    I have 2 small questions concerning the 16s sequencing protocol for MiSEq:

    1) Some samples have very low concentration so we are considering loading a 2pM library instead of the more standard 4pM. Anyone having bad (low) clustering doing this or no problem when using the v3 MisEq kit (2 X 300 bp) ?

    2) There are various older threads discussing the PHIX % to be added to the library. The protocol says 5 % (v3 kits again), but the rep says 10-20 %. Any comments on this ?

    thanks

    Greg

  • #2
    Anyone having bad (low) clustering doing this or no problem when using the v3 MisEq kit (2 X 300 bp) ?
    We saw inconsistent cluster density. Illumina recommends concentration by SpeedyVac. Other options would be concentrating them using column or AMPure beads by reducing elution volume.

    The protocol says 5 % (v3 kits again), but the rep says 10-20 %. Any comments on this ?
    Illumina's current official recommendation with the latest software is 5% PhiX spike-in, but 10% spike-in increases the likelihood of successful run.

    Comment


    • #3
      Thanks for this nucacidhunter. So is 4 pM the optimal (clustering) conc for amplicon Seq ?

      cheers

      Comment


      • #4
        Illumina’s guide for clustering libraries on MiSeq can be find by google searching “Preparing Libraries for Sequencing on the MiSeq”. 4 pM is the concentration for denaturation step. Final pM concentration for clustering varies among labs as the quantification systems and practices differ. For amplicons less input improves PF% and also quality scores which usually is achieved with cluster densities below 1k. Illumina also recommends incubation of HT1 diluted denatured libraries at 96°C for 2 min before loading to improve clustering GC rich fragments

        Comment


        • #5
          I had another question concerning low cleaned amplicon concentration before indexing PCR. Can the Nextera indexing PCR cycle (8 cycles) be increased (to 10-12) without any issues to increase the concentration before final bead clean-up ? Also, what's the contamination likelihood using a SpeedVac ?

          cheers
          Last edited by BioGenomics; 07-30-2014, 12:50 AM.

          Comment


          • #6
            If concentration of amplicon PCR after clean-up is 200 pg/ul, using 5 ul in index PCR will result in over 200 ng amplicon which after clean-up will be at concentration of 4ng/ul. This amount will be equivalent to 12 nM (if you are using Illumina validate region primers). If you are not obtaining these figures, the problem might be in input DNA. Increasing cycle number may cause biases in amplification of various species DNA present in sample. In any case the metagenomics profile is always one view of community and there is no common agreement on it. If increasing amplicon concentration is the aim, it can be achieved easily by decreasing elution volume from beads.

            I do not know the likelihood of cross contamination in SpeedyVac (I think pretty low), but it should not matter because the amplicons already have been indexed.

            Comment


            • #7
              Originally posted by nucacidhunter View Post
              Illumina’s guide for clustering libraries on MiSeq can be find by google searching “Preparing Libraries for Sequencing on the MiSeq”. 4 pM is the concentration for denaturation step.
              You mean 4nM.

              --
              Phillip

              Comment


              • #8
                Yes, 4 nM. Thanks for correction.

                Comment


                • #9
                  Originally posted by nucacidhunter View Post
                  We saw inconsistent cluster density. Illumina recommends concentration by SpeedyVac. Other options would be concentrating them using column or AMPure beads by reducing elution volume.
                  Yes.
                  The real issue here is probably the concentration of NaOH in the sample loaded into the cassette. If it is too high, then it would interfere with binding of the amplicons to the flowcell.

                  I wish Illumina would bear down and create a denaturation procedure that doesn't require a subsequent 50X dilution of the sample with buffer prior to loading.

                  --
                  Phillip

                  Comment


                  • #10
                    I wish Illumina would bear down and create a denaturation procedure that doesn't require a subsequent 50X dilution of the sample with buffer prior to loading.
                    It seems that Illumina looking into different methods for denaturing libraries for clustering. For HiSeq X system, library is mixed with three solutions (PCX1-3) and RSB (without an initial denaturation) and loaded into cBot. Libraries for this system must be denatured in cBot perhaps by heating (if someone have the related cBot recipe can check this). Also, llumina’s current protocol for clustering and sequencing on NextSeq 500 includes a naturalisation with Tris-Cl pH7.0 after NaOH denaturation step.

                    Comment


                    • #11
                      Hi nucacidhunter et al, anybody has experience with using a 0.5 nM 16s library concentration (instead of 2nM) on the MiSeq ? The NextSeq500 has this protocol (updated October 2014, so more recent) and we thought to add the 200mM TRis buffer Ph 7.0 too to hydrolyse the higher NaOH concentration needed initially. Thanks

                      Comment


                      • #12
                        For subnanomolar libraries we use the protocol found in the supplemental section (Protocol 12) in the paper below. Its a fantastic protocol and I have used it for libraries with 100 pM (diluted to 4 pM for the final load) and had successful runs.



                        Nature Methods 5, 1005 - 1010 (2008)
                        Published online: 25 November 2008 | doi:10.1038/nmeth.1270

                        A large genome center's improvements to the Illumina sequencing system



                        Supplementary info;
                        Supplementary Protocol 12: Modified hybridisation buffers

                        Comment


                        • #13
                          We use the NextSeq protocol with the 200mM Tris pH 7 all of the time for our MiSeq runs.

                          Comment


                          • #14
                            Hi,
                            I am wondering for subnanomolar concentration library, what was the final volume you (acockburn) loaded on the sequencing machine? The miseq protocol says 600ul, but curious what others have tried loading. Thank you.

                            Comment


                            • #15
                              I usually make 700 uL and load 600 uL. I have never tried less.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Best Practices for Single-Cell Sequencing Analysis
                                by seqadmin



                                While isolating and preparing single cells for sequencing was historically the bottleneck, recent technological advancements have shifted the challenge to data analysis. This highlights the rapidly evolving nature of single-cell sequencing. The inherent complexity of single-cell analysis has intensified with the surge in data volume and the incorporation of diverse and more complex datasets. This article explores the challenges in analysis, examines common pitfalls, offers...
                                Today, 07:15 AM
                              • seqadmin
                                Latest Developments in Precision Medicine
                                by seqadmin



                                Technological advances have led to drastic improvements in the field of precision medicine, enabling more personalized approaches to treatment. This article explores four leading groups that are overcoming many of the challenges of genomic profiling and precision medicine through their innovative platforms and technologies.

                                Somatic Genomics
                                “We have such a tremendous amount of genetic diversity that exists within each of us, and not just between us as individuals,”...
                                05-24-2024, 01:16 PM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Today, 08:18 AM
                              0 responses
                              7 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, Today, 08:04 AM
                              0 responses
                              6 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 06-03-2024, 06:55 AM
                              0 responses
                              13 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 05-30-2024, 03:16 PM
                              0 responses
                              27 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X