    Problem with small RNA library preparation

    Hi,

    I'm hoping someone can provide me with some answers regarding the small RNA library preparation.

    I have now used both the Illumina small RNA v1.5 and the Bioo Scientific AIR small RNA sequencing kit. However, I only ever end up with adapter dimer and no small RNA bands around 100 bp. I have cut the gel at 100 bp and run it on a DNA chip on the Agilent Bioanalyzer to confirm that there is nothing there. I also find that with the Bioo Scientific AIR kit, there is significantly less adapter dimer and the band tends to be tighter than with the Illumina kit.

    I am preparing samples from total RNA extracted from leukocytes using the trizol method and have tried concentrations ranging from 1 to 10 µg. Despite this, every time I prepare a library, I end up with an adapter dimer band (with the Illumina kit, this runs to almost 100 bp, the Bioo Scientific kit it is much tighter around 75-85 bp) and no small RNA library. At this point the obvious question is whether or not I have small RNA's in my samples. However, I am convinced that I do and have validated this on gels and using the Agilent RNA chip. I know the 3' adapter is supposed to be pre-adenylated to bind specifically with the small RNA fragment of my sample but how specific is this as it doesn't seem to happen in my sample?

    I'd appreciate any advice if there is anyone else who has had this issue.

    Many Thanks!
  • link1
    Member
    • May 2010
    • 21

    #2
    sounds like your sample might be lacking small RNA? I would recommend running your sample on a small agilent chip:
    http://www.chem.agilent.com/Library/...G_smallRNA.pdf

    make sure its the small chip you run. you might be losing the smalls in your trizol extraction. on a gel you won't be able to see the micrRNAs.
    I've made ~20 small libraries with both vendor's kits with success so far. In the beginning I was losing the smalls in my trizol extraction. Using the small chip helped me identify which trizol preps were good and which were bad.

    Comment

    • bpa400
      Junior Member
      • Aug 2010
      • 3

      #3
      Originally posted by link1 View Post
      sounds like your sample might be lacking small RNA? I would recommend running your sample on a small agilent chip:
      http://www.chem.agilent.com/Library/...G_smallRNA.pdf

      make sure its the small chip you run. you might be losing the smalls in your trizol extraction. on a gel you won't be able to see the micrRNAs.
      I've made ~20 small libraries with both vendor's kits with success so far. In the beginning I was losing the smalls in my trizol extraction. Using the small chip helped me identify which trizol preps were good and which were bad.
      I have run some of my samples on the Agilent small RNA chip now. The electropherogram shows a slight bump between ~10 - 30 nt and I'm getting micro RNA concentrations between 150-750 pg/ul. Is this a sufficient amount of micro RNA for the library preparation? Would reducing the adapter concentrations help to reduce adapter dimer while increasing the amplification efficiency of my small during PCR?

      Comment

      • pbluescript
        Senior Member
        • Nov 2009
        • 224

        #4
        If you are not already, make sure you add your ligase to the reaction last. I have seen that cause problems with adapter dimer before, especially with low input samples. Lowering the adapter concentrations would help as well.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          The Factors Reshaping Genome Assembly
          by SEQadmin2



          When the Human Genome Project declared the human genome finished in 2003, portions of the sequence were still missing, including many of its most repetitive and structurally complex regions.1 Still, for the next two decades, researchers analyzed their sequencing data against the single linear reference that came out of this effort. Released in 2013 as GRCh38, that reference has powered countless discoveries, but it has gaps and cannot capture the genetic diversity of the world’s...
          11-24-2026, 07:04 AM
        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2





          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
          10-30-2026, 06:46 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 11-23-2026, 11:11 AM
        0 responses
        39 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 11-17-2026, 06:27 AM
        0 responses
        47 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 11-12-2026, 07:59 AM
        0 responses
        68 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 11-05-2026, 09:25 AM
        0 responses
        68 views
        0 reactions
        Last Post SEQadmin2  
        Working...