    E120 reactions

    Hello All, is anyone running E120 reactions? I've tried 2 now and each time I only have 400M beads after enrichment instead of 2.2B.

    We are using the EZ bead system and following the aqueous phase set up as per the calculator provided by AB (with a total of around 4.4ng of template, ave size 290bp).

    What are the key factors in getting a successful reaction? if the instruments are set up correctly and the input material is correct what else can I tweak?

    JPC
  • RickC7
    Member
    • Feb 2010
    • 31

    #2
    E120

    Hi,

    I have done several successful E120s, but when I have a problem it is usually too many beads...

    How are you quantifying your library? If this is not accurate, then you will never get good beads.

    Are you using a scale to prepare your oil mix or using the 6 falcon tubes? I heard some labs were having trouble with scales, so that's the reason for the falcon tube method.

    Do you see any beads falling out of emulsion after amplification? Beads seen at bottom of pouch?

    Just a few things I can think of...

    Richard

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      You probably should try qPCR on the library. EZ Bead tends to give some beads, even with no library (unlike the manual method). Your library may not be good.

      --
      Phillip

      Comment

      • JPC
        Senior Member
        • May 2008
        • 116

        #4
        Thanks for the replies, we are quantifying with the Qubit, diluting our sample down to around 35pg/ul so that we are adding 125 ul of a 290bp product, I didn't want to dilute any further in case we drop below the qubit HS range.

        'Yes' we are prep'ing in batches of 6 falcon tubes and 'No' we don't see nay bead fall out at all.

        I'll give the qPCR a go to check the library, I have no reason to think there is a problem with it (we check regularly with the bioabnalyser, nanodrop and qubit) but it seems like a worthwhile exercise.

        thanks both

        JC

        Comment

        • yhey
          Junior Member
          • Oct 2010
          • 3

          #5
          Originally posted by JPC View Post
          Hello All, is anyone running E120 reactions? I've tried 2 now and each time I only have 400M beads after enrichment instead of 2.2B.

          We are using the EZ bead system and following the aqueous phase set up as per the calculator provided by AB (with a total of around 4.4ng of template, ave size 290bp).

          What are the key factors in getting a successful reaction? if the instruments are set up correctly and the input material is correct what else can I tweak?

          JPC
          I have heard recently from my FAS that one of the improvements to the enricher with the EZBead upgrade to v1.9, which has just been released, is to filter the E120 scale beads in several batches. Currently it filters in one batch and with the large scale emulsions there are just too many beads which puts excessive pressure on the filter causing it to be inefficient.
          If you feel your libraries are OK then maybe this is your problem and you are loosing 'good' beads.

          Yvonne

          Comment

          • JPC
            Senior Member
            • May 2008
            • 116

            #6
            We were told by AB to increase our titration point to 0.6picomoles this means our input library amount went from 3.62ng to 4.85ng for a 239 bp library. Unfortunately the Enricher (now running v1.9) shook the Hyb tube from it's place holder and the resulting Output tube solution is cloudy and not very brown....waiting to here from AB if I can collect the beads back together and save the emPCR somehow?!

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            61 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            43 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            76 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            74 views
            0 reactions
            Last Post SEQadmin2  
            Working...