  • Justin87
    Junior Member
    • Jan 2017
    • 6

    #1

    16S Sequencing Run - Under-Clustered; Higher PhiX Alignment than expected

    Good day,

    We recently performed a 16S metagenomics sequencing run (288 samples) on the Illumina MiSeq using a v3 (600 cycle) kit.

    We quantified the library using qPCR and calculated the size-adjusted concentration using the average fragment size from the Bioanalyzer.

    As we have typically done for previous runs, we spiked-in our PhiX at 15% at the same loading concentration as our pooled libraries (~5.5 pM).

    However, compared to previous runs, we achieved a much lower cluster density (~350 K/mm2) and data output (4.8 Gb) than normal (~700 K/mm2 and 7.5 GB, respectively). Also, the previous runs contained 384 samples.

    We are highly confident that our qPCR and Bioanalyzer results are accurate. Obviously, one would question the manual steps involved in denaturation and dilution of the libraries/PhiX. We also ensured the 0.2 N NaOH was made up fresh!

    The number of reads that aligned to PhiX was 23.94% rather than the expected 15%. I know that % of PhiX increases when the concentration of the library is low causing PhiX to take over.

    One concern I have is that the 20 pM PhiX library (prepared the week before and used for a sequencing run that worked) was stored as suggested in the Illumina guide: 'You can store the denatured 20 pM PhiX library up to 3 weeks at -15°C to -25°C'.

    I am wondering whether a more diluted form of PhiX was stored (rather than the 20 pM) after the previous run, and that perhaps the Phix was spiked into the present library at a lower concentration than the library. Would this influence the QC metrics of the run?

    Any suggestions/guidance would be greatly appreciated.

    Thank you!

    Kind regards,
    Justin
  • thermophile
    Senior Member
    • Apr 2015
    • 243

    #2
    Your phix isn't too low, it could be too high []. But if you used exactly that tube on a previous run, then I think it's fine and your library quant is off.
    Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

    Comment

    • Justin87
      Junior Member
      • Jan 2017
      • 6

      #3
      PhiX conc.

      Thank you, thermophile!

      Sorry for the ignorant question, but just out of interest, could I ask what would happen IF the phix was not spiked into the library at an equimolar concentration and was much lower than expected? Other than the % alignment, which other stats would be affected (cluster density, percentage of clusters passing filter, etc.)? Are any of these measurements based on PhiX conc.?

      Regards,
      Justin

      Comment

      • thermophile
        Senior Member
        • Apr 2015
        • 243

        #4
        % base and clusters PF are the biggest ones. A run of just 16S will likely fail because many of the first 20 cycles are only one base and the machine won't be able to focus for the other bases.

        One thing to check, how old is your 2N NaOH that you use to make fresh 0.2N? If it's over 6months you might want to throw it out and make fresh 2N from 10N stock
        Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          10-30-2026, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        42 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        25 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        74 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Working...