Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • rabiecqa
    Junior Member
    • Feb 2011
    • 8

    #1

    RNA quality (bioanalyser) for agilent microarray

    hi,
    i have this profile of RNA quality to preceed for microarray , one strange peak is apearing every time.
    can you help me?
    the picture is there.
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    There are several possibilities. It probably is not a result of RNA degradation because you have clean large and small sub-unit rRNA peaks that appear in roughly the expected stoichiometry.

    Possibly the sample has been ribo-depleted prior to the Bioanalyzer run? A single cycle of most ribo-depletion methodologies will leave some rRNA behind. The result could look like the image you show above. Then the answer would be "that is your non-ribosomal RNA".

    Alternatively there may be some other RNA that is being transcribed to extremely high levels in your samples. These could include some highly induced gene, viral RNA, contamination from previous RNA preps done in the same centrifuge tube.

    Finally, you do not say what species this RNA is from. The large sub unit rRNA looks a little smaller than I would expect for a mammal. If this RNA is of mammalian origin, the sample may have been eukaryotic-rRNA depleted (eg, Invitrogen ribo-minus) but still retain some mitochondrial rRNA peaks. Or, alternatively, there may be some bacterial component to the total RNA. Again, a ribo-depletion method that was eukaryotic rRNA-specific would remove the eukaryotic rRNA, but perhaps not the bacterial rRNA.

    If you gave us a little more information about the sample (is it total RNA or has it been ribo-depleted? what species and tissue is it from? Etc.) the possibilities would be narrowed.

    --
    Phillip

    Comment

    • rabiecqa
      Junior Member
      • Feb 2011
      • 8

      #3
      hi,
      The samples were freezed in 10% DMSO and Fetal bovine serum .Then the total RNA was extracted from human Peripheral Blood Mononuclear cells(PBMC) by using "RNaseasy Qiagen mini Kit.The problem is majority of the samples have this additional peak.Please give me some suggesions.

      Comment

      • rabiecqa
        Junior Member
        • Feb 2011
        • 8

        #4
        some of additional results there:
        1:NK cells
        2: PBMCs cells
        Last edited by rabiecqa; 04-06-2011, 06:10 AM.

        Comment

        • TonyBrooks
          Senior Member
          • Jun 2009
          • 303

          #5
          How about possible DNA contamination? Have you DNase treated your samples?

          Comment

          • rabiecqa
            Junior Member
            • Feb 2011
            • 8

            #6
            i have not treated with DNAse. do you think i should treat it ?
            because the experiment was in sterile environment (PSM: post security microbiology Level 2)

            Comment

            • TonyBrooks
              Senior Member
              • Jun 2009
              • 303

              #7
              I would. You could always treat a few samples with DNase I, stop the reaction with EDTA, then re-run on the Bioanalyser to check if there's an improvement.

              Comment

              • rabiecqa
                Junior Member
                • Feb 2011
                • 8

                #8
                Yes I'll treat with DNAse I because there some informations contort this idea and seem to be the same problem of mine.
                link of information: http://www.ambion.com/techlib/tn/112/10.html

                Comment

                • rabiecqa
                  Junior Member
                  • Feb 2011
                  • 8

                  #9
                  hi, it's was a DNA contamination.. The use of DNAse I solve the problem and there is no strange peak..
                  thanks for all!

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Good call Tony!
                    Mysterious to me though. Genomic DNA I expect to be mainly >50kb -- which would likely not even show up on a nanochip. If something degraded the bacterial chromosome down to ~1 kb, I guess that is what you would see. Maybe there was some DNAse in earlier steps, but it did not result in a complete digestion?

                    --
                    Phillip

                    Comment

                    • Zapp
                      Junior Member
                      • Mar 2011
                      • 6

                      #11
                      I would rather assume that the RNeasy columns selectively bind low molecular DNA.

                      Zapp

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM
                      • SEQadmin2
                        New Genomics Technologies Take Aim at Long-Standing Limits
                        by SEQadmin2


                        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                        ...
                        09-28-2026, 10:25 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, Yesterday, 11:13 AM
                      0 responses
                      16 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, Yesterday, 10:48 AM
                      0 responses
                      18 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-15-2026, 10:52 AM
                      0 responses
                      35 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-14-2026, 05:39 AM
                      0 responses
                      42 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...