    Here is what I mean:

    Click image for larger version

Name:	NegativePeak.png
Views:	1
Size:	28.4 KB
ID:	309147

    This is some input DNA obtained by reverse-crosslinking sonicated chromatin submitted by someone interested in our doing library construction for them. It was run on an Agilent Bioanalyzer High Sensitivity Chip.

    Everything looks great except -- the peak produced has negative amplitude!
    The Agilent Bioanalyzer Trouble shooting manual suggests this could happen as a result of the sample containing "detergents" or "dyes".

    Anyone seen this and been able to figure out the cause?

    (Yes, I know I answered the post of someone asking nearly the same question 4 years ago. But having seen this a few time recently in our lab, I'm finding all the responses in that thread, including my own, unhelpful.)

    --
    Phillip
    Last edited by pmiguel; 10-19-2016, 10:39 AM.

  • #2
    We have seen this before (slightly less dramatic) also with ChIP DNA from customers. Perhaps there is a reagent in the Diagenode ChIP kits that can have this effect? The library prep and sequencing went well nevertheless.

    Comment


    • #3
      I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.

      Comment


      • #4
        Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.

        Comment


        • #5
          Originally posted by RickC7 View Post
          Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.
          We re-ran the samples, loading 1/4th the amount. Same effect:
          Click image for larger version

Name:	NegativePeak2.png
Views:	1
Size:	27.2 KB
ID:	305173

          --
          Phillip

          Comment


          • #6
            Originally posted by nucacidhunter View Post
            I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
            Interesting!

            Thanks,
            --
            Phillip

            Comment


            • #7
              Originally posted by nucacidhunter View Post
              I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
              Alas, after checking with the people doing the sample prep, they don't use Tween. They use SDS. They also do a minelute clean-up afterwards.

              --
              Phillip

              Comment


              • #8
                Phillip,

                Have you checked this resource from Agilent - slide 12??



                Maybe the sample needs one more clean up??
                SeqDavis

                Comment


                • #9
                  This suggests multiple causes including detergents and possibly chaotropic salts remains from column purification. Probably columns need to be washed more times than manufacturer recommends.

                  Comment

                  Latest Articles

                  Collapse

                  • seqadmin
                    Essential Discoveries and Tools in Epitranscriptomics
                    by seqadmin




                    The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
                    04-22-2024, 07:01 AM
                  • seqadmin
                    Current Approaches to Protein Sequencing
                    by seqadmin


                    Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                    04-04-2024, 04:25 PM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by seqadmin, Yesterday, 11:49 AM
                  0 responses
                  13 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-24-2024, 08:47 AM
                  0 responses
                  16 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-11-2024, 12:08 PM
                  0 responses
                  61 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-10-2024, 10:19 PM
                  0 responses
                  60 views
                  0 likes
                  Last Post seqadmin  
                  Working...
                  X