    Hi Phillip,

    It was a pre-stained agarose gel. Thanks!

    Sonal

    Leave a comment:


  • pmiguel
    replied
    Originally posted by pbluescript View Post
    Sure. This one was 1.5% agarose post-stained with EtBr. I've actually tried it with 1-2% agarose post-stained with EtBr, Sybr-safe, and Sybr-gold and get identical results. I have also used the Invitrogen E-gels and see the same pattern.

    I generally take an aliquot of each library pre-PCR and titrate the number of cycles to decide how many each library needs.
    Then I am not sure what to think. I thought most of the double peak phenomenon was the result of the extra sieving resistance BioAnalyzer chip polymer had for ssDNA vs. dsDNA. But you are showing the same phenomenon in agarose -- agarose without run-time staining. Strange.

    --
    Phillip

    Leave a comment:


  • pbluescript
    replied
    Originally posted by pmiguel View Post
    Hi pbluescript,
    That is a great picture. Could we get info about the sieving matrix (agarose or acrylamide?) and whether you pre or post-stained it? And what the stain was (EtBr, SyBR-safe, SyBr-gold, etc)?

    --
    Phillip
    Sure. This one was 1.5% agarose post-stained with EtBr. I've actually tried it with 1-2% agarose post-stained with EtBr, Sybr-safe, and Sybr-gold and get identical results. I have also used the Invitrogen E-gels and see the same pattern.

    I generally take an aliquot of each library pre-PCR and titrate the number of cycles to decide how many each library needs.

    Leave a comment:


  • pmiguel
    replied
    Originally posted by pbluescript View Post
    You can see the gel I have attached to this post for an example of the appearance of these double sized bands as the number of PCR cycles goes from 6-18.
    Hi pbluescript,
    That is a great picture. Could we get info about the sieving matrix (agarose or acrylamide?) and whether you pre or post-stained it? And what the stain was (EtBr, SyBR-safe, SyBr-gold, etc)?

    --
    Phillip

    Leave a comment:


  • pmiguel
    replied
    Originally posted by ssing View Post
    Hi there,

    I got the exact same pattern you describe (couldn't see the image). See this post for what I found, what others suggested, and to see the end results. Let me know if you'd like more information.

    Techniques and protocol discussions on sample preparation, library generation, methods and ideas


    Best of luck!
    Hi ssing,
    Glad that worked for you. However, about the gel you ran after heating/cooling -- was that agarose? If so, did you pre-stain (eg, EtBr in loading dye and/or gel and/or running buffer) or post-stain the gel?

    --
    Phillip

    Leave a comment:


  • pbluescript
    replied
    It is due to too many cycles of PCR. When you have a high concentration of DNA fragments with complimentary ends, they can hybridize with each other, leading to the longer fragment sizes. Did you amplify each library with the same number of cycles or did you titrate the cycle number?
    If you titrate the number of PCR cycles, you can be sure you don't get these effects and introduce minimal PCR bias into your samples.
    You can see the gel I have attached to this post for an example of the appearance of these double sized bands as the number of PCR cycles goes from 6-18.

    Edit:
    The gel is 1.5% agarose post-stained with EtBr. The ladder is an Invitrogen 1kb+ ladder, so the sizes visible are 100,200,300,400,500,650.
    Attached Files
    Last edited by pbluescript; 03-05-2012, 04:59 AM. Reason: Added more information about the gel.

    Leave a comment:


  • MLog
    replied
    Originally posted by ssing View Post
    Hi there,

    I got the exact same pattern you describe (couldn't see the image). See this post for what I found, what others suggested, and to see the end results. Let me know if you'd like more information.

    Techniques and protocol discussions on sample preparation, library generation, methods and ideas


    Best of luck!
    Thanks, ssing! This is a very useful link. In my case the reason is not the beads - just because I don't use them. So I will certainly try heating and cooling.
    PS: there was indeed a problem with the attached image but now it should work.

    Leave a comment:


  • ssing
    replied
    Hi there,

    I got the exact same pattern you describe (couldn't see the image). See this post for what I found, what others suggested, and to see the end results. Let me know if you'd like more information.

    Techniques and protocol discussions on sample preparation, library generation, methods and ideas


    Best of luck!

    Leave a comment:


  • MLog
    started a topic RNA-seq libraries - double peak after size selection

    RNA-seq libraries - double peak after size selection

    Hi everyone,
    I get some weird results when measuring my RNA-seq libraries on Bioanalyzer and now desperately need an advice...
    The libraries look like a peak of expected length (300-400 or 400-500) and a higher peak (>1000 bp according to BA) (the image is attached).
    I'm using TruSeq RNA v.2 kit, following the protocol (except for using qiagen columns instead of ampure beads and a size-selection step after PCR enrichment).
    This is the first time that I'm making RNA-seq libraries but I have some experience with standard DNA libraries - and none of them looked like this. It was always a peak of expected length, sometimes with traces of adapter dimers at 120 bp - but never any peaks higher than expected.
    What puzzles me the most is that I got this AFTER size selection. I'm pretty sure that I cut the band about 300-500 bp and certainly not 1500-2000 bp! So this should be something generated from my 300-500 bp fragments. What could it be? A single-stranded DNA? If so, will it affect the sequencing resuts? Does it make sense to try to run these libraries or it would be safer to redo them?
    Thank you very much in advance!
    Attached Files
    Last edited by MLog; 02-08-2012, 02:46 PM.

Latest Articles

Collapse

  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM
  • SEQadmin2
    New Genomics Technologies Take Aim at Long-Standing Limits
    by SEQadmin2


    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
    ...
    09-28-2026, 10:25 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
37 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
28 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-15-2026, 10:52 AM
0 responses
41 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-14-2026, 05:39 AM
0 responses
45 views
0 reactions
Last Post SEQadmin2  
Working...