  • der_eiskern
    Member
    • Jul 2009
    • 46

    #1

    Is the flowcell's layout known to affect perceived cluster density on a given tile?

    I just did my first sequencing run on the GAIIx. I was doing some quality control calculations from the data contained in the Summary File and I noticed something that I didnt' expect to see.

    i had very clear peaks (maxima) in the cluster density around Tile 1 and Tile 120 with a minima in the middle. In other words, the graph of Tile # (X-axis) versus Cluster Density (Y-axis) looks like a parabola.

    For lanes whose cluster density ranged from 150,000 to 220,000 the amount of mappable sequence was proportional (though not linearly) to the cluster density; while for the lanes that had between 230,000 and 250,000 clusters per tile, the trend seemed to be inversely proportional.

    As expected, Lane 8 was yielded <50% mappable reads of the ones that passed filter with a large amount not passing filters but it clearly had this parabolic distribution across the lane's tiles.

    My question is, does the parabolic cluster density reflect problems with the optics of detecting clusters, a problem with the flowcell, or a problem with the pump on the cluster generation station? or is this a normal pattern in your experience? Secondly, what order are the tiles on the flowcell? is it a 2x60 or a 1x120 or 3x40 design for each lane?

    Thanks,
    der.

    PS. i should clarify that these numbers are before QC. the numbers after QC are not parabolic when graphed.
    Last edited by der_eiskern; 10-20-2009, 06:23 PM. Reason: clarification
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    Originally posted by der_eiskern View Post
    Secondly, what order are the tiles on the flowcell? is it a 2x60 or a 1x120 or 3x40 design for each lane?
    The tiles for each lane are arranged in 2 columns of 60 tiles. The numbering starts with tile 1 at the bottom (nearest the fluid inlet) up to 60 at the top then shifts over to column 2 and continues 61-120 from top to bottom. Tiles 1 and 120 are next to each other. Your "parabola" is telling you that the clusters are more dense at the ends of the lane nearest the inlet holes.

    It is not surprising that the amount of useable data from lanes with >230,000 clusters per tile is proportionately lower. The recommended cluster density is ~200-210 thousand clusters per tile. Beyond that there is more signal overlap between tiles which lowers the quality of the base calls.

    Comment

    • james hadfield
      Moderator
      Cambridge, UK
      Community Forum
      • Feb 2008
      • 224

      #3
      I always supposed that we see cluster denisity gradients since the DNA we are hybridising is reducing in concentration as it flows down the flow cell and DNA 'sticks' to the oligos on the flow cell. This was always the basis of our smiley face showing that the flowcell was the right way round in the instrument! A sad face inicated that someone had made a mistake.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
        by SEQadmin2




        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
        Today, 06:46 AM
      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 10-28-2026, 09:57 AM
      0 responses
      22 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-28-2026, 09:51 AM
      0 responses
      14 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 11:13 AM
      0 responses
      68 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 10:48 AM
      0 responses
      48 views
      0 reactions
      Last Post SEQadmin2  
      Working...