  • peromhc
    Senior Member
    • Sep 2009
    • 108

    #1

    Library Quantification Confusion!

    I have just finished making some Illumina mRNA libraries using the v2 TruSeq Kit. I used the QUBIT to quantify samples, and got high concentrations (200-600ng/uL). Bioanalyzer traces suggest that there is a lot of material there, as well, and of the appropriate size.

    I contrast, I used the Kapa Bio library quantification kit, and got very different numbers concentrations, ranging from .01-90ng/uL. At the end of library prep, after enrichment PCR, the library should be enriched for dsDNA fragments that have adaptors on both ends, and thus are PCR competent. There should not be a lot of dsDNA floating around I don't think..

    So, why such big differences?

    Attached you'll see a bioanalyzer trace (1:10 dilution) of a library. QUBIT told me concentration was 321ng/uL, but Kapa qPCR says only .12ng/uL..

    Can anybody suggest a reason why this may be?
    Attached Files
  • niceday
    Member
    • Apr 2010
    • 68

    #2
    do you have quantitation for pre and postPCR enrichment?
    DEpending on the number of cycles you should see an increase in the amount of library.
    If the increase is low then the amount of properly ligated library is low before PCR or the PCR hasn't worked and you haven't amplified your correctly ligated material over your incorrectly ligated material.

    Comment

    • Eric@Kapa
      Member
      • Dec 2009
      • 11

      #3
      There are many possible explanations for this, but assuming there was nothing wrong with the qPCR assay, I would suggest that you proceed with caution -- the KAPA qPCR primers correspond to the flow cell oligo sequences, so anything that does not amplify during qPCR will not produce clusters on your flow cell.

      Without seeing the details of your qPCR assay, it's impossible for me to comment on potential problems. Please feel free to send full details to [email protected]:

      * details of how the assay was set up, which qPCR instrument was used, etc.
      * screen-captures of amplification profiles for standards and unknowns
      * screen-captures of standard curves
      * spreadsheet containing data including Ct scores of replicates, calculated concentrations, etc.

      Hopefully we can help to shed some light on your problem!

      Best regards,

      Eric

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        The simple explanation would be that you generated a large amount of cDNA and that relatively little of it was converted to library molecules. Presuming there was no issue with blunting/A-tailing your ds-cDNAs and ligation went as expected, then the limiting factor would be the amount of adapter available. In another thread the relative concentrations of adapter in the DNA vs RNA kits was determined to be 60:1.

        Once the adapter:insert molar conc. ratio drops below 2 you get into problematic territory. That is, if there is only 1 adapter available for each insert molecule, chances of getting a molecule with an adapter on each end drops to 25% (0.5^2).

        That said, the 10x difference in concentration between your Agilent chip (30 ng/ul -- taking into account the 1:10 dilution) and your QBIT is mysterious.

        --
        Phillip

        Comment

        • peromhc
          Senior Member
          • Sep 2009
          • 108

          #5
          perhaps more enlightening (see image)

          this is a different sample than the one I reference above, but the problem is the same. Note that between the top image (sample before PCR enrichment) and below (after PCR enrichment) that there is obvious peak enhancement.. Despite this, this the qPCR conc. value is very low.
          Attached Files

          Comment

          • HESmith
            Senior Member
            • Oct 2009
            • 512

            #6
            Did you include a positive control sample (e.g., phiX) in your qPCR? If that value is off, it would indicate a problem with your instrument or reagents.

            Comment

            • NextGenSeq
              Senior Member
              • Apr 2009
              • 482

              #7
              Note the small RNA adapters have a mismatch at the 3' end for the KAPA QPCR reatltime PCR primers. This gave us a huge difference in quantification of a small RNA library.

              Anyway, double check that the KAPA primers are correct for your library prep.

              Comment

              • pmiguel
                Senior Member
                • Aug 2008
                • 2328

                #8
                Also, check the calibration on your QBIT.

                --
                Phillip

                Comment

                • sehrrot
                  Member
                  • Jul 2010
                  • 58

                  #9
                  Yes recalibration for QuBit I would recommend. Even though it's done, QuBit gives not too accurate value and somtimes differs by a type of QuBit.

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    The OP is getting 100x higher numbers on the QBIT than on the BioAnalyzer. That suggests the a decimal place slip and/or the standards are 100x off.

                    --
                    Phillip

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM
                    • SEQadmin2
                      New Genomics Technologies Take Aim at Long-Standing Limits
                      by SEQadmin2


                      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                      ...
                      09-28-2026, 10:25 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    26 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    20 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-15-2026, 10:52 AM
                    0 responses
                    36 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-14-2026, 05:39 AM
                    0 responses
                    43 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...