  • salyamoff
    Junior Member
    • Sep 2013
    • 4

    SureSelect Wash 2

    Hi everyone! I had a failure with sureselect exome enrichment - no DNA in captured sample. I reread manual, then i read Faircloth (2013) "Target Enrichment of Illumina Libraries" and Holmberg (2005) "The biotin-streptavidin interaction can be reversibly broken using water at elevated temperatures" and i see that:
    1) Samples need be washed 30 minutes total at 65C. (Manual)
    2) Wash 2 buffer is 15 mM NaCl (Faircloth)
    3) On this conditions streptavidin-biotin interaction will be broken with 100% chances. (Holmberg)
    I don't think that SureSelect protocol is unreliable, but i think that my problem due to washing step.
    Last edited by salyamoff; 10-13-2016, 09:15 PM.
  • seqdavis
    Junior Member
    • Sep 2015
    • 5

    #2
    Hello salyamoff,

    We currently use Agilent's SureSelect XT and QXT protocols on a regular basis and have come across this a few times. Here are a few questions for your troubleshooting:

    1. What was your target yield (ng) going into hybridization? We have found and the literature will tell you that one needs at least 250ng of pre-capture library going into hybridization. If this is not the case, you will have to adjust your bait ratio.

    2. Did you perform the initial 95C denaturing step prior to the addition of your baits? If this doesn't occur, there are no single stranded molecules for the bait to anneal to therefore, no material will get 'pulled down.'

    3. During your washing at 65C, did you make sure the beads were resuspended prior to the 10 minute incubation (which is done 3 separate times)?

    4. Did you use water anywhere in the washing?

    Take a look at these sections of your experiment and see if anything stands out. From my experience, this is where the majority of issues arise with the SureSelect protocols.
    SeqDavis

    Comment

    • salyamoff
      Junior Member
      • Sep 2013
      • 4

      #3
      I repeated a enrichment and capture with another beads. Sequensing was done good. It looks like my dynabeads was depraved. But this beads succesfully captured biotenylated proteins recently. I think it because protein capture conditions aren't so extreme.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
        by SEQadmin2



        Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
        ...
        07-09-2026, 11:10 AM
      • SEQadmin2
        Cancer Drug Resistance: The Lingering Barrier to Rising Survival
        by SEQadmin2



        Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

        There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
        07-08-2026, 05:17 AM
      • GATTACAT
        Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
        by GATTACAT
        Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
        07-01-2026, 11:43 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 07-13-2026, 10:26 AM
      0 responses
      27 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-09-2026, 10:04 AM
      0 responses
      37 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-08-2026, 10:08 AM
      0 responses
      24 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-07-2026, 11:05 AM
      0 responses
      35 views
      0 reactions
      Last Post SEQadmin2  
      Working...