    Total reads from MiSeq 16S V4 PE run

    Hi everyone

    Probably a stupid question, but i'm finding it hard to find meaningful info on sequencing depth and read numbers on the MiSeq.

    Illumina states that a 16S V4 run should deliver ~20m reads.

    Assuming 96 samples are run, that's 208,333 raw reads per sample.

    Therefore, if I run 24 samples, should I expect 833,333 raw reads per sample?

    I guess my question is why does the number of sequences fall with increased samples?

    Thank you
  • GW_OK
    Senior Member
    • Sep 2009
    • 411

    #2
    An Illumina sequencer will, or at least should, generate the same amount of reads for every run, respective to the chemistry loaded on it. A v3 Miseq kit should give about ~20M reads per run.

    Much like a pizza, every time you split it it the slices become smaller.

    Slicing a pizza (evenly) 24 times will give you larger pieces over slicing it 96 times.

    Comment

    • Egansbay
      Member
      • Jan 2015
      • 23

      #3
      Thank you for your reply GW_OK

      So from the number of ~20m reads, Illumina states that you should get ~100,000 seqs per sample, assuming 96 samples were run.

      Do you think that this number refers to 100,000 OTU classified seqs after quality filtering, given that the raw reads from 96 samples should be >200,000 per sample?

      Thank you again

      Comment

      • GW_OK
        Senior Member
        • Sep 2009
        • 411

        #4
        I don't know the context of Illumina's statement, so I can't comment.

        Comment

        • Egansbay
          Member
          • Jan 2015
          • 23

          #5
          I am just referring to the statement in the MiSeq wet lab protocol where it reads "The MiSeq run output is >20m reads, and assuming 96 indexed samples, can generate >100,000 reads per sample, commonly recognized as sufficient for metagenomic surveys".

          http://www.illumina.com/content/dam/...15044223-b.pdf

          Comment

          • GW_OK
            Senior Member
            • Sep 2009
            • 411

            #6
            Sounds about right, then.

            I wouldn't run 2x300's on the Miseq, though, if you're planning on doing that.

            Comment

            • Egansbay
              Member
              • Jan 2015
              • 23

              #7
              Thank you - I have run 2 x 300bp, but have to trim as the quality of the reverse read is not that good. Assume this is what you're referring to

              Comment

              • thermophile
                Senior Member
                • Apr 2015
                • 243

                #8
                20M is if you are loading the max which you really shouldn't for amplicons. I run v2 2x250, aiming for 700-800k clusters/mm, ~80% passing filter, so ~8-10M reads which includes the phiX (I'm running 15%). Also remember that your sequences will not be perfectly distributed between the samples.
                Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                Comment

                • Egansbay
                  Member
                  • Jan 2015
                  • 23

                  #9
                  Thanks Thermophile - I think the seq co must have loaded the max, as 24 samples yielded 22m raw reads.

                  Not sure if the seq was too deep though, as when I run rarefaction down to 100,000 reads/sample it still looks ok

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                    by SEQadmin2




                    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                    Today, 06:46 AM
                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-28-2026, 09:57 AM
                  0 responses
                  24 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-28-2026, 09:51 AM
                  0 responses
                  15 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 11:13 AM
                  0 responses
                  68 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 10:48 AM
                  0 responses
                  48 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...