Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • DRYTCYV
    Member
    • Apr 2011
    • 50

    #1

    MiSeq fungal contamination

    Hello,
    We detected some kind of biofilm on the Miseq manual sipper and have reported to Illumina support. The reply from the supposed Illumina support was a little bit unsupported. They said that they don’t have any official recommendation but we can use a bleach dilution to soak it in (manual sipper). They informed that tween solution can lead to a fungal growth, and we need to be aware that now all the lines (all sippers) can be contaminated, and there’s no official procedure for this kind of problems.
    After reported several times some white spots in the images, looks like we discovered that it was fungus.
    Anyone with the same problem?
    Cheers,
  • thermophile
    Senior Member
    • Apr 2015
    • 243

    #2
    We had a film on the syringe plunger (didn't attempt to id it, it's either biofilm or chemical precipitation). I'm leaning towards precipitation because we do a bleach wash between each run. Illumina replaced the syringe and we stopped having runs fail due to bubbles.
    Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

    Comment

    • nangel
      Member
      • Nov 2014
      • 15

      #3
      Hi, we had noticed a similar fungal like growth on the side of the bottle containing the sipper after we had the machine in stand by for a few weeks. Although, I swapped all the solutions over before starting the sequencing run and had also done a bleach line template wash and 2 maintenance washes the next sequencing runs were very low in quality even with cluster densitys of around 800k/mm2 PF 94%. Illumina tech support had a look at the SAV files and have advised that fungal contamination is the issue here. Specifically the prephasing, high error rates, and the pattern of the Q scores dropping suggest this. Is this the same kind of run metrics you were seeing?

      Comment

      • DRYTCYV
        Member
        • Apr 2011
        • 50

        #4
        Hi nangel,

        Thank you for your reply

        Originally posted by nangel View Post
        Illumina tech support had a look at the SAV files and have advised that fungal contamination is the issue here. Specifically the prephasing, high error rates, and the pattern of the Q scores dropping suggest this. Is this the same kind of run metrics you were seeing?
        Now there´s some metrics that make sense, the prephasing increase, Q scores problems etc. I'm gona check the last runs to confirm this.

        Now the techsupport said to wash everything (all lines) with a bleach solution, but there´s no official protocol.

        Cheers,

        Comment

        • JBKri
          Member
          • Jan 2014
          • 80

          #5
          Originally posted by nangel View Post
          Hi, we had noticed a similar fungal like growth on the side of the bottle containing the sipper after we had the machine in stand by for a few weeks. Although, I swapped all the solutions over before starting the sequencing run and had also done a bleach line template wash and 2 maintenance washes the next sequencing runs were very low in quality even with cluster densitys of around 800k/mm2 PF 94%. Illumina tech support had a look at the SAV files and have advised that fungal contamination is the issue here. Specifically the prephasing, high error rates, and the pattern of the Q scores dropping suggest this. Is this the same kind of run metrics you were seeing?
          We too had some fungal growth in the wash bottle. We believe it may have been related to the water we used; the supposedly nanopure water turned out to have quite a lot of dirt inside. We switched to HPLC-grade bottled water; since then we haven't seen any fungus.

          Comment

          • nangel
            Member
            • Nov 2014
            • 15

            #6
            We are also doing the full line wash with bleach. My concern was that on going back to review all of the amplicon runs on the machine the pre-phasing and error rate were in exactly the same range from the very 2nd run on a brand new machine to the latest runs that were suspected to suffer from fungal contamination. I guess all I see is a rapid fall in the sequence quality as anything I can note as a difference. Still fingers crossed that fixes the problem. I guess the silver lining is that I now understand all of the run metrics much better and it should improve our run monitoring and loading (trying to keep that glass half full). I will also look at our water quality but we have changed SOP to autoclaving the MilliQ water in bottles and adding the Tween directly to this and keeping any stocks in foil in the fridge. With more rapid and recorded turnover, we are hoping this may also keep things ultra clean!

            Comment

            • ajeffs
              Junior Member
              • Jan 2012
              • 3

              #7
              We don't do the bleach wash (yet), but did switch to keeping the MiSeq wash buffers in the fridge a few months ago after noticing stuff growing in the buffer bottles.

              Comment

              • docphil
                Junior Member
                • Aug 2014
                • 8

                #8
                The quality of our MiSEQs runs has been consistently decreasing over the last 6 months, especially after the flip for paired end sequencing (read2). I am wondering what specific run metrics come with this fungal contamination and what kind of protocol are you using for a bleach wash for every line. Thanks!

                Comment

                • nangel
                  Member
                  • Nov 2014
                  • 15

                  #9
                  We didn't see a huge improvement after the bleach washes or using another machine that should not have been affected by fungal contamination. I think our issue is something different after all. As far as bleach wash for the machine, I would make sure that the protocol came from Illumina tech support so that you are covered if it caused any problems with your machine. In essence it is just using diluted bleach in every position in the wash tray for 3 washes and then at least 10 MilliQ washes immediatley after. It was a very long day of washing!

