with a strong industrial presence :-)

    As I said elsewhere, if only 75% of what they claim is true the achievement is still impressive. The best part for them commercially is the ability to dip your toes and play with the instrument for a relatively low upfront cost. This is also the biggest risk, as the barrier to exit are just as low as the barrier to entrance. If they have overhyped this the reaction will be swift and rather merciless.

    Comment

    • clivey
      Member
      • Jul 2008
      • 24

      #32
      BBoy. You weren't at the talk so I will clairy if I can.

      You raise the point about parallelism a lot. What you fail to consider is speed. Not all sensing circuits are the same. On a Nanopore system, the chemistry is not cyclical and not synchronized, its free running. How many bases per second you can measure depends on how quickly you can read the channel and at what noise. 8k channels at 1000 bases per second per channel is in fact more data than a million sensors at 1 base per second per channel (obviously). Not all chips or circuits are the same. And there are significant constraints on the kinds of circuits you can pack onto silicon without making trade offs in speed and noise. Theres the rub in terms of sensor design. If your noise is too high, at the given sample rate, you won't be sequencing. One answer proposed elsewhere for this is to "floss" DNA and read it several times. Good, but of course if you read it 4 times that's then like running 1/4 the number of sensors once in terms of throughput.

      So with a Nanopore systems parallelism is only half the story -- speed (and signal to noise at that speed) are the other. Both are important, not just the parallelism. A sensor needs to be judged on both.

      The other important and often overlooked feature is sensor lifetime. Small volume wells or bubbles won't last very long. Minutes or hours. You won't get much data from that. Larger volume systems run for longer and give more data per chip, lowering cost per base.

      Any of the rules that apply to cyclical chemistries, like density of features, are not quite the same on a real time system. Nanopore systems are very different.
      Last edited by clivey; 02-25-2012, 02:16 AM. Reason: Clearer wording

      Comment

      • clivey
        Member
        • Jul 2008
        • 24

        #33
        Originally posted by GW_OK View Post
        If I recall correctly from his talk he mentioned 25% of the wells (pore binding sites?, array positions? not quite sure what to call them...) had a single pore.

        edit : one could call them membranes, I suppose.
        You guys ! Tsk!

        What I actually said was that we use a 4:1 input multiplex for each circuit. So 4 array wells per circuit. When you poisson load 4 wells, 1 gets no pores, 1 gets 1, 1 gets 2 and 1 gets 3 - on average of course. The array can then switch the circuit to read the well with 1 pore in it, ignoring the others. So, in fact, beating poisson w.r.t the mapping of pores to circuits and ensuring every circuit is being used. So when we say 8k pores, it means just that, single pores being read.

        I can see the are a lot of questions and misconceptions which we can easily answer. If you have further questions please email me directly from your institutions email account.
        Last edited by clivey; 02-25-2012, 02:15 AM. Reason: Clearer wording

        Comment

        • gringer
          David Eccles (gringer)
          • May 2011
          • 845

          #34
          Originally posted by clivey View Post
          I think this is the last misconception I'm going to answer here. If you have further questions or assertions please email me directly from your institutions email account.
          Aww, I wanted to hear that the lambda sequencing was actually done by 5 passes on an Ion Proton.
          Last edited by gringer; 02-25-2012, 08:58 AM. Reason: Removed irrelevant quote that was subsequently removed

          Comment

          • krobison
            Senior Member
            • Nov 2007
            • 734

            #35
            Originally posted by larissa View Post
            That's one of the reasons why 4% error rate is not acceptable in the clinic, for guiding treatment/dosing and inclusion/exclusion criteria. It will be acceptable for lots of other things, all R&D oriented, in academia and private sector.
            Two things to note

            1) "The clinic" is not monolithic. You seem to use this as a poor shorthand for "calling missense/nonsense and short indel mutations". There are a number of other applications with clinical value, such as detecting CNVs, chromosomal rearrangements, and transcription states, for which 4% base calling error would be quite tolerable.

            2) The 4% error rate is dominated by indels in specific contexts; for assays looking for missense/nonsense mutations outside these contexts the system might be acceptable; simply toss any reads which show an indel in the neighborhood of what you are interested in. This is a strategy not unheard of in the 454/Ion Torrent world, due to their indel issues.

            Comment

            • BBoy
              Member
              • Oct 2010
              • 50

              #36
              Originally posted by clivey View Post
              BBoy. You weren't at the talk so I will clairy if I can.

              You raise the point about parallelism a lot. What you fail to consider is speed.
              Let me clarify that I appreciate the speed and asynchronous nature of the readout very much, and perhaps by continuously nitpicking on paralelism I am leaving the wrong impression of how I view ONT's announcement. The whole technology is rather impressive sounding, even if (probably) somewhat puffed up for presentation reasons :smile:

              However, there are certain applications where parallelism matters and others where long reads do. The trivial example is a 100-page book that is read "competitively" by 100 people reading 1 pg/min and 1 person reading 50 pg/min. If you want the contents of the whole book you definitely want the latter, it will be much easier to piece the storyline together. However, if you want the contents of a single page the former is preferable. This is something that pretty much everyone on Seqanswers appreciates.

