    Weird library profile

    Hello,

    We are seeing a weird library profile - double peak + some high molecular weight smear after the range we cut. It's a chip-seq library amplified for 18 cycles - first time we see this specific profile, should we sequence it?

    Thanks!

    Click image for larger version

Name:	weird-library.jpg
Views:	1
Size:	80.9 KB
ID:	308053
  • kbusseq
    Junior Member
    • Mar 2013
    • 7

    #2
    It's possible that the double peak is caused by running too much sample for the chip (I'm guessing your image is from a Bioanalyzer chip or something similar). We've seen some seriously bizarre peaks on our Bioanalyzer (some double like yours and one that was actually a triple peak) that are remedied simply by re-running the sample at a 1:5 or 1:10 dilution.

    Sometimes this eliminates all trace of a high MW smear as well, but typically I would guess that the presence of high MW fragments is due to over-amplifying the library. Do you usually amplify for 18 cycles? That is a little high in my experience (which, granted, I don't have tons of yet... ), but if it normally works for you than it might not be the issue.

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
      by SEQadmin2



      Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
      ...
      07-09-2026, 11:10 AM
    • SEQadmin2
      Cancer Drug Resistance: The Lingering Barrier to Rising Survival
      by SEQadmin2



      Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

      There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
      07-08-2026, 05:17 AM
    • GATTACAT
      Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
      by GATTACAT
      Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
      07-01-2026, 11:43 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, 07-13-2026, 10:26 AM
    0 responses
    21 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 07-09-2026, 10:04 AM
    0 responses
    32 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 07-08-2026, 10:08 AM
    0 responses
    20 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 07-07-2026, 11:05 AM
    0 responses
    34 views
    0 reactions
    Last Post SEQadmin2  
    Working...