    What fragment size do you use for your Illumina libraries?

    Illumina recommends insert sizes of ~200bp for SE libraries and ~300bp for PE libraries but clearly indicates in their literature that you can go larger (500-600bp). I'd like to ask the NextGen community a couple of questions:

    1. What insert size do you routinely use for your SE & PE libraries?

    2. Have you noticed any relationship between overall sequence quality and library fragment size? (Actual data correlating these would be terrific.)

    Thanks.
  • kainsteven
    Member
    • Jul 2009
    • 16

    #2
    For RNA-Seq applications, we have seen better results by excising material corresponding to the peak of the bioanalyzer trace for library preparation. In most cases, this corresponds to ~ 250 – 350 bp in length. We have used this range for both SR and PE libraries, but only with short read lengths of 32-36 bp. For longer read lengths and PE sequencing, I would imagine you would want to lean towards the longer end of this range.

    Best,
    Steve

    Comment

    • greigite
      Senior Member
      • Mar 2009
      • 145

      #3
      I don't have any data on this myself, but this paper reports the successful use of a 600 bp library, which reduced overall coverage variability: Harismendy & Frazer 2009

      Comment

      • maureen
        Junior Member
        • Feb 2010
        • 3

        #4
        Solexa recommends 200-250 bp and a tight range, but I definitely know of successful routine sequencing of libraries 300-700 bp. In fact my own library is 300-600 bp.

        Comment

        • der_eiskern
          Member
          • Jul 2009
          • 46

          #5
          illumina may recommend smaller library sizes but i'm pretty sure there are people using 1kb or 2kb insert sizes in parallel with the 200-700 bp sizes for structural variant discovery.

          Comment

          • edge
            Senior Member
            • Sep 2009
            • 199

            #6
            Hi kainsteven,

            I have an Illumina RNA-seq, pair-end read, 2x90bp....
            However, I not really sure about the fragment size/insert size
            It might affect my assembly result.

            Based on your experience, do you have any idea regarding the fragment size of Illumina pair-end RNA-seq data if my read length is 90nt.

            Many thanks for advice.

            Originally posted by kainsteven View Post
            For RNA-Seq applications, we have seen better results by excising material corresponding to the peak of the bioanalyzer trace for library preparation. In most cases, this corresponds to ~ 250 – 350 bp in length. We have used this range for both SR and PE libraries, but only with short read lengths of 32-36 bp. For longer read lengths and PE sequencing, I would imagine you would want to lean towards the longer end of this range.

            Best,
            Steve

            Comment

            • RPiddock
              Lab Monkey
              • Jun 2011
              • 16

              #7
              Our libraries team generally shear for about 500bp with Illumina TruSeq libraries.

              Comment

              • edge
                Senior Member
                • Sep 2009
                • 199

                #8
                Thanks for replying
                Is there got any other way to inferred insert size by undergo some statistic analysis?
                Thanks.

                Originally posted by RPiddock View Post
                Our libraries team generally shear for about 500bp with Illumina TruSeq libraries.

                Comment

                • tony19871929
                  Junior Member
                  • Feb 2012
                  • 1

                  #9
                  question about illumina seq.

                  hello every one , i really want to know the different Illumina standard (300bp insert) from overlapping 180bp.? can you help me ,thank you

                  Comment

                  • kainsteven
                    Member
                    • Jul 2009
                    • 16

                    #10
                    Thank you for the question. As mechanical or enzymatic fragmentation methods will produce a distribution of fragment sizes we typically choose a median size of 350 to 400 bp for 2 x 90 bp reads. There is no harm is going to this length, and it reduces the likelihood of any forward and reverse sequencing overlaps.

                    Steve

                    Comment

                    • apeterson
                      Junior Member
                      • Jan 2011
                      • 3

                      #11
                      Originally posted by der_eiskern View Post
                      illumina may recommend smaller library sizes but i'm pretty sure there are people using 1kb or 2kb insert sizes in parallel with the 200-700 bp sizes for structural variant discovery.
                      Does anyone have any more information on making 1kb insert size libraries?

                      Comment

                      • cement_head
                        Senior Member
                        • Mar 2012
                        • 264

                        #12
                        Originally posted by edge View Post
                        Hi kainsteven,

                        I have an Illumina RNA-seq, pair-end read, 2x90bp....
                        However, I not really sure about the fragment size/insert size
                        It might affect my assembly result.

                        Based on your experience, do you have any idea regarding the fragment size of Illumina pair-end RNA-seq data if my read length is 90nt.

                        Many thanks for advice.
                        It should not affect your assembly, IF you have enough (ideal) coverage. What will affect your assembly is if you have overlapp in your SBS. For example, if your fragment length was 300 bp and your paired-end reads were 200 bp you will have a problem because the reads actually have 100 bp overlap and the assembler will have difficulty discerning whether or not it's an overlap or two different reads.

                        What you need to calculate, or consider, is how much of the end of the fragment that you need to sequence in order to identify it as unique. In a haploid organism that answer is not very much. In a diploid organism that answer is more complicated. You have to consider variants (like alleles) and how you will be able to assign reads to a given gene. Then you might consider SBS to within a few bases of each other (e.g. paired-end sequencing of 140 bases of SBS on a 300 bp fragment) to cover most of the fragment.

                        How did you prepare your library? If you used Nextera, you have sizes that range from 300 to 1400 bp. If you used disruption, you should have a much narrower distribution of sizes.

                        Hope this helps,
                        Andor

                        Comment

                        • chiaraf
                          Member
                          • Oct 2012
                          • 10

                          #13
                          Hi!
                          I routinely sequence libreries with a mean fragment size of 450 bp for a 100PE for DNA and libreries witha mean fragment size of 350-400 for a 100PE for RNA seq.

                          Comment

                          • pc2009open
                            Junior Member
                            • Jun 2009
                            • 3

                            #14
                            Hi everyone,

                            I am also interested in 1Kb-1.5Kb library. Do you know any lab that has the expertise?

                            Thanks a lot.



                            Originally posted by apeterson View Post
                            Does anyone have any more information on making 1kb insert size libraries?

                            Comment

                            Latest Articles

                            Collapse

                            • SEQadmin2
                              Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                              by SEQadmin2


                              I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                              Here are nine questions we think about, in roughly the order they matter, before...
                              06-18-2026, 07:11 AM
                            • SEQadmin2
                              From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                              by SEQadmin2


                              Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                              The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                              ...
                              06-02-2026, 10:05 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by SEQadmin2, Today, 11:10 AM
                            0 responses
                            6 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 06-17-2026, 06:09 AM
                            0 responses
                            41 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 06-09-2026, 11:58 AM
                            0 responses
                            102 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 06-05-2026, 10:09 AM
                            0 responses
                            124 views
                            0 reactions
                            Last Post SEQadmin2  
                            Working...