  • read length of GS FLX Titanium

    So far, the average read length of 454 GS FLX Titanium system in our lab is 400 base pair, with most less than 400 base pair. Lately, a run got around 320 base pair with less homopolymers in the sequenced genome. The library prep is ok. We don't know whether it is common for 454 GS FLX Titanium, or there is something else we can improve.

    How about your read length?
    Last edited by anabio; 12-11-2009, 12:56 AM.

  • #2
    With us running different projects from different PI with different library preps it is hard to break out any good statistics. But I would say that if you have been averaging 400 bp reads then you are doing good. Ours tend to be worse but then as I said -- multiple projects in any given run. Your run with 320 bp ... well ... I would be tempted to just chalk it up to a bad day unless subsequent runs are equally poor.

    As for our stats, here are the last 3 runs off of the machine. Take the numbers as you will. Some of the preps were fairly bad.

    Total Number of Reads Basecalled:,610840
    Average Sequence Length:,340.105
    Standard Deviation:,131.506

    Total Number of Reads Basecalled:,1170569
    Average Sequence Length:,385.581
    Standard Deviation:,122.721

    Total Number of Reads Basecalled:,461593
    Average Sequence Length:,290.199
    Standard Deviation:,161.673

    Comment


    • #3
      400~500bp is expected with 200 cycles. According to westerman's stats, the standard Deviation is considerable. I think there must be a QC limiting the read length to ensure high quality data. After cycles, the signal from certain wells were too low to be recorded because of nonsynchronous extension or weak signal from lost copies during wash. If this is the truth, how to improve?If not...Could anyone got longer read length give some suggestion?

      Comment


      • #4
        I do want to repeat that our runs tends to be mixed samples. Thus it becomes hard to troubleshoot problems since we are not running the same stuff over and over again. For example that last run had five samples in four regions (two samples were barcoded together). Looking at the per-region stats shows [length, std. dev]:

        region 1:,259.646,169.349
        region 2:,328.566,146.516
        region 3:,62.784,39.604
        region 4:,280.595,156.995

        The controls in all 4 regions were mostly the same with about 330 bases, s.d., 146 bases.

        So obviously the second sample was great (i.e., ran as good as the controls) while the third sample had prep problems.

        Our second to the last run we did much better both for the samples and for the overall run.

        region 1: 389.056,125.754
        region 2: 380.148,129.73

        With the control close to 445 bases, sd of 90.

        -------------

        Going back to anabio's original message, what library and controls stats do you normally get?

        Comment


        • #5
          Originally posted by westerman View Post
          I do want to repeat that our runs tends to be mixed samples. Thus it becomes hard to troubleshoot problems since we are not running the same stuff over and over again. For example that last run had five samples in four regions (two samples were barcoded together). Looking at the per-region stats shows [length, std. dev]:

          region 1:,259.646,169.349
          region 2:,328.566,146.516
          region 3:,62.784,39.604
          region 4:,280.595,156.995

          The controls in all 4 regions were mostly the same with about 330 bases, s.d., 146 bases.

          So obviously the second sample was great (i.e., ran as good as the controls) while the third sample had prep problems.

          Our second to the last run we did much better both for the samples and for the overall run.

          region 1: 389.056,125.754
          region 2: 380.148,129.73

          With the control close to 445 bases, sd of 90.

          -------------

          Going back to anabio's original message, what library and controls stats do you normally get?
          What Rick isn't telling you is that these samples are all cDNAs using the SMART kit. With which we frequently see a bimodal read length distribution with a tail extending towards 40 bases. I'm hoping all these issues go away with the Roche cDNA protocol.

          The Titanium controls comprise several template lengths, I believe. Might want to focus on the long ones only if you want to troubleshoot read length.

          Even if you do that, it is possible that the issue I'm going to call "polyA glare" of nearby sample beads could negatively impact the read length of the control beads.

          --
          Phillip

          Comment

          Latest Articles

          Collapse

          • seqadmin
            New Genomics Tools and Methods Shared at AGBT 2025
            by seqadmin


            This year’s Advances in Genome Biology and Technology (AGBT) General Meeting commemorated the 25th anniversary of the event at its original venue on Marco Island, Florida. While this year’s event didn’t include high-profile musical performances, the industry announcements and cutting-edge research still drew the attention of leading scientists.

            The Headliner
            The biggest announcement was Roche stepping back into the sequencing platform market. In the years since...
            03-03-2025, 01:39 PM
          • seqadmin
            Investigating the Gut Microbiome Through Diet and Spatial Biology
            by seqadmin




            The human gut contains trillions of microorganisms that impact digestion, immune functions, and overall health1. Despite major breakthroughs, we’re only beginning to understand the full extent of the microbiome’s influence on health and disease. Advances in next-generation sequencing and spatial biology have opened new windows into this complex environment, yet many questions remain. This article highlights two recent studies exploring how diet influences microbial...
            02-24-2025, 06:31 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, Today, 12:50 PM
          0 responses
          10 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 03-03-2025, 01:15 PM
          0 responses
          181 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 02-28-2025, 12:58 PM
          0 responses
          276 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 02-24-2025, 02:48 PM
          0 responses
          663 views
          0 likes
          Last Post seqadmin  
          Working...
          X