  • bruce01
    Senior Member
    • Mar 2011
    • 160

    Truseq RNA Prep Kit Questions

    Hi, hoping to get a few answers/see what others are getting from the Truseq RNA sample kit.

    I have just completed several trial runs, from my results using Bioanalyser to qualify success the libraries seem to be ~1/4 of what is shown in the handbook, thankfully they are ~260bp. So I am confident the protocol has worked, but am now wondering what to do to increase yield.

    Does anyone use any modifications to the standard procedure? I have been thinking:
    (i) increase in PCR enrichment cycles; anecdotally the PCR enrichment step does not suffer from PCR duplicates: is this really the case?
    (ii) use low-bind plates; I have used what we had in the lab but should probably find some low-bind plates: has anyone found this to make a difference in yield? Any recommendations for plates?
    (iii) is it because my RNA is not good quality? RINs ~8 which seems to be acceptable. Was going to try with some very high quality RNA but wanted to know others experience. Surely if it is fragmented during lib. prep. it will not matter so much about the RNA integrity if it isn't badly degraded(?)

    I will sequence and run qPCR but would like advice about issues of yield. First time doing libraries and the lab I am in has a DIY which has worked very well so not sure whether to go along with Truseq or just do the longer (but working) lab method.

    Thanks in advance,

    Bruce.
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Hi Bruce,
    (i) If you need more library to fill out your run -- getting that extra library by increasing enrichment cycles probably will result in additional PCR duplicates. Stands to reason...
    (ii) Depends on the binding capacity of normal plates. I don't know what that is. Maybe 1 ng?
    (iii) Yes. The TruSeq RNA kit entails a polyA RNA binding. If your RNA is degraded, then only that RNA adjacent to the polyA tail will be retained. The 5' ends of the RNA strands will be lost. That, among other things, will lower yield.

    --
    Phillip

    Comment

    • bruce01
      Senior Member
      • Mar 2011
      • 160

      #3
      Thanks for your reply Phillip,

      yes, it does stand to reason that PCR dupes are created with additional cycles, for our DIY method we have reduced to 7 or 8 cycles as this is a big problem. I have heard from technicians that dupes are reduced in the Illumina kit but have not seen any proof of this (yet!)

      I am going to try a sample with very high RIN and 28/16 ratio to determine if degradation plays a major part in my lower yield. I can see how degradation at 5' would affect yield, just hoping the method isn't very sensitive to this.

      Illumina techsupport have suggested a vacuum concentrator or preparing more library and pooling, but I am not too keen as we are cutting it fine with number of preps in the kit as it is.

      Bruce.

      Comment

      • bruce01
        Senior Member
        • Mar 2011
        • 160

        #4
        Hi Phillip,

        thanks for the reply. Was waiting to test a few more samples prior to replying myself.

        (i) Illumina techsupport suggest: making another library to add to the previous one to boost yield; use a vacuum concentrator (without heat). Agree with increasing PCR cycles, do not want to bias more than we already do.

        (ii+iii) constructed a very nice library (akin to the image in the handbook) using a very good quality RNA sample (RIN 9.5, 28/18 2.1). Also left XP steps for longer to increase binding which may have helped.

        There is a 'bump' after the ~260bp library, ~550-2000bp which I assume is insert/adapter concatenators(?)

        Cheers,

        Bruce.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
          by SEQadmin2


          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
          ...
          Yesterday, 10:05 AM
        • SEQadmin2
          Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
          by SEQadmin2


          With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


          Introduction

          Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
          05-22-2026, 06:42 AM
        • SEQadmin2
          Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
          by SEQadmin2

          Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


          Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
          05-06-2026, 09:04 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Yesterday, 12:03 PM
        0 responses
        19 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, Yesterday, 11:40 AM
        0 responses
        14 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 05-28-2026, 11:40 AM
        0 responses
        29 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 05-26-2026, 10:12 AM
        0 responses
        31 views
        0 reactions
        Last Post SEQadmin2  
        Working...