    qubit and picogreen discrepancies

    We're a small lab that has done most of our quantitation using nanodrop and qubit, depending on the application. However, as we try to prepare multiplexed illumina libraries for dsDNA, we're trying to get a picogreen up and running. After some tech support help we were able to get a beautiful standard curve that seemed to be working well. But then we decided to double check our samples using the qubit, and found some very significant discrepancies. Some samples had similar readings on both machines, others would be tens of ng lower on the qubit, and yet others would be tens of nanograms higher on the qubit. So the concentration readings aren't consistently off in one direction or the other.

    We double checked the raw qubit standards and 2 sets of raw picogreen standards on the nanodrop, and all the readings were 92 ng/ul give or take a ng, so the standards seem to be consistent, and we're assuming the nanodrop is just underestimating them slightly. But when we tried qubiting the pico standard, it was about 86 ng/ul. When we tried replicating samples on the picogreen with the same standards, the values were sometimes significantly different.

    We're trying our best to ensure that all samples and standards are incubated to the same temperature for the same amount of time. We also do our best to minimize light exposure, even going so far as to pipette into the picogreen plates with the lights off and only a little ambient light coming in through closed shades

    Any thoughts or suggestions on what we might be doing wrong or things we could try would be appreciated. Thanks.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Some possible causes:

    1- variation in pipetting and pipette being out of calibration
    2- using low sample volume. For comparisons it is better to use minimum 3ul sample or the standard with two replicates
    3- using unsuitable Qubit reagent and buffer, best practice would be using sample with concentration around 10 ng/ul with HS reagents

    Comment

    • thermophile
      Senior Member
      • Apr 2015
      • 243

      #3
      Remember quibit is a single point calibration (with blank-s1 0ng, s2 10ng). So the further you are from 10ng/ul the worse the quantification will be. Are you doing replicates with your picogreen? (you can do 25ul diluted dye, 25ul diluted DNA in 384 well plates to minimize costs and allow you to do replicates)

      The nanodrop is going to be way off, don't worry about the numbers it comes up with.
      Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

      Comment

      • anjama
        Member
        • Mar 2016
        • 15

        #4
        Thanks for the input. Based on input here and experimenting in the lab, it looks like we'll be incorporating replication into our Picogreen protocol to ensure good consistency.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          Today, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Today, 05:39 AM
        0 responses
        10 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, Yesterday, 09:56 AM
        0 responses
        11 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-09-2026, 11:14 AM
        0 responses
        25 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 09-29-2026, 09:51 AM
        0 responses
        77 views
        0 reactions
        Last Post SEQadmin2  
        Working...