    index contamination in multiplexed run

    Hi all,

    I have an issue with adapter/index contamination in a multiplexed lane:

    I multiplexed 8 samples in a single HiSeq run. Unfortunately >90% of my reads are associated with an adapter/index that I did not use. The majority of these reads are my organism (not a common one) so contamination from another user's sample is unlikely. I have not used this particular index in a year however and have sequenced samples since then without problem. I think this means that my library prep reagents are not contaminated. This leaves the adapters themselves, or... something else? Perhaps something about the indexing read???

    Thanks for any help!
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    Did you do the demultiplexing yourself or was it done using CASAVA? It may simply be a case of an incorrect samplesheet being used (easily fixable by a re-run of CASAVA).

    It is not clear from your post if all 8 samples are affected or just one.

    Comment

    • btb
      Junior Member
      • Jun 2011
      • 9

      #3
      Thanks for the response, GenoMax.

      I did not do the demultiplexing myself. Perhaps I can get the facility to re-run CASAVA. Though if it's an incorrect samplesheet it's still not clear this will fix my problem.

      >90% of the reads are associated with index #1 (an index I did not use). The remaining reads are split between my 8 samples. So in that regard, all 8 of my samples are affected.

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        So in this case the demultiplexing actually has worked, with a tag that you did not expect to be there.

        Did you make these libraries or did the facility make them? In case you did then this outcome does not make sense.

        Comment

        • btb
          Junior Member
          • Jun 2011
          • 9

          #5
          I guess the demultiplexing has worked. It's just given me an unexpected result.

          I made the libraries myself. And I agree it does not make sense! I'm struggling to figure out what happened. Like I said in the OP, index #1 is not one I used in this particular library prep (though I have used it in the past, nearly a year ago). I have had several successful Illumina runs in the meantime without this index #1 contamination. Most of the library kit reagents are new and therefore unlikely to be contaminated by index #1. So it kind of leaves me with contaminated adapters (some of which are new) as the likely source of the problem. But the contamination would have to be pretty severe to leave me with >90% reads associated with index#1. Perhaps several of the adapters I used contained index#1 by mistake? I quantified these samples by qPCR, so if they were loaded evenly by the facility then it's all the more confusing.

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            So was index 1 was determined to be present based on the reads that ended up the "undetermined" bin because it should not have been included in the normal samplesheet since you had not used it. What is stranger is that if it was someone else's sample that contaminated your pool those reads are aligning to your genome (unless the contaminating sample also happens to be from the same species).

            Comment

            • btb
              Junior Member
              • Jun 2011
              • 9

              #7
              You're correct. All of these index#1 reads are in the Undetermined bin.

              I'd be surprised if someone else was using this genome for NGS in our campus facility (I would think our lab would know as they would have likely acquired the organism from us). Contamination from someone else's sample would certainly explain my results but I'm thinking the real answer is elsewhere. I guess I could do some QC on the adapters I used. Though the company has agreed to send new adapters.

              Comment

              • Chipper
                Senior Member
                • Mar 2008
                • 323

                #8
                Check if your expected indexes are present in the fastq for index 1. I would guess someone made an error in the samplesheet, e.g. by using the wrong base 7 in the index reads.

                Comment

                • MU Core
                  Member
                  • Apr 2008
                  • 60

                  #9
                  I'm assuming that you provided a pool of your 8 libraries to the facility to sequence. It sounds as though your remaining 10% of total reads were evenly split between the expected 8 indexes. If you had a contamination issue with the index 1 adapter at the library stage, it would be highly unlikely you would be seeing >90% of the reads for this one index. I would more likely suspect a mix-up at the dilution step immediately prior to clustering. You can test this by remaking the pool and submitting again. The facility may be willing to test the new pool as a spike-in to the control lane. If the new pool shows no sign of index 1 and only your expected 8 indexes, you have your answer.

                  Comment

                  • btb
                    Junior Member
                    • Jun 2011
                    • 9

                    #10
                    Hi Chipper. The demultiplexing should have pulled all reads with my expected indices so I don't think they'll be in the Undetermined bin (the fastq that contains, among other things, reads with index#1).

                    Hi MU Core. I provided individual samples. The facility runs Bioanalyzer as QC, then dilutes and pools. They have the individual libraries so re-pooling and sequencing is a possibility, as is sequencing the individual libraries singly. I'm leaning towards prepping new libraries using new barcoded adapters and seeing if I can spike a pooled sample in a control lane.

                    Comment

                    • btb
                      Junior Member
                      • Jun 2011
                      • 9

                      #11
                      I wanted to follow up with some additional information. The data was re-run through CASAVA and the results were no different. The run statistics are attached in a pdf. Anything here stand out as a potential cause? The '% Perfect Index Reads' seem off. And the '% of raw clusters per lane' seem low as well, though this may just reflect the low read number #'s for most samples.

                      Would appreciate any thoughts. Still trying to make sense of all this.
                      Attached Files

                      Comment

                      • MU Core
                        Member
                        • Apr 2008
                        • 60

                        #12
                        The high % of one mismatch numbers suggest a run or chemistry issue. You definitely need to sit down with the folks generating the sequence and review the run metrics. There may have been some issues with the focus, etc. that could explain what appears to be poor base calling for at least the index read.

                        Comment

                        • GenoMax
                          Senior Member
                          • Feb 2008
                          • 7142

                          #13
                          Originally posted by btb View Post
                          Would appreciate any thoughts. Still trying to make sense of all this.
                          This is where you look through the "undetermined" pile to see what tags are represented there. If there are N's then not much you can do but there could be tags that you just did not expect (but are there anyway).

                          Comment

                          • parulagwl
                            Junior Member
                            • Jul 2014
                            • 6

                            #14
                            Hi btb,
                            Did you figure out the source of the unused index contamination in your libraries. I ran into exactly same issue and would really appreciate your input.
                            Last edited by parulagwl; 12-03-2014, 02:37 PM.

                            Comment

                            • GenoMax
                              Senior Member
                              • Feb 2008
                              • 7142

                              #15
                              Can you confirm if you have exactly the same situation btb had (i.e. majority of the reads have unexpected barcodes and thus end up in the "undetermined" bin)?

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                Today, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              14 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...