    NGS DNA Library Purification via Dynabeads

    Hello, I could use some help from the community.

    I have built 48 RAD-Seq libraries using a protocol from Tin et al. (2014) designed for use with museum specimens. I have confirmed that the libraries have been built correctly via vector cloning of a positive control. After amplification using Phusion HF DNA Polymerase (40 cycles), all samples have been run on a gel to confirm that bands are produced around ~200 bp.

    Then, I purified the PCR products via Dynabeads MyOne Carboxylic Aid as recommended by Tin et al. (2014). The protocol is as follows:

    1. 10 µl per sample are removed from the bottle. The storage buffer is removed using a magnetic stand. Then, the beads are washed twice in EB and resuspended in EB maintaining the same volume.
    2. 10 µl beads, 40 µl PCR product, and 100 µl of 14.5% PEG-6000 (0.9 M NaCl, 10 mM Tris, pH 6) are added.
    3. The solution is incubation at room temperature for 10 minutes. The supernatant is removed using a magnetic stand and discarded. The beads are washed twice with 70% ethanol and air dried at room temperature.
    4. Beads are eluted in 15 µl EB and incubated for 5 minutes at room temperature. The supernatant is kept.

    To quantify, I ran a SYBER I Green Assay on all samples and got values between 0.2 ng/µl to 2 ng/µl. Next, I ran some samples of varying quantity (but within the range of the chip’s capabilities) on a Bioanalyzer High Sensitivity Chip. Typically, I rarely get a distribution, and if I do, the concentration is very small. Other times, I am getting unusual distributions that do not seem to have a pattern.

    I have tried reamplifying for an additional 20 cycles and purifying again with no luck. I even replaced Phusion with GoTaq, which gave me unusual distributions. I am not sure what to make of this and believe something is wrong with the purification step. Does anyone have any experience with these beads or know of a solution?

    I have contacted the manufacturers of the beads (Life Technologies), and they suggest high volume of beads and longer incubation times. I have already tried the longer incubation times with a positive control without any luck.

    Thank you in advance.
    Attached Files
  • austinso
    Member
    • Jun 2012
    • 77

    #2
    Frankly, I think there is a typo in the protocol. It shouldn't be pH 6...it really should be pH 8.

    I'm not sure how you were even able to pH Tris to 6.0, since its pKa is 8.3...? i.e. it is no longer a buffer at that pH.

    Comment

    • cement_head
      Senior Member
      • Mar 2012
      • 265

      #3
      What would be the disadvantage of using a spin-column purification, or alternate bead based purification? (i.e. AMPureXP or ProNex)?

      Comment

      • cmbetts
        Senior Member
        • Jun 2012
        • 120

        #4
        That protocol looks like a scaled down version of the various homebrew AmpureXP recipes out there. It's probably a cost savings measure to avoid the premium on the real deal. I'd agree with the spin column or commercial beads recommendation. You otherwise may end up burning way more cash in troubleshooting than you'd save using the more expensive reagent up front. I'm probably too spoiled by corporate labs, but I find homebrew only saves money if you're going to be running enough reactions in the long run that the math heavily favors DIY.

        Comment

        • cement_head
          Senior Member
          • Mar 2012
          • 265

          #5
          Originally posted by cmbetts View Post
          That protocol looks like a scaled down version of the various homebrew AmpureXP recipes out there. It's probably a cost savings measure to avoid the premium on the real deal. I'd agree with the spin column or commercial beads recommendation. You otherwise may end up burning way more cash in troubleshooting than you'd save using the more expensive reagent up front. I'm probably too spoiled by corporate labs, but I find homebrew only saves money if you're going to be running enough reactions in the long run that the math heavily favors DIY.
          Yeah, we came to the same conclusion. We tried the new Promega ProNex beads and they worked very well, and they are very, very fast at separating. Thanks.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          60 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          42 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          75 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          74 views
          0 reactions
          Last Post SEQadmin2  
          Working...