  • Does TE buffer impact on covaris shearing?

    Dear all,

    I used bioanlayzer after covaris shearing. I put TE buffer in DNA samples, as protocol recommends, but result was not good. I just wonder whether salt in TE buffer impact on fragment size.

    Is there anyone who use water in DNA sample in covaris shearing instead TE buffer?

  • #2
    I shear in 10mM TrisHCl 8.5, and the manufacturer recommended protocols perform pretty well.

    More often problems come from not enough volume in the micro tube (<120ul will give air bubbles and thus variability). What consumable are you using and what volume?

    Comment


    • #3
      I did covaris shearing for NibleGen exomcapture library, which recommends a total volume of 100 ul with 1X TE buffer. I mixed app. 4 ul DNA and 96 ul TE buffer in microtube (6 x 16 mm, no. 520045).

      To prevent bubble, do I need to increase a total volume? how much volume is recommended or suitable?

      Comment


      • #4
        TE and Qiagen buffer EB ( 10mm Tris HCl pH 8.5) generate the most reproducible results. Deionized and MiliQ water quality and pH vary from lab to lab, and it tends to affect fragmentation reporducibiltiy.
        We have also increased the processing volume in the 6x16mm tube to 130ul in order to prevent the formation of an airgap during processing which can widen the size distribution of fragments.
        What settings were you using, and what did the bioanalyzer trace look like?

        Thank you

        Hamid

        Comment


        • #5
          Covaris for fragmentation of DNA

          Does any one know how much the covaris is?

          Comment


          • #6
            Thank you all

            I found what problem was. Basically, my problem was that a graph of LCL samples in BioA appeared to begin from little bit upper of the bottom unlike blood samples. It was not a fault of TE buffer. This was due to remaining small proteins during shearing.

            But I found that shearing in water gave more constricted shape than in TE. I used Pfizer medical DW water rather than the other DW. It also worked very well in further steps.

            Also I found that 130 ul was the best volume (it gave narrower peak) but it was too big for a tube and little bit of samples bumped out of a tube, which may contaminate a water-tank. So I chose 120 ul or 125 ul. Also 100 ul volume, which is recommended by NimbleGen exmoe capture manual, showed a kind of birdnesting in higher concentration samples (300 ng/ul or 500ng/ul)

            Comment


            • #7
              Originally posted by Gillow View Post
              Does any one know how much the covaris is?
              Approx. $45000, for the S2.

              Comment

              Latest Articles

              Collapse

              • seqadmin
                Addressing Off-Target Effects in CRISPR Technologies
                by seqadmin






                The first FDA-approved CRISPR-based therapy marked the transition of therapeutic gene editing from a dream to reality1. CRISPR technologies have streamlined gene editing, and CRISPR screens have become an important approach for identifying genes involved in disease processes2. This technique introduces targeted mutations across numerous genes, enabling large-scale identification of gene functions, interactions, and pathways3. Identifying the full range...
                08-27-2024, 04:44 AM
              • seqadmin
                Selecting and Optimizing mRNA Library Preparations
                by seqadmin



                Sequencing mRNA provides a snapshot of cellular activity, allowing researchers to study the dynamics of cellular processes, compare gene expression across different tissue types, and gain insights into the mechanisms of complex diseases. “mRNA’s central role in the dogma of molecular biology makes it a logical and relevant focus for transcriptomic studies,” stated Sebastian Aguilar Pierlé, Ph.D., Application Development Lead at Inorevia. “One of the major hurdles for...
                08-07-2024, 12:11 PM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by seqadmin, 08-27-2024, 04:40 AM
              0 responses
              16 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 08-22-2024, 05:00 AM
              0 responses
              293 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 08-21-2024, 10:49 AM
              0 responses
              135 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 08-19-2024, 05:12 AM
              0 responses
              124 views
              0 likes
              Last Post seqadmin  
              Working...
              X