Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • edwardwong1070584
    Junior Member
    • Aug 2016
    • 5

    #1

    PhiX spike-in issue

    Hi all,

    I have come across issues with the Illumina PhiX v3 in my project. I need to sequence my low diverse library on NextSeq 500. To balance out the diversity, I tried to spike-in 50% of the Illumina PhiX v3 control along with my library. But, the run wasn't good and I only managed to see a 9% alignment in my first run. At first, I thought it was resulted from the degradation of my PhiX. Thereby, I quantified my PhiX using picrogreen and the reading seems fine after conversion and the bioanalyzer showed no sign of degradation. I repeated my run the 2nd time. This time the Illumina FAS came over and supervised my library denaturation method to ensure I followed exactly the instructions recommended in the protocol, ie. library denaturation->library dilution to 1.8pM followed by PhiX denaturation->dilution of denatured PhiX to 1.8pm and, eventually, mixed both library and PhiX at 50:50 ratio. Everything was fine but the results of the 2nd run was still the same as in we couldn't achieved near 50% of PhiX spike-in. In addition, we observed a high error rate too. Does anyone have any suggestion on this issue? Thank you in advance for the guidance.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Three issues to consider:
    1- your library and PhiX
    2- sequencing chemistry
    3- sequencing hardware

    For correct diagnosis you need to provide some information from runs SAV metrics such as cluster density, PF%, % aligned to PhiX, Q30 scores

    Comment

    • edwardwong1070584
      Junior Member
      • Aug 2016
      • 5

      #3
      Hi @nucacidhunter,

      Thank you for your reply. Please see below for the metrics. Thank you in advance for your guidance.
      Attached Files

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        The issues that is apparent:

        1- Low cluster density that could be due to errors in quantification, denaturation or loading resulting in under loading or overloading (over clustering where several cluster are identified as a single cluster). Over clustering possibility can be checked by looking at images and also selecting intensity in Flow Cell Chart from analysis page which should show higher values (yellowish) in upper tiles in comparison to lower ones. In standard clustering lower tiles has higher density than the upper ones.

        2- PF% very low: indicates low diversity (as you have mentioned) that has not been improved by PhiX spike-in because PhiX% is not high enough (this can be checked by looking at %Base in Data By Cycle plot in analysis tab).

        3- High error rate: could be result of over clustering or poor sequencing primer compatibility (if custom primer was used)

        4- If you can try quantifying PhiX by qPCR

        I hope this and looking to similarities in second run helps you to identify the issue. If you need more help please post images from those analysis page plots.

        Comment

        • edwardwong1070584
          Junior Member
          • Aug 2016
          • 5

          #5
          Thank you very much for your reply, @nucacidhunter. The Illumina FAS is currently checking through the run metrics. He is also suspecting that the library sample itself might be overclustered, resulting in the template generation to perform poorly and also exlaining on why the alignment didn't make sense. We had actually run the 3rd time, in which we lowered the library sample loading concentration from 1.8 to 1.5pM and it did improve the cluster density. But, the error rate was still high.

          We did quantify our PhiX and the concentration was fine. So I guess we could omit the possibility of the degraded PhiX being used. I have attached the plot of data by cycle for this run for your reference.

          Thank you
          Attached Files

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          38 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          28 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          41 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          45 views
          0 reactions
          Last Post SEQadmin2  
          Working...