                  Have your runs slowly been decreasing in quality especially for R2? Ours was a very clear change from previous runs to all the runs going forward after a specific date. There are so many run metrics you can look at but perhaps the clearest was looking at the %perfect reads in R2.

                  I should also say that we are only currently running 2 x 300bp V3 kits so we are never looking at the most fantastic R2 sequencing.

                  Comment

                  • docphil
                    Junior Member
                    • Aug 2014
                    • 8

                    #10
                    Yes we have seen a dramatic decrease in R2 quality, about a 30-35% decrease in %>Q30 compared to runs from ~6months ago. We routinely use V2 500cycle kits for 2x251 reads. My PI has spoken to a large commercial sequencing company who also use MiSEQ for amplicon sequencing and they have noticed a sharp decline in run quality in recent months as well. Perhaps illumina has changed the kit chemistry?

                    Comment

                    • DRYTCYV
                      Member
                      • Apr 2011
                      • 50

                      #11
                      Hi everyone, sorry for the late reply and I need more time to check the previous Runs metrics.


                      I contacted tech support and they confirmed that the fungal contamination could be causing the white spots, that we saw in the images, and could potentially affect the metrics.

                      We started a new bleach protocol for all lines (1:100 dilution) and after that 2 water washes.

                      The last Run, was the first after the new wash protocol and looks like one of the worst runs in terms of quality. R2 Q30 ~57%. The cluster density was a bit high 1483 (V3 Kit), but i think this is not enough for a huge decrease in quality.

                      Originally posted by nangel View Post
                      As far as bleach wash for the machine, I would make sure that the protocol came from Illumina tech support so that you are covered if it caused any problems with your machine. In essence it is just using diluted bleach in every position in the wash tray for 3 washes and then at least 10 MilliQ washes immediatley after. It was a very long day of washing!
                      Maybe we need to increase the water washes...

                      Originally posted by nangel View Post
                      Have your runs slowly been decreasing in quality especially for R2? Ours was a very clear change from previous runs to all the runs going forward after a specific date. There are so many run metrics you can look at but perhaps the clearest was looking at the %perfect reads in R2.

                      I should also say that we are only currently running 2 x 300bp V3 kits so we are never looking at the most fantastic R2 sequencing.
                      Originally posted by docphil View Post
                      Yes we have seen a dramatic decrease in R2 quality, about a 30-35% decrease in %>Q30 compared to runs from ~6months ago. We routinely use V2 500cycle kits for 2x251 reads. My PI has spoken to a large commercial sequencing company who also use MiSEQ for amplicon sequencing and they have noticed a sharp decline in run quality in recent months as well. Perhaps illumina has changed the kit chemistry?
                      R2 seems a identified problem that Illumina dont like to talk. To avoid a reduced R2 quality we need to keep the clustering arround 1000-1200 (R2 Q30~75%).

                      Cheers,

                      Comment

                      • nangel
                        Member
                        • Nov 2014
                        • 15

                        #12
                        We only run our amplicons at 800 or below clustering with at least 10% PhiX. The recent decline in R2 quality has been confirmed by our FAS as a problem world wide with some batches of 2 x 250 bp and 2 x 300 bp MiSeq reagents. They are currently investigating. FAS have provided details of support and work arounds until the issue is resolved. Maybe check with your FAS to see if your reagents are similarly affected?

                        Comment

                        • dgovil
                          Junior Member
                          • Aug 2010
                          • 1

                          #13
                          DRYTCYV,

                          I concur with your statements. Our lab has been experiencing deteriorating Run quality on 2x250 and 2x300 kits.

                          the libraries that ran perfectly on older kits are failing and there are unofficial conformation regarding reagent quality from our FAS bit nothing official yet.

                          We have resorted yo running critical samples on the HiSeq instead.

                          Would love to hear other experiences as well

                          Comment

                          • JBKri
                            Member
                            • Jan 2014
                            • 80

                            #14
                            Originally posted by dgovil View Post
                            DRYTCYV,

                            I concur with your statements. Our lab has been experiencing deteriorating Run quality on 2x250 and 2x300 kits.

                            the libraries that ran perfectly on older kits are failing and there are unofficial conformation regarding reagent quality from our FAS bit nothing official yet.

                            We have resorted yo running critical samples on the HiSeq instead.

                            Would love to hear other experiences as well
                            We also seem to have a problem. We just finished a 2x300 run, and we got only 68 % > Q30. The signal intensities were lower than our previous v3 run. The cluster density also turned out lower than expected from the loading concentration.

                            Comment

                            • Vinz
                              Member
                              • Dec 2010
                              • 80

                              #15
                              Illumina currently has serious problems with MiSeq V2 and V3 reagents... Some lots perform terribly bad. Others are fine. It seems they are not communicating that well. I guess this is a problem particularly for smaller labs with lower throughput. Maybe it would be useful if we would come up with a shared spreadsheet where bad lots are identified?

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              75 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              49 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...