    Does TE buffer impact on covaris shearing?

    Dear all,

    I used bioanlayzer after covaris shearing. I put TE buffer in DNA samples, as protocol recommends, but result was not good. I just wonder whether salt in TE buffer impact on fragment size.

    Is there anyone who use water in DNA sample in covaris shearing instead TE buffer?
  • ECO
    --Site Admin--
    • Oct 2007
    • 1360

    #2
    I shear in 10mM TrisHCl 8.5, and the manufacturer recommended protocols perform pretty well.

    More often problems come from not enough volume in the micro tube (<120ul will give air bubbles and thus variability). What consumable are you using and what volume?

    Comment

    • sehrrot
      Member
      • Jul 2010
      • 58

      #3
      I did covaris shearing for NibleGen exomcapture library, which recommends a total volume of 100 ul with 1X TE buffer. I mixed app. 4 ul DNA and 96 ul TE buffer in microtube (6 x 16 mm, no. 520045).

      To prevent bubble, do I need to increase a total volume? how much volume is recommended or suitable?

      Comment

      • Hamid
        Senior Member
        • Sep 2009
        • 108

        #4
        TE and Qiagen buffer EB ( 10mm Tris HCl pH 8.5) generate the most reproducible results. Deionized and MiliQ water quality and pH vary from lab to lab, and it tends to affect fragmentation reporducibiltiy.
        We have also increased the processing volume in the 6x16mm tube to 130ul in order to prevent the formation of an airgap during processing which can widen the size distribution of fragments.
        What settings were you using, and what did the bioanalyzer trace look like?

        Thank you

        Hamid

        Comment

        • Gillow
          Junior Member
          • Mar 2010
          • 6

          #5
          Covaris for fragmentation of DNA

          Does any one know how much the covaris is?

          Comment

          • sehrrot
            Member
            • Jul 2010
            • 58

            #6
            Thank you all

            I found what problem was. Basically, my problem was that a graph of LCL samples in BioA appeared to begin from little bit upper of the bottom unlike blood samples. It was not a fault of TE buffer. This was due to remaining small proteins during shearing.

            But I found that shearing in water gave more constricted shape than in TE. I used Pfizer medical DW water rather than the other DW. It also worked very well in further steps.

            Also I found that 130 ul was the best volume (it gave narrower peak) but it was too big for a tube and little bit of samples bumped out of a tube, which may contaminate a water-tank. So I chose 120 ul or 125 ul. Also 100 ul volume, which is recommended by NimbleGen exmoe capture manual, showed a kind of birdnesting in higher concentration samples (300 ng/ul or 500ng/ul)

            Comment

            • shurjo
              Senior Member
              • Jan 2009
              • 132

              #7
              Originally posted by Gillow View Post
              Does any one know how much the covaris is?
              Approx. $45000, for the S2.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                by SEQadmin2



                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                ...
                07-31-2026, 11:01 AM
              • SEQadmin2
                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                by SEQadmin2


                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                The systematic characterization of the human proteome has
                ...
                07-20-2026, 11:48 AM
              • SEQadmin2
                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                by SEQadmin2



                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                ...
                07-09-2026, 11:10 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 10:13 AM
              0 responses
              14 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-31-2026, 02:55 AM
              0 responses
              29 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-24-2026, 12:17 PM
              0 responses
              22 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-23-2026, 11:41 AM
              0 responses
              21 views
              0 reactions
              Last Post SEQadmin2  
              Working...