    Improving Sequencing Results

    We are a core facility that receives samples from multiple labs. We have one lab in particular that submits which are ran on a 3730XL. We are trying to determine what is happening with these particular results. The first time that we run their plate the results don't look particularly good but if we put that same plate back into the ABI we see a much better signal. I've spoken with another facility who thought it was due to some contamination. I don't really know where this would be coming from as we run the bigdye reaction and cleanseq on all the plates and it's only seen with this particular lab's plates. I would really like to try to help troubleshoot this issue for them. I have attached images of the first and second run. Subsequent plates look pretty good on the first run but could also be improved on the second run. These are also being ran back to back.
    Attached Files
  • Markiyan
    Senior Member
    • Sep 2010
    • 126

    #2
    If the second run from the same plate is better, also the signal is declining with reads length, than it is salts being injected instead of labelled DNA fragments on the first run.
    I would suggest to use 200ul of 70% ethanol during the ethanol wash stage (if using ethanol precipitation in plates for sequencing reaction products purification). If doing precipitation in 1.5 ml tubes - than use 1ml of 70% ethanol on the wash step.

    Comment

    • Miranda
      Junior Member
      • Jan 2026
      • 4

      #3
      Originally posted by jblain_3 View Post
      We are a core facility that receives samples from multiple labs. We have one lab in particular that submits which are ran on a 3730XL. We are trying to determine what is happening with these particular results. The first time that we run their plate the results don't look particularly good but if we put that same plate back into the ABI we see a much better signal. I've spoken with another facility who thought it was due to some contamination. I don't really know where this would be coming from as we run the bigdye reaction and cleanseq on all the plates and it's only seen with this particular lab's plates. I would really like to try to help troubleshoot this issue for them. I have attached images of the first and second run. Subsequent plates look pretty good on the first run but could also be improved on the second run. These are also being ran back to back.
      Based on the fact that signal quality improves upon re-running the same plate, the issue is most consistent with sample-related inhibitory factors (e.g., high salt, EDTA, or residual ethanol) originating from the submitting laboratory rather than an instrument or reagent failure. The first run likely reduces these inhibitory components, resulting in improved signal quality in subsequent runs.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
        by SEQadmin2




        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
        Today, 06:46 AM
      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 10-28-2026, 09:57 AM
      0 responses
      22 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-28-2026, 09:51 AM
      0 responses
      14 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 11:13 AM
      0 responses
      68 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 10:48 AM
      0 responses
      48 views
      0 reactions
      Last Post SEQadmin2  
      Working...