    Hello all,

    I have received paired-end reads for 40 samples.
    The reads are supposed to be 100bp per end.
    Instead, 20 of my samples are 101bp per end, while the other 20 are 100bp as expected.

    Because of this, we assume that these 20 samples were all in the same lane; and that by accident there was an extra iteration in the illumina sequencing.
    However, we also see a strong negative correlation with read length and quality; the samples in which we had 101bp per end lose about 30% of the reads in the trimmomatic quality step.

    My question is: Has this happened to anyone else? Does it occur often, and if so, does it often affect quality? We are really quite puzzled by this.

    Any ideas / clues appreciated.

    Thank you!
    -Thies Gehrmann
  • mastal
    Senior Member
    • Mar 2009
    • 666

    #2
    RNASeq: Read length different from expected

    The read length of 101 bp instead of 100 bp would probably be because the operator thought there were enough reagents left to sequence the extra base.


    Presumably the run that gave the 101 bp reads had some problems that resulted in poorer base qualities than the run or lane that gave the 100 bp reads.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      That is odd. I do not think one can set up lanes on a single flowcell to run for different number of cycles.

      Have you checked with your provider to confirm if all the samples ran together (were they multiplexed) or if there was a batch difference (2 separate runs?) that can account for your observation. I think it would probably be the latter case.

      Comment

      • mastal
        Senior Member
        • Mar 2009
        • 666

        #4
        The read identifiers for each sample should tell you if they were on the same flow cell or lane.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Originally posted by mastal View Post
          The read identifiers for each sample should tell you if they were on the same flow cell or lane.
          Great point. This should be easy to check.

          Comment

          • kmcarr
            Senior Member
            • May 2008
            • 1181

            #6
            It is somewhat standard practice for Illumina sequencing that if you want read lengths of N bases you run N+1 cycles. This has to do with the way base calling works on Illumina; to properly call the base at position n in a read you need data from cycle n+1. The last base in a read will always have a lower Q-score reflecting the added uncertainty in the base call. To mitigate this you run N+1 cycles but just report N bases per read, dropping the last, low quality base. This practice is ingrained in the Illumina run recipes; the standard PE100 recipe (with indexing) on the HiSeq runs a 209 cycles (101 + 7 + 101) adding an extra cycle each to read 1, the index read and read 2. (Interestingly the MiSeq recipes still add the extra cycle to reads 1 & 2 but not to the index read.)

            Some core labs simply call and report all 101 cycles, some stick to the original practice of clipping the last base. It may be that your samples were all run together on the same flow cell for 2x101 cycles but for one set of 20 they reported all 101 cycles and the other they clipped the last base. You need to check the IDs of your samples to identify the flow cell and lanes used for each.

            Comment

            • gogodidi
              Junior Member
              • Aug 2012
              • 4

              #7
              Originally posted by mastal View Post
              The read identifiers for each sample should tell you if they were on the same flow cell or lane.
              Indeed, they were!
              Thank you (Don't know why it didn't occur to me).

              Comment

              • gogodidi
                Junior Member
                • Aug 2012
                • 4

                #8
                Originally posted by kmcarr View Post
                It is somewhat standard practice for Illumina sequencing that if you want read lengths of N bases you run N+1 cycles. This has to do with the way base calling works on Illumina; to properly call the base at position n in a read you need data from cycle n+1. The last base in a read will always have a lower Q-score reflecting the added uncertainty in the base call. To mitigate this you run N+1 cycles but just report N bases per read, dropping the last, low quality base. This practice is ingrained in the Illumina run recipes; the standard PE100 recipe (with indexing) on the HiSeq runs a 209 cycles (101 + 7 + 101) adding an extra cycle each to read 1, the index read and read 2. (Interestingly the MiSeq recipes still add the extra cycle to reads 1 & 2 but not to the index read.)

                Some core labs simply call and report all 101 cycles, some stick to the original practice of clipping the last base. It may be that your samples were all run together on the same flow cell for 2x101 cycles but for one set of 20 they reported all 101 cycles and the other they clipped the last base. You need to check the IDs of your samples to identify the flow cell and lanes used for each.
                Thank you for this information.
                It is strange then, that they cut it for one lane, and not for the other.
                I'll have to call them to ask about that.

                Thank you so much, all of you!

                Comment

                Latest Articles

                Collapse

                • GATTACAT
                  Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                  by GATTACAT
                  Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                  07-01-2026, 11:43 AM
                • SEQadmin2
                  Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                  by SEQadmin2


                  I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                  Here are nine questions we think about, in roughly the order they matter, before...
                  06-18-2026, 07:11 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Today, 11:05 AM
                0 responses
                6 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-02-2026, 11:08 AM
                0 responses
                27 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-30-2026, 05:37 AM
                0 responses
                25 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-26-2026, 11:10 AM
                0 responses
                25 views
                0 reactions
                Last Post SEQadmin2  
                Working...