    Agilent Bioanalyzer Negative Peaks?!?

    Here is what I mean:

    Click image for larger version

Name:	NegativePeak.png
Views:	1
Size:	28.4 KB
ID:	309147

    This is some input DNA obtained by reverse-crosslinking sonicated chromatin submitted by someone interested in our doing library construction for them. It was run on an Agilent Bioanalyzer High Sensitivity Chip.

    Everything looks great except -- the peak produced has negative amplitude!
    The Agilent Bioanalyzer Trouble shooting manual suggests this could happen as a result of the sample containing "detergents" or "dyes".

    Anyone seen this and been able to figure out the cause?

    (Yes, I know I answered the post of someone asking nearly the same question 4 years ago. But having seen this a few time recently in our lab, I'm finding all the responses in that thread, including my own, unhelpful.)

    --
    Phillip
    Last edited by pmiguel; 10-19-2016, 10:39 AM.
  • luc
    Senior Member
    • Dec 2010
    • 469

    #2
    We have seen this before (slightly less dramatic) also with ChIP DNA from customers. Perhaps there is a reagent in the Diagenode ChIP kits that can have this effect? The library prep and sequencing went well nevertheless.

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.

      Comment

      • RickC7
        Member
        • Feb 2010
        • 31

        #4
        Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by RickC7 View Post
          Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.
          We re-ran the samples, loading 1/4th the amount. Same effect:
          Click image for larger version

Name:	NegativePeak2.png
Views:	1
Size:	27.2 KB
ID:	305173

          --
          Phillip

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by nucacidhunter View Post
            I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
            Interesting!

            Thanks,
            --
            Phillip

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by nucacidhunter View Post
              I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
              Alas, after checking with the people doing the sample prep, they don't use Tween. They use SDS. They also do a minelute clean-up afterwards.

              --
              Phillip

              Comment

              • seqdavis
                Junior Member
                • Sep 2015
                • 5

                #8
                Phillip,

                Have you checked this resource from Agilent - slide 12??



                Maybe the sample needs one more clean up??
                SeqDavis

                Comment

                • nucacidhunter
                  Jafar Jabbari
                  • Jan 2013
                  • 1250

                  #9
                  This suggests multiple causes including detergents and possibly chaotropic salts remains from column purification. Probably columns need to be washed more times than manufacturer recommends.

                  Comment

                  Latest Articles

                  Collapse

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, Today, 06:09 AM
                  0 responses
                  11 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 06-09-2026, 11:58 AM
                  0 responses
                  33 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 06-05-2026, 10:09 AM
                  0 responses
                  38 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 06-04-2026, 08:59 AM
                  0 responses
                  43 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...