    Weird bioanalyser profile

    Dear all,
    I am trying to set up ChIP-seq in our lab. My problem is that, following ChIP, the bioanalyser profile of my input and samples is very different. The input shows plenty of material in the 2-300bp range. However, all my ChIP samples show a peak at around 90bp and go to baseline after that (see attached).
    Just to give you some info on the procedure I am using:
    Starting with 7 x 10^7 cells (Drosophila Kc cells)
    Fix with 1% formaldehyde for 10 minutes
    Sonicate with the Bioruptor for 7 pulses (30secs ON 30secs OFF)
    (I've optimised the number of pulses to make sure I get a smear of between 100bp and 1Kb when I look at an aliquot on a gel following reversal of crosslinking)
    I block samples with Rabbit IgG for 3-5 hours and treat with the antibody for the TF I am studying O/N.
    The beads are agarose G. I do the RNase, reversal of crosslinks and Proteinase K steps, and do Phenol-Chloroform extraction to purify using Glycoden and carrier tRNA.
    I have also done this exact procedure with the positive control H3K4me3. qPCR indicated an enrichment of a positive site by 40-50% so I assumed the ChIP had worked. However, the bioanalyser profile is the same as for my samples with a peak at 90bp and nothing bigger.
    If anyone has seen this kind of thing before and has any ideas I'd be really grateful.
    Thank you!
    Attached Files
  • Chipper
    Senior Member
    • Mar 2008
    • 323

    #2
    The spike and dip is present also in input, and is likely instrument related. I would not expect to see a mear on the dna1000 chip, we usually can not detect our sample on hs chips before library amplification.

    Comment

    • nisha barak
      Junior Member
      • Oct 2011
      • 8

      #3
      Thanks for your reply. I forgot to add: I also measured my samples on the Qubit and am getting very high (unusually for ChIP) concentrations, around 80 ng per microlitre (in a total of 30 microlitres). Attempts at library preparation (done by the sequencing centre) of these samples has failed as there is not enough material at around 200-300bp.

      Comment

      • lterhune
        Member
        • Sep 2011
        • 19

        #4
        Originally posted by nisha barak View Post
        Thanks for your reply. I forgot to add: I also measured my samples on the Qubit and am getting very high (unusually for ChIP) concentrations, around 80 ng per microlitre (in a total of 30 microlitres). Attempts at library preparation (done by the sequencing centre) of these samples has failed as there is not enough material at around 200-300bp.
        Hi Nisha, I also recently did a ChIP experiment that we have done previously, and got 52ng/ul in a total of 50ul on the Qubit. We haven't tried the library prep yet but are wondering if it is a waste of time and reagents. Let me know if you find anything else out!

        Comment

        • nisha barak
          Junior Member
          • Oct 2011
          • 8

          #5
          Hi Iterhune,
          I have discovered that the high amount of DNA indicated by the Qubit was actually the tRNA carrier I was using during phenol-chloroform extraction. When I ran my samples on a gel I just got a concentrated 90 bp band. On the other hand if I used spin columns (and therefore did not use a tRNA carrier) instead of phenol-chloroform extraction, this band disappeared. Now I know that my ChIP had too low a DNA amