    PhiX spike-in issue

    Hi all,

    I have come across issues with the Illumina PhiX v3 in my project. I need to sequence my low diverse library on NextSeq 500. To balance out the diversity, I tried to spike-in 50% of the Illumina PhiX v3 control along with my library. But, the run wasn't good and I only managed to see a 9% alignment in my first run. At first, I thought it was resulted from the degradation of my PhiX. Thereby, I quantified my PhiX using picrogreen and the reading seems fine after conversion and the bioanalyzer showed no sign of degradation. I repeated my run the 2nd time. This time the Illumina FAS came over and supervised my library denaturation method to ensure I followed exactly the instructions recommended in the protocol, ie. library denaturation->library dilution to 1.8pM followed by PhiX denaturation->dilution of denatured PhiX to 1.8pm and, eventually, mixed both library and PhiX at 50:50 ratio. Everything was fine but the results of the 2nd run was still the same as in we couldn't achieved near 50% of PhiX spike-in. In addition, we observed a high error rate too. Does anyone have any suggestion on this issue? Thank you in advance for the guidance.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Three issues to consider:
    1- your library and PhiX
    2- sequencing chemistry
    3- sequencing hardware

    For correct diagnosis you need to provide some information from runs SAV metrics such as cluster density, PF%, % aligned to PhiX, Q30 scores

    Comment

    • edwardwong1070584
      Junior Member
      • Aug 2016
      • 5

      #3
      Hi @nucacidhunter,

      Thank you for your reply. Please see below for the metrics. Thank you in advance for your guidance.
      Attached Files

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        The issues that is apparent:

        1- Low cluster density that could be due to errors in quantification, denaturation or loading resulting in under loading or overloading (over clustering where several cluster are identified as a single cluster). Over clustering possibility can be checked by looking at images and also selecting intensity in Flow Cell Chart from analysis page which should show higher values (yellowish) in upper tiles in comparison to lower ones. In standard clustering lower tiles has higher density than the upper ones.

        2- PF% very low: indicates low diversity (as you have mentioned) that has not been improved by PhiX spike-in because PhiX% is not high enough (this can be checked by looking at %Base in Data By Cycle plot in analysis tab).

        3- High error rate: could be result of over clustering or poor sequencing primer compatibility (if custom primer was used)

        4- If you can try quantifying PhiX by qPCR

        I hope this and looking to similarities in second run helps you to identify the issue. If you need more help please post images from those analysis page plots.

        Comment

        • edwardwong1070584
          Junior Member
          • Aug 2016
          • 5

          #5
          Thank you very much for your reply, @nucacidhunter. The Illumina FAS is currently checking through the run metrics. He is also suspecting that the library sample itself might be overclustered, resulting in the template generation to perform poorly and also exlaining on why the alignment didn't make sense. We had actually run the 3rd time, in which we lowered the library sample loading concentration from 1.8 to 1.5pM and it did improve the cluster density. But, the error rate was still high.

          We did quantify our PhiX and the concentration was fine. So I guess we could omit the possibility of the degraded PhiX being used. I have attached the plot of data by cycle for this run for your reference.

          Thank you
          Attached Files

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by SEQadmin2


            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

            Here are nine questions we think about, in roughly the order they matter, before...
            06-18-2026, 07:11 AM
          • SEQadmin2
            From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
            by SEQadmin2


            Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


            The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
            ...
            06-02-2026, 10:05 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 06-17-2026, 06:09 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-09-2026, 11:58 AM
          0 responses
          100 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-05-2026, 10:09 AM
          0 responses
          121 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-04-2026, 08:59 AM
          0 responses
          113 views
          0 reactions
          Last Post SEQadmin2  
          Working...