    Bioanalyzer RNA QC

    Hi everyone

    I am ready to start my first library prep with the TruSeq v2 kit, but already got stopped after a Bioanalyzer Pico run to verify the quality of my extracted RNAs. The RIN is N/A, undetermined because of a baseline drop or increase. I was so far not able to find such a profile in troubleshooting examples and am hoping that someone here can help figuring out what is going on with some of my samples. Unfortunately, I do not have access to the instrument and software. Samples were extracted from C. elegans using Qiazol followed by RNeasy mini, Bioanalyzer file is attached (Please do not consider 23 and 24).

    Thanks a lot
    Caro
    Attached Files
  • DRYTCYV
    Member
    • Apr 2011
    • 50

    #2
    Hi,

    First of all, Ladder peaks size are not adjusted.
    Sample 3 need to be adjusted too.
    The profile looks ok, but a strange baseline (negative values).

    This is my quickest answer

    Cheers

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Yes, the person that ran your bioanalyzer chip needs to force the software to recognize the correct position of the 25 base marker in the molecular weight marker. Same thing for sample Ev-III, the dumbo software mis-called the 25 base marker peak. It needs to be forced to recognize the right peak.

      My guess is that your samples are fine. Pico chips are the worst for irregularities. But they are very sensitive...

      If you poke around on the Agilent site, you can find the Expert 2000 software available for download. It is kind of obnoxious -- has some weird licensing modules that stay resident on the computer you infect with this package. But the software itself is very powerful. If you had it, you could just ask your agilent facility for the .xad file and make the adjustments yourself.

      Myself, I feel like I have both arms tied behind my back just looking at the .pdf.
      --
      Phillip

      Comment

      • cahe
        Junior Member
        • May 2014
        • 2

        #4
        Hi
        Thank you for your comments. Seems like a rather bad service from a core facility that they don't even make adjustments to the software.
        They provided me with the .xad files, so I will try to get the software.

        Thanks
        Caro

        Comment

        • kerplunk412
          Senior Member
          • Jun 2012
          • 119

          #5
          I guess you have probably solved your problem by now, but I thought I would comment anyway in case it can help anyone else. Although you didn't get any RIN values from the BA, the 28S and 18S peaks in your samples are sharp and the 28S is roughly twice the intensity of the 18S, which indicates very high quality RNA that should work well for library prep. For comparison, sample 23 in the pdf is highly degraded and sample 24 is somewhat degraded but still probably okay for library prep. I have run a decent amount of RNA on the Bioanalyzer, and I am confident that all of your samples would have a RIN of >9.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          32 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          23 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          40 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          45 views
          0 reactions
          Last Post SEQadmin2  
          Working...