              Things get a bit more interesting when you increase the size of the book, throw in errors, and introduce a random selection of segments. If you are interested in a certain short stretch of pages the statistical advantage of massively parallel short & slow reads becomes considerable. It is my understanding that this is what clinical applications are all about, and this was the context in which I made my latest remark. Several people have already stated that if ONT's technology is anywhere close to what was presented it is likely to thrive by creating its own niches of new applications. However, displacing short-reads does not seem to be one of them.

              Originally posted by clivey View Post
              So with a Nanopore systems parallelism is only half the story -- speed (and signal to noise at that speed) are the other. Both are important, not just the parallelism. A sensor needs to be judged on both.
              Absolutely. Both = throughput, and for many applications this is all that matters. Walmart makes a ton of money over billions of transactions where they make a few cents on each. Oracle makes its money by making thousands of dollars on thousands of transactions.

              However, for certain applications the metrics are different, and this is where I find ONT's "run until" marketing a bit over the top. If you are after only a certain information then the error of the read can matter, and long reads on a randomly cleaved strand can be a disadvantage when the accuracy is <100%.

              Originally posted by clivey View Post
              The other important and often overlooked feature is sensor lifetime. Small volume wells or bubbles won't last very long. Minutes or hours. You won't get much data from that. Larger volume systems run for longer and give more data per chip, lowering cost per base.
              This is not inherent, but a characteristic of the particular sensor technology. I suspect that you are referring to nanopore technology specifically.

              In any case, thanks for taking the time to write. Your presence in these debates is much appreciated, and very different from the approach other companies are using.

              Comment

              • nxgsqq
                Junior Member
                • Jan 2011
                • 3

                #37
                Hi Clive,

                If your membranes break (or pores clog) can you destroy and reform them? I don't know your email to write to you directly.

                Comment

                • BBoy
                  Member
                  • Oct 2010
                  • 50

                  #38
                  Originally posted by nxgsqq View Post
                  I don't know your email to write to you directly.
                  As Randy Pausch said

                  The brick walls are there for a reason. The brick walls are not there to keep us out. The brick walls are there to give us a chance to show how badly we want something. Because the brick walls are there to stop the people who don’t want it badly enough. They’re there to stop the other people.
                  You can either PM him here for his email, try through LinkedIn and the additional information he provides there, or just guess. If you know one email address at a company you more or less know them all, they typically conform to a strict pattern.

                  Comment

                  • GW_OK
                    Senior Member
                    • Sep 2009
                    • 411

                    #39
                    Originally posted by clivey View Post
                    You guys ! Tsk!
                    In my defense you were talking crazy fast (and in your defense you only had 17 minutes of talk time).

                    I don't suppose there's any chance of the AGBT talk being released and/or re-done for release?

                    Comment

                    • JamesH
                      Member
                      • Aug 2010
                      • 19

                      #40
                      I've been thinking more about the disruption that the nanopore tech could cause, there's been a bit of discussion regarding the "niche" of de novo genome assembly, and granted this will be the first thing that people want to get their hands on a system to do (because of the long reads). But for me, "run until", and no library prep are what will probably make the bigger difference, you don't have to just do long reads: there's no reason not to put shorter fragments in. In our lab we do a lot of population genetics involving non-model plants and animals, this mostly involves genotyping a lot of individuals (usually at a cost of $10-30 per individual). While the cost of sequencing has been coming down the time/cost/hassle of library prep has been the main barrier for us in going NGS (having to maintain 100's of barcoded primers etc). Straight away I can see several ways that I can make restricted libraries for each individual, not bother with PCR and any associated errors, then run a Gb or two until sufficient coverage is reached and do genotyping by sequencing for close to the cost of what we currently do. The ability to do RNAseq gene expression studies in the lab without having to send off to a core facility (even if you have to make cDNA) will also be pretty awesome. I can also see applications for the USB stick in agriculture to allow field-based monitoring of resistance to insecticides/herbicides. I think that putting this kind of sequencing power straight into the hands of researchers is going to be a big game changer, and the more I think about it the less I can see myself doing a lot of the things that I currently do in the lab!

