  • Tobias Kockmann
    Member
    • Sep 2011
    • 10

    #1

    no index reads

    Hello HiSeq2000 users,

    we just finished a multiplexed mRNA-seq run on the HiSeq2000 with puzzling results, maybe somebody encountered this phenomenon before:

    We prepared 6 mRNA-seq libraries using the illumina TruSeq kit incl. the indexed RNA adapters and sequenced using a 50x cycle single-end run on the HiSeq 2000 on a single lane. The insert cycles (1-50) yielded sequences of reasonable quality, although the cluster density on that lane was pretty high (around 1'000 clusters/mm^2). From cycles 51-58 (the index/barcode) we do not get any sequence information (NNNNNNNN), which means the do-convolution of the samples failed. On the same flow cell there was another multiplexed sample, using the same indexing system (illumina TruSeq) and this worked out pretty fine.

    Any idea, what went wrong here???
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Never had this issue with 6 indexes in a lane -- usually only a problem when there are less than 4 indexes for us.

    My speculation is that one or two of the six samples produced far more clusters than any of the the others. Under such circumstances I find the instrument frequently fails to obtain sequence of the indexes.

    I blame this on low diversity of sequence during scanning. But thus far I don't seem to be convincing many people -- especially at Illumina tech support.

    Are you sure that all 6 bases were unreadable? I just found out about a parameter:

    --use-bases-mask

    that can be used to "turn-off" failed bases in the bar code and force demultiplexing base only on the bases that work. Well, I may be making sound more sophisticated than it is. You have to determine which are the bad bases yourself (or with the help of tech support) and then feed the switch above a string that directs it not to use certain base numbers.

    --
    Phillip

    Comment

    • Tobias Kockmann
      Member
      • Sep 2011
      • 10

      #3
      Unfortunately all of the index cycles failed to deliver sequences, which means I can not use partial information to demultiplex the samples.

      Never the less, thx for the hint!!!

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        So only your sample failed to generate sequence for the index read cycles whereas another set of samples on the same flowcell worked right?

        Did all your reads end up in "unknown" directory after de-multiplexing?

        Originally posted by Tobias Kockmann View Post
        Unfortunately all of the index cycles failed to deliver sequences, which means I can not use partial information to demultiplex the samples.

        Never the less, thx for the hint!!!

        Comment

        • Tobias Kockmann
          Member
          • Sep 2011
          • 10

          #5
          Originally posted by GenoMax View Post
          So only your sample failed to generate sequence for the index read cycles whereas another set of samples on the same flowcell worked right?

          Did all your reads end up in "unknown" directory after de-multiplexing?
          Exactly, for my libraries we didn't get any sequence information from the index, but there was another multiplexed sample on the same flow cell that worked without any problem. Since, all the lanes on the HiSeq 2000 share the same reagents for sequencing, I consider this as a sort of positive control, which pinpoints to a problem which seems to be library specific.

          Comment

          • Tobias Kockmann
            Member
            • Sep 2011
            • 10

            #6
            ...and yes, all of my reads ended up in the "unknown" directory after demultiplexing! Without exceptions!

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              This sounds like a problem that is specific to your samples then since the other samples are working fine. Guess you could try running your samples again to see if this was a one time issue.

              Originally posted by Tobias Kockmann View Post
              Exactly, for my libraries we didn't get any sequence information from the index, but there was another multiplexed sample on the same flow cell that worked without any problem. Since, all the lanes on the HiSeq 2000 share the same reagents for sequencing, I consider this as a sort of positive control, which pinpoints to a problem which seems to be library specific.

              Comment

              • Tom Barker
                Member
                • Jun 2011
                • 16

                #8
                Hi,

                I have recently had a similar problem. The samples all clustered at ~600k/mm2 and the index reads were very good quality according to SAV however when the data was passed onto our bioinformatics team two of the lanes gave no sequence information for the index reads. I was just wondering if you ever got to the bottom of this.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  One thing to check is whether you get any demultiplexing using all 24 TruSeq v2 indexes. If, for instance, the wrong adapter got added to a library, everything would go into the unknown directory.

                  --
                  Phillip

                  Comment

                  • Jon_Keats
                    Senior Member
                    • Mar 2010
                    • 279

                    #10
                    Can you look at the on machine base composition stats for the index read? If so do you see the typical craziness or a flat line at zero for this lane in question? As suggested I'd check that you didn't submit the wrong indexes to the demultiplex step so maybe try submitting all 24 or double/triple check the library prep to ensure it is right.

                    Comment

                    • HESmith
                      Senior Member
                      • Oct 2009
                      • 512

                      #11
                      Using the wrong indices for demultiplexing would not produce Ns for the index read (see the first post). We've seen a similar phenomenon (good read one but no index bases called), but only at high (1000K) cluster densities. Any chance that the sample was loaded at too high a concentration (pmiguel is right that the reported densities are inaccurate in this situation)? The %PF would be lower than normal, and the thumbnails would show the actual cluster density.

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        Originally posted by HESmith View Post
                        Using the wrong indices for demultiplexing would not produce Ns for the index read (see the first post). We've seen a similar phenomenon (good read one but no index bases called), but only at high (1000K) cluster densities.
                        Actually, this is a concatenation of 2 threads. The first was started by Tobias Kockmann in September of last year. The second was started by Tom Barker a few days ago. Tom's results are qualitatively different than Tobias's. Tom has high quality values and a density of 600K/mm2. I am presuming that is his raw cluster density, though he does not specify.

                        --
                        Phillip

                        Comment

                        • Tom Barker
                          Member
                          • Jun 2011
                          • 16

                          #13
                          Hi,

                          I was originally told that the index reads only produced N's but have recently found out that they did produce sequence and as it turned out it was the wrong indexes. Could have saved me alot of time looking into this if he had of said the indexes don't match as oppose to they're all N's. Thanks for all your input anyway.

                          Tom

                          Comment

                          • pmiguel
                            Senior Member
                            • Aug 2008
                            • 2328

                            #14
                            I have never seen sequence reads generated during demultiplexing. Perhaps there is a switch to make that happen? Would be handy in cases like these.

                            Anyway, that is why if I see there was "no information" from the index reads, I interpret it as "none of the indexes I specified were found during demultiplexing".

                            --
                            Phillip

                            Comment

                            • cbird
                              Member
                              • May 2012
                              • 10

                              #15
                              possible explanation

                              I'm a novice, but I think that high cluster densities could lead to successful reads of your target DNA and unsuccessful reads of your indexes.

                              The Illumina sequencers use the first few sequencing cycles to identify clusters, and I believe this happens on the fwd read, the rev read (paired end), and the index read. There are only 6 unique signals in an index read (if six indices are used) versus 4^4 potentially unique signals in the target DNA reads. Consequently, high cluster density could lead to deconvolution problems. That could explain the pattern the original poster reported - that is, of course, if I understand what the sequencer is doing.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                Today, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              15 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...