  • lilletine
    Junior Member
    • Jan 2012
    • 7

    #1

    Guidelines - DNA input

    Hello - I am very new to this, our lab has just set up til 5500xl and we are about to start the first run. I am looking all over for general guidelines regarding library DNA input for the ePCR but i am unable to find documentation for this other than the 'ePCR calculator' which suggests 500 pM input DNA. We learn that it is really important to get the correct ratio of DNA and beads so i would like to see recommendations in writing. I am very grateful for any help regarding this!
  • JPC
    Senior Member
    • May 2008
    • 116

    #2
    Hi, the ePCR calculator is the only way to calculate your input DNA. For an E120 reaction we use a Titration point of 0.616 picomoles (which you get my adjusting the "Desired P2% post enrich" to 18) the enter your library size and concentration (ideally around 70-80 pg/ul). Ultimately for an E120 we end up loading around 5,000 - 6,000 pg of DNA depending on your product size.

    In terms of general advice, we normally use with 2ug of DNA to make a library, we found using the max (5ug) often caused problems, as did other users we know. Also after you have ligated your adaptors you come to an 'optional' amplification step.... we have found that this is NOT optional. No matter how much library you have at this stage I recommend doing the number of cycles listed in the table (page 26 of the Frag Lib prep User Guide).

    Good luck with your run

    JC
    Last edited by JPC; 02-08-2012, 04:18 AM.

    Comment

    • JPC
      Senior Member
      • May 2008
      • 116

      #3
      also, I'm not sure if it says this in the manuals but we tend to try and load 400M beads per lane

      JC

      Comment

      • lilletine
        Junior Member
        • Jan 2012
        • 7

        #4
        Thank you very much for your help!

        Comment

        • yhey
          Junior Member
          • Oct 2010
          • 3

          #5
          James, I am interested to know why you think the optional amplification is not optional. What problems have you encountered by not amplifying?

          Thanks
          Yvonne

          Comment

          • JPC
            Senior Member
            • May 2008
            • 116

            #6
            Hi Yvonne, skipping this step was the cause of the problems we experienced at the start (with the Enricher not returning the beads we were expecting). AB suggested that at the end of library prep there is still a significant proportion of fragments with no adaptor or just 1 adaptor so the 'optional amp' enriches for the fragments with 2 adaptors (P1-P2, P1-P1 and P2-P2).

            Comment

            • yhey
              Junior Member
              • Oct 2010
              • 3

              #7
              Originally posted by JPC View Post
              Hi Yvonne, skipping this step was the cause of the problems we experienced at the start (with the Enricher not returning the beads we were expecting). AB suggested that at the end of library prep there is still a significant proportion of fragments with no adaptor or just 1 adaptor so the 'optional amp' enriches for the fragments with 2 adaptors (P1-P2, P1-P1 and P2-P2).
              OK I thought that might be your reply. I have just had a similar experience with my first unamplified fragment libraries in that an E80 enrichment gave a bead yield less than expected from E20 but the sequence quality looks fine. I'll definitely amplify next time though!

              Thanks for the post James.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                by SEQadmin2



                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                ...
                07-31-2026, 11:01 AM
              • SEQadmin2
                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                by SEQadmin2


                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                The systematic characterization of the human proteome has
                ...
                07-20-2026, 11:48 AM
              • SEQadmin2
                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                by SEQadmin2



                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                ...
                07-09-2026, 11:10 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 08-03-2026, 10:13 AM
              0 responses
              15 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-31-2026, 02:55 AM
              0 responses
              32 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-24-2026, 12:17 PM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-23-2026, 11:41 AM
              0 responses
              21 views
              0 reactions
              Last Post SEQadmin2  
              Working...