  • anjama
    Member
    • Mar 2016
    • 15

    qubit and picogreen discrepancies

    We're a small lab that has done most of our quantitation using nanodrop and qubit, depending on the application. However, as we try to prepare multiplexed illumina libraries for dsDNA, we're trying to get a picogreen up and running. After some tech support help we were able to get a beautiful standard curve that seemed to be working well. But then we decided to double check our samples using the qubit, and found some very significant discrepancies. Some samples had similar readings on both machines, others would be tens of ng lower on the qubit, and yet others would be tens of nanograms higher on the qubit. So the concentration readings aren't consistently off in one direction or the other.

    We double checked the raw qubit standards and 2 sets of raw picogreen standards on the nanodrop, and all the readings were 92 ng/ul give or take a ng, so the standards seem to be consistent, and we're assuming the nanodrop is just underestimating them slightly. But when we tried qubiting the pico standard, it was about 86 ng/ul. When we tried replicating samples on the picogreen with the same standards, the values were sometimes significantly different.

    We're trying our best to ensure that all samples and standards are incubated to the same temperature for the same amount of time. We also do our best to minimize light exposure, even going so far as to pipette into the picogreen plates with the lights off and only a little ambient light coming in through closed shades

    Any thoughts or suggestions on what we might be doing wrong or things we could try would be appreciated. Thanks.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Some possible causes:

    1- variation in pipetting and pipette being out of calibration
    2- using low sample volume. For comparisons it is better to use minimum 3ul sample or the standard with two replicates
    3- using unsuitable Qubit reagent and buffer, best practice would be using sample with concentration around 10 ng/ul with HS reagents

    Comment

    • thermophile
      Senior Member
      • Apr 2015
      • 243

      #3
      Remember quibit is a single point calibration (with blank-s1 0ng, s2 10ng). So the further you are from 10ng/ul the worse the quantification will be. Are you doing replicates with your picogreen? (you can do 25ul diluted dye, 25ul diluted DNA in 384 well plates to minimize costs and allow you to do replicates)

      The nanodrop is going to be way off, don't worry about the numbers it comes up with.
      Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

      Comment

      • anjama
        Member
        • Mar 2016
        • 15

        #4
        Thanks for the input. Based on input here and experimenting in the lab, it looks like we'll be incorporating replication into our Picogreen protocol to ensure good consistency.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM
        • SEQadmin2
          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
          by SEQadmin2


          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
          ...
          06-02-2026, 10:05 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        12 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-09-2026, 11:58 AM
        0 responses
        106 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-05-2026, 10:09 AM
        0 responses
        125 views
        0 reactions
        Last Post SEQadmin2  
        Working...