Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • sehrrot
    Member
    • Jul 2010
    • 58

    #1

    LCL vs whole blood samples? (shearing differences)

    Dear all,
    I sheared lymphoblastoid cell line (LCL) samples and whole blood samples by covaris S2. Then, I ran bioanalyzer to check their length.
    Interestingly, unlike whole blood samples that showed typical graph after covaris shearing, sheared LCL samples looked different
    First, a bit wider than whole blood samples
    Second, low fluorescent unit -> I measured conc' again with nanodrop but same as whole blood.
    Third, their graph started from little bit higher than baseline.



    I assumed that fragmentation of episome is a reason why a bunch of small fragments are. But I cannot understand low concentration and wider distribution.
    Attached Files
  • upenn_ngs
    Member
    • Sep 2009
    • 70

    #2
    What is the downstream application? Looks like the LCL sample concentration was lower than expected, so although the same covaris profile was used the two samples run on the bioanalyzer much differently.

    Also, if you are sequencing be aware LCL samples have a high number of SNP artifacts from transformation.
    Last edited by upenn_ngs; 09-20-2010, 07:28 AM.

    Comment

    • sehrrot
      Member
      • Jul 2010
      • 58

      #3
      Originally posted by upenn_ngs View Post
      What is the downstream application? Looks like the LCL sample concentration was lower than expected, so although the same covaris profile was used the two samples run on the bioanalyzer much differently.

      Also, if you are sequencing be aware LCL samples have a high number of SNP artifacts from transformation.
      I used same protocols, including time, intensity, and volume, and application (covaris S2). Also, I ran bioa' with other samples in same condition.

      These LCL samples are transformed by EBV. I haven't found any papers that EBV change the host genome. If you know any papers for EBV integration, could you cite them on here?

      Comment

      • upenn_ngs
        Member
        • Sep 2009
        • 70

        #4
        Good question-- I presume the variability is due to a single cell, or subset of cells taking over the culture during transformation, and additional mutations during division. For whole exome resequencing, we have seen 10,000-30,000 more SNPs pass filter in LCL samples vs. blood.

        Comment

        • sehrrot
          Member
          • Jul 2010
          • 58

          #5
          Originally posted by upenn_ngs View Post
          Good question-- I presume the variability is due to a single cell, or subset of cells taking over the culture during transformation, and additional mutations during division. For whole exome resequencing, we have seen 10,000-30,000 more SNPs pass filter in LCL samples vs. blood.
          Yes, we used two LCL samples, which of one has been cultured for a long time. We are also expecting EBV-transformed one has some SNPs but I'm just wonder why it was shown broader. Maybe LCL-transformed cell's genome might have become fragile cause of virus integration (consistently it would take more CNV I think).

          Comment

          • Hamid
            Senior Member
            • Sep 2009
            • 108

            #6
            Hi Sehrrot,

            I think the diffrences you are noticing are simply due to concentration differences of the samples. I think it is essential not only to carry out a nanodrop analysis, but also a Qbit or similar picogreen analysis for concentration. My guess is that if you run a Qbit analysis you will notice that the concentration of the two samples will be different unlike the uv/vis analysis on the nanodrop.
            what settings, volume, buffer type, and tube type are you using for the fragmentation?

            Thank you

            Hamid

            Comment

            • sehrrot
              Member
              • Jul 2010
              • 58

              #7
              Hi Hamid,

              Thank you for your reply.
              I did QBit, nanodrop and Epoch and also did 5 different concentration in bioa' analysis.
              Even Qbit gave more accurate results, all three method showed almost same result.

              But my question is why LCL transformed sample showed little bit higher start region (10bp~80bp) and why LCL transformed samples generally showed lower concentration than whole blood sample.

              We used recommended covaris setting with various setting, including 100, 120 ul, 130 ul shearing and TE buffer, MiliQ water shearing.

              Comment

              • josdegraaf
                Member
                • Mar 2010
                • 33

                #8
                your lcl samples fragmented to much smaller sizes, and you lost a lott of those during cleanup?

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                  by SEQadmin2





                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
                  10-30-2026, 06:46 AM
                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Today, 07:59 AM
                0 responses
                13 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 11-05-2026, 09:25 AM
                0 responses
                32 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 11-03-2026, 11:54 AM
                0 responses
                45 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-28-2026, 09:57 AM
                0 responses
                169 views
                0 reactions
                Last Post SEQadmin2  
                Working...