    Qubit - wrong values due to GFP?

    Hi,

    I performed a Qubit measurement on genomic DNA (BR) and took 2.5 µg according to the measurement for a reaction with a GFP-tagged protein.
    Afterwards I did another measurement (HS) (because I thought due to clean-up steps I would probably have less DNA), but weirdly now I have about double the amount of DNA in there.

    I also measured the original DNA again with HS (comes out at about the same amount as BR).

    Now I was wondering if maybe residual GFP-tagged protein could influence the measurment? Does anybody here know anything about that? Or maybe experienced something similar?

    Thanks in advance! Anne
  • cliffbeall
    Senior Member
    • Jan 2010
    • 144

    #2
    I am pretty sure it will interfere, the Qubit fluorophore is green (I believe it's picogreen?). You might be able to measure the sample without adding the fluorophore and subtract that out as background?

    That being said, I wouldn't expect GFP to survive a typical DNA purification -they usually involve proteases and denaturants.

    Comment

    • bombardior
      Member
      • Oct 2014
      • 10

      #3
      that qubit reagent is orange in color, which, coincidentally, is the same color as sybr green, so i think gfp may indeed interfere with qubit readings.

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        Now I was wondering if maybe residual GFP-tagged protein could influence the measurment? Does anybody here know anything about that?
        Loss of DNA during purification is expected. Increase in measured DNA amount after clean up is most likely false. It could be due to contaminants (salts, proteins and other substances) that has not been completely removed. There is a list of common contaminants and their effect on quantification with PicoGreen reagent in the product manual: http://tools.lifetechnologies.com/co...ls/mp07581.pdf

        Comment

        • AnneH
          Member
          • Sep 2014
          • 10

          #5
          I guess the next time I use this GFP-tagged protein, I will set some of it aside and do an extra Qutbit measurement with just this protein and without DNA.

          Thanks nucacidhunter for the Link, unfortunately the only thing from that list that might still be in there could be sodium chloride, and this would lead to a decrease, not an increase :-/

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            I do not know details of your protocol, but PicoGreen normally is excited at 480 nm and fluorescence emission is collected at 520 nm. GFP unlike PicoGreen is a protein and has to be in soluble form to be active and fluoresce under correct excitation condition. You may check fluorescence spectral viewer (http://www.lifetechnologies.com/au/e...#product=P7589) to see if your GFP type can be exited and emit at the same wavelength as PicoGreen.

            Comment

            • bilyl
              Member
              • Aug 2013
              • 52

              #7
              Originally posted by AnneH View Post
              Hi,

              I performed a Qubit measurement on genomic DNA (BR) and took 2.5 µg according to the measurement for a reaction with a GFP-tagged protein.
              Afterwards I did another measurement (HS) (because I thought due to clean-up steps I would probably have less DNA), but weirdly now I have about double the amount of DNA in there.

              I also measured the original DNA again with HS (comes out at about the same amount as BR).

              Now I was wondering if maybe residual GFP-tagged protein could influence the measurment? Does anybody here know anything about that? Or maybe experienced something similar?

              Thanks in advance! Anne
              If you are really concerned, try spiking in a microliter of Proteinase K into your Qubit tube and see if that changes the signal. Try it on some "no-GFP" DNA to have your negative control.

              I wouldn't put the ProK into your main DNA stock, as it might interfere with downstream assays unless you clean it up really well.

              Comment

              • IdahoRAD
                Junior Member
                • Jan 2014
                • 9

                #8
                So, you are making sure to only consider the amount of DNA, and not the concentration? Changes in elution volumes could make it "look" like the DNA content is increasing.

                Comment

                • kerplunk412
                  Senior Member
                  • Jun 2012
                  • 119

                  #9
                  Treating some of your sample with ProtK sounds like a good idea to me, but you might want to do that in a separate tube and then just use an aliquot of that reaction for Qubit. Although just doing it in the same tube would probably work as well.

                  Also, have you tried just using the Nanodrop? It sounds like your DNA may be concentrated enough that old fashioned spectroscopy should give you a fairly accurate reading.

                  Comment

                  • AnneH
                    Member
                    • Sep 2014
                    • 10

                    #10
                    So, unfortunately I don't have anything left of that sample and I treated it with Proteinase K before the measurement.
                    @IdahoRAD: Yes of course I consider the amount. I know that I went into the reaction with about 2.5 µg and came out with about 5 µg.

                    I also measured the DNA with Nanodrop, and here I measure an even higher increase. Unfortunately i didn't write down 260/280 nm ratio :-/

                    Comment

                    • AnneH
                      Member
                      • Sep 2014
                      • 10

                      #11
                      I performed a measurement, with different input of the GFP-tagged protein (without DNA) and I can in fact measure "dsDNA" with the High Sensitivity kit.

                      It wasn't high enough though to explain my miraculous increase in DNA in this sample completly though.
                      But well at least one question answered: GFP can influence DNA measurement with Qubit.

                      Comment

                      Latest Articles

                      Collapse

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 06-09-2026, 11:58 AM
                      0 responses
                      23 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 06-05-2026, 10:09 AM
                      0 responses
                      29 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 06-04-2026, 08:59 AM
                      0 responses
                      39 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 06-02-2026, 12:03 PM
                      0 responses
                      61 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...