Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • ShyN
    Junior Member
    • Feb 2019
    • 7

    #1

    iSeq, High Q30% but Low PF%

    So there's not a lot of data for iSeq since its pretty new - but if anyone has had this issue on MiSeq or in general sequencing please help!

    We did an iSeq run today and the results had high Q30% (94%97%) but a VERY low PF% (19%)

    What could cause this kind of variation? I've found answers for the other way around but not this.

    We're considering the demultiplexing having gone wrong and we're working to correct that but if anyone else has experienced this error - what was the reason, and if it was demultiplexing how did you resolve or get a higher PF%? Is that something we can calculate as we can't get that data from the machine after our data work.
  • misterc
    Member
    • Jan 2016
    • 46

    #2
    Load up the run with SAV and see what the reported occupancy is. If it's below 90%, you've got too little library in there and need to increase the loading concentration. If it's 96% or higher, you've way over-loaded library and need to do the opposite.

    Comment

    • ShyN
      Junior Member
      • Feb 2019
      • 7

      #3
      For some reason, we can't see the reported occupancy on SAV for iSeq but we can for NovaSeq
      Last edited by ShyN; 02-06-2019, 03:03 PM. Reason: Wrong response

      Comment

      • Ahutch08
        Junior Member
        • Mar 2019
        • 1

        #4
        You need to load the run into BaseSpace to see SAV metrics for iSeq. Ability to view SAV metrics on instrument is coming in later software update.

        Did you determine whether your input library concentration was too low or too high?

        Comment

        • kbp
          Junior Member
          • Mar 2019
          • 2

          #5
          Used iSeq/MiSeq Cartridges

          Hi ShyN and/or others,

          Any chance you have a spent iSeq and/or MiSeq cartridge(s) that my colleagues and I could have or borrow, perhaps in exchange for a beer? We are in the Bay Area and could come by to pick up.

          Thanks!

          Comment

          • misterc
            Member
            • Jan 2016
            • 46

            #6
            Originally posted by Ahutch08 View Post
            You need to load the run into BaseSpace to see SAV metrics for iSeq. Ability to view SAV metrics on instrument is coming in later software update.

            Did you determine whether your input library concentration was too low or too high?
            SAV 2.4.7 (now released) now gives the %occupancy without having to deal with BS.

            Comment

            • ShyN
              Junior Member
              • Feb 2019
              • 7

              #7
              Originally posted by kbp View Post
              Hi ShyN and/or others,

              Any chance you have a spent iSeq and/or MiSeq cartridge(s) that my colleagues and I could have or borrow, perhaps in exchange for a beer? We are in the Bay Area and could come by to pick up.

              Thanks!
              Hi there - yes! We have a couple used cartridges that we give you. I've sent you an email via SeqAnswers.

              Comment

              • kbp
                Junior Member
                • Mar 2019
                • 2

                #8
                Used iSeq/MiSeq Cartridges

                Thanks Shy! Feel free to ping me at [email protected]. Thanks!!!

                Comment

                • ShyN
                  Junior Member
                  • Feb 2019
                  • 7

                  #9
                  Originally posted by Ahutch08 View Post
                  You need to load the run into BaseSpace to see SAV metrics for iSeq. Ability to view SAV metrics on instrument is coming in later software update.

                  Did you determine whether your input library concentration was too low or too high?

                  We're going to try to so BaseSpace now - we used a local application before but I think its time we give base space a chance.

                  We ended up concluding that our loading concentration was wayyy too low. So out PF was low because of this. Our %Q30 was high because out of the samples that passed (low as they were), most of them were useable.

                  Thanks everyone for your input!
                  Last edited by ShyN; 03-21-2019, 08:10 AM. Reason: Incorrect response

                  Comment

                  • ShyN
                    Junior Member
                    • Feb 2019
                    • 7

                    #10
                    Originally posted by misterc View Post
                    Load up the run with SAV and see what the reported occupancy is. If it's below 90%, you've got too little library in there and need to increase the loading concentration. If it's 96% or higher, you've way over-loaded library and need to do the opposite.
                    Hi, can you tell me where you found this data? Thanks!

                    Comment

                    • misterc
                      Member
                      • Jan 2016
                      • 46

                      #11
                      It's more empirically driven. We've been running NovaSeq (also patterned flow cell with this occupancy metric generated by Illumina) for a couple years now. iSeq performs very, very similarly.

                      Comment

                      • a.boese
                        Junior Member
                        • Mar 2019
                        • 1

                        #12
                        I recently started using the iSeq and I had the same problem. I felt that my library prep concentration was too high and I got these results by changing the concentrations (I was using PCR products fragmented with the NEBNext Ultra II FS library prep kit):

                        200pM % PF 39.26% AVG %Q30 82.64%
                        100pM % PF 67.54%; AVG %Q30 89.15%
                        50pM %PF 73.92%; AVG %Q30 91.81%

                        Comment

                        • ShyN
                          Junior Member
                          • Feb 2019
                          • 7

                          #13
                          Originally posted by a.boese View Post
                          I recently started using the iSeq and I had the same problem. I felt that my library prep concentration was too high and I got these results by changing the concentrations (I was using PCR products fragmented with the NEBNext Ultra II FS library prep kit):

                          200pM % PF 39.26% AVG %Q30 82.64%
                          100pM % PF 67.54%; AVG %Q30 89.15%
                          50pM %PF 73.92%; AVG %Q30 91.81%
                          We had some more issues with PF and they were corrected with the loading concentration being changed (raised or lowered depending on the % occupancy). We readily have 70% to 80% PF and the %Q30 have been in the 90th percentile.

                          The problem really ended up being miscalculation of our starting concentrations via qPCR due to an equation error.

                          50pM is the perfect loading concentration! Glad you (and I) were able to figure it out

                          Comment

                          • yoderto
                            Junior Member
                            • Feb 2019
                            • 2

                            #14
                            Hey ShyN-- We've had some yield issues we believe are related to loading concentration. I'm just curious, what plexities have you been running on your iSeq? Thanks!

                            Comment

                            Latest Articles

                            Collapse

                            • SEQadmin2
                              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                              by SEQadmin2




                              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                              10-30-2026, 06:46 AM
                            • SEQadmin2
                              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                              by SEQadmin2



                              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                              ...
                              10-14-2026, 09:55 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by SEQadmin2, 10-28-2026, 09:57 AM
                            0 responses
                            37 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-28-2026, 09:51 AM
                            0 responses
                            23 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-22-2026, 11:13 AM
                            0 responses
                            72 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-22-2026, 10:48 AM
                            0 responses
                            48 views
                            0 reactions
                            Last Post SEQadmin2  
                            Working...