    Did mRNA capture (Dynabead) for my RNA extract samples (mushroom). When run mRNA pico assay for QC, these kind of result appeared. Sometime it work for some of the samples. But repeated assay using same sample seem to produce different result. Contacted field application specialist and made the listed changes but problem persists.

    Changes made:

    1. washed electrode with RNAzap and consequently with RNase free water and air dried for more than 10 minutes. Even diagnosed for short circuit before run.

    2. check priming station for failure. It seem to be working because it bounce normally when the latch was released

    3. my sample were treated with DNase from MoBio.

    4. diluted my samples.

    Attached here are pdf. containing results for last few run.

    2100 expert_mRNA Pico_DE13805702_2016-06-27_18-03-51.pdf

    2100 expert_mRNA Pico_DE13805702_2016-06-14_17-28-00.pdf

    2100 expert_mRNA Pico_DE13805702_2016-06-27_16-40-47.pdf
  • DStephens@NuGEN
    Registered Vendor
    • Jun 2016
    • 3

    #2
    Hi sbchua.1990,
    We have seen traces such as these when the chip priming/loading was not performed correctly. Did you ensure that the chip priming station plunger was set to the highest notch and the plunger was released after exactly 30 seconds? Was the chip loaded with fresh reagents?

    A couple of other possibilities include overloading of the chip and salt or detergent carryover. You noted that you already tried diluting the samples; how much material was used? We typically load 1-5 ng/well onto the Pico chip.

    Finally, have you tried an additional clean-up of your samples?

    Hope that helps!
    Denise Stephens
    Technical Support Scientist at NuGEN Technologies

    Comment

    • sbchua.1990
      Junior Member
      • Sep 2015
      • 3

      #3
      Thanks for reply, I am pretty sure that the chip priming/loading was performed correctly. Reagents are still within expiration date.

      My samples concentration should be around 300pg - 1500pg. I just loaded 1 ul without dilution.

      No I did not try additional clean-up for my samples.

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        The samples most likely are contaminated with residual removal resin. You would need to contact MoBio for cleanup instructions.

        Comment

        • sbchua.1990
          Junior Member
          • Sep 2015
          • 3

          #5
          Hi nucacidhunter, did you encounter such problem in the past? Or what make you think so? Thanks for your reply.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            Profile indicates presence of insoluble contaminants which could be protein (residual DNase) or removal resin. I suspect the resin because it can be carried over easily if spinning was not long enough or one tried to take the supernatant without leaving any liquid behind in the tube. There is also possibility of protein protectant carry over which is used to stabilise DNase at RT.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              10-30-2026, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            41 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            25 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            74 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...