    Trouble with RNA quality

    I recently did a RNA extraction using Trizol, I followed all the steps, except I skipped the optional step. I intend to do a RNA seq with these samples using the miseq. The trouble I'm running into are the RIN numbers i get from the bioanalyzer are really bad across samples. The samples are pretty fresh (less than 3 weeks old), were immediately frozen on dry ice, and then kept in a -80 till extraction. Extractions were then also done on ice.

    I also nano dropped these samples and the 280/260 values are decent(>1.8), however, the 260/230 are very low, always less than 1.

    I was wondering if anyone can help interpret some of my traces and offer any advice as to how to fix these samples so that they are still useable.

    I attached the traces.

    Thanks,
    Mike
    Attached Files
  • sara_ahmed
    Member
    • Nov 2011
    • 13

    #2
    I think it might help to rerun these on the Bioanalyzer at a higher concentration. The FU on the y-axis seem super low and I am afraid you may just be looking at background rather than actual ribosomal peaks from the sample.

    Sara Ahmed, PhD | Director of Sequencing | Cofactor Genomics
    4044 Clayton Ave | St. Louis, MO 63110 | tel. (314) 531-4647

    Comment

    • ELoomis
      Junior Member
      • Sep 2011
      • 9

      #3
      What does your nanodrop trace look like? Low 260/230 could mean phenol in your samples (Trizol is ~50% phenol). Check out this link for a good guide to understanding nanodrop reports (scroll down for examples of phenol contamination):


      If it is indeed phenol, you can extract with chloroform again until it goes away. (Dilute your samples to a few hundred uL, but no more than half the tube volume, then add an equal volume chloroform, mix, separate phases, etc. as before, then precipitate as before)

      Hope that helps!

      Comment

      • Mike2188
        Member
        • Oct 2013
        • 27

        #4
        A few questions:

        a) what samples are you working with?
        - certain tissues have different forms of contamination, for instance I work with leaf and seed so often carbohydrates may be an issue.

        b) what yields are you obtaining (concentrations from the nanodrop?)
        - if your concentrations are high you can try a LiCl reprecipitation, or redo the extraction from the chloroform phase seperation
        - if your concentrations are low you can do a column purification

        c) are you familiar with the bioanalyzer?
        - have you performed these successfully before? IMHO I think you may have overprimed the chip, it looks like it is all over the place from the gel, and some samples have signals coming before the marker which is indicative of this. Rerun the chip, make sure that you have the injector set to the proper position (as far up as it can go) -- it may have still been set for a high sensitivity chip. Make sure you prime for 30 seconds EXACTLY. Also, make sure you mix the dye very well (10 entire seconds of intense vortexing and you mix the gel-dye extremely well (10 entire seconds of intense vortexing). Clumps of dye can also clog the chip.

        If you want you can run your RNA on a gel if your concentrations are high enough and look for the ribosomal bands.

        Also, would you be able to actually upload the .xad file from your bioanalyzer run?

        Comment

        • Mike2188
          Member
          • Oct 2013
          • 27

          #5
          Also, Sara is mistaken, your concentrations are sufficient, as they reach or exceed the marker.

          Comment

          • michael-steffen
            Member
            • Dec 2012
            • 18

            #6
            Sorry. Didn't realize I had more responses to this post.

            The tissue I am using is mental gland tissue from salamanders(pheromone gland).

            My nano drop concentrations aren't horrible(100ng/ul).

            I do think you may have helped me with the problem though. I am pretty sure I have been over priming the chip. Been doing HS chips lately, and its got a 60s injection step. I've also been doing that same time for the RNA I realize now. Going to give it another try with 30s and see if that fixes it for me.

            Thanks,
            Mike

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Yesterday, 11:13 AM
            0 responses
            16 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, Yesterday, 10:48 AM
            0 responses
            17 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            35 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            42 views
            0 reactions
            Last Post SEQadmin2  
            Working...