                      Comment

                      • BBoy
                        Member
                        • Oct 2010
                        • 50

                        #41
                        Originally posted by JamesH View Post
                        I've been thinking more about the disruption that the nanopore tech could cause, there's been a bit of discussion regarding the "niche" of de novo genome assembly, and granted this will be the first thing that people want to get their hands on a system to do (because of the long reads). But for me, "run until", and no library prep are what will probably make the bigger difference, you don't have to just do long reads: there's no reason not to put shorter fragments in. In our lab we do a lot of population genetics involving non-model plants and animals, this mostly involves genotyping a lot of individuals (usually at a cost of $10-30 per individual). While the cost of sequencing has been coming down the time/cost/hassle of library prep has been the main barrier for us in going NGS (having to maintain 100's of barcoded primers etc). Straight away I can see several ways that I can make restricted libraries for each individual, not bother with PCR and any associated errors, then run a Gb or two until sufficient coverage is reached and do genotyping by sequencing for close to the cost of what we currently do. The ability to do RNAseq gene expression studies in the lab without having to send off to a core facility (even if you have to make cDNA) will also be pretty awesome.
                        Yes, this is something that in my sniping about parallelism I seriously overlooked. There is nothing that prevents you from heavily fragmenting DNA and running short reads. In fact, there is nothing preventing you from running PCR and then sequencing if you have a specific need for a specific short segment. Shows you how easily I develop myopia...

                        Originally posted by JamesH View Post
                        I can also see applications for the USB stick in agriculture to allow field-based monitoring of resistance to insecticides/herbicides. I think that putting this kind of sequencing power straight into the hands of researchers is going to be a big game changer, and the more I think about it the less I can see myself doing a lot of the things that I currently do in the lab!
                        Absolutely, as long as the bio-informatics eventually are brought up to iPhone level, i.e., a big green and a big red button with little thinking in between.

                        Comment

                        • Jeremy
                          Senior Member
                          • Nov 2009
                          • 190

                          #42
                          Originally posted by clivey View Post
                          I can see the are a lot of questions and misconceptions which we can easily answer. If you have further questions please email me directly from your institutions email account.
                          Many of us are very interested in the discussion and your answers. Why not continue them here instead of answering the same questions over and over by email?

                          I wasn't able to attend the talk, but am interested in how much sequence you can expect out of a single gridION cartridge. From what I have read on the site, if you want to maximise the amount of sequence you set it to run until read quality falls below X. How many days would that typically take (assuming you set x to whatever would be reasonable for a large (2-3 Gb) whole genome assembly) and what sort of sequence volume are we talking?

                          Comment

                          • gringer
                            David Eccles (gringer)
                            • May 2011
                            • 845

                            #43
                            How many days would that typically take (assuming you set x to whatever would be reasonable for a large (2-3 Gb) whole genome assembly) and what sort of sequence volume are we talking?
                            In my post on the "disruptive technology" thread, I came up with about 200kb per pore per hour:

                            Sequencing technologies without a commercially released platform (Oxford Nanopore, Halcyon Molecular, etc.)


                            This was from an estimate of GridION processing speed of 10Gb per day on 2000 pores ("tens of gigabases per day"). Just for comparison, using this 10Gb estimate the MinION (which has a run-time limit of about 6 hours, and 1/4 the number of pores) should be able to process about 600Mb before the reagents run out.

                            Comment

                            • Jeremy
                              Senior Member
                              • Nov 2009
                              • 190

                              #44
                              10Gb per day does sound very impressive. But everything I have read still only says a run can take "days" and that you just stop it when you are ready. It doesn't actually state how many days you can get good quality data from a single cartridge assuming enough good quality sample is available. That's a pretty significant factor if you want to sequence a whole genome.

                              Comment

                              • gringer
                                David Eccles (gringer)
                                • May 2011
                                • 845

                                #45
                                Originally posted by Jeremy View Post
                                It doesn't actually state how many days you can get good quality data from a single cartridge assuming enough good quality sample is available. That's a pretty significant factor if you want to sequence a whole genome.
                                This is not particularly important. Quoting myself from this thread:
                                Samples can be ejected during a run and shifted to other cartridges, cartridges can be replaced, runs can be started at any time, and stopped due to a number of different desired factors, so any drop-off in occupancy probably won't have too much impact on result output (as long as there's still money to burn).
                                As long as you have money and monkeys, you can sequence as many bananas as you want. With the "release date" 2000-pore GridION, a 20-node system should do a human genome in an hour (you'd need to split the sample into 20 for this), or a 5-node system in 4 hours:



                                I guess that's assuming 5-20x coverage for the human genome, depending on how you interpret "tens of gigabases", and also suggests a genome should be possible in one day on a single node with reasonable coverage.

                                With 20 MinIONs, it would take about 4 hours assuming similar processing speed.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                  by SEQadmin2




                                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                  10-30-2026, 06:46 AM
                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-28-2026, 09:57 AM
                                0 responses
                                39 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-28-2026, 09:51 AM
                                0 responses
                                23 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 11:13 AM
                                0 responses
                                73 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 10:48 AM
                                0 responses
                                48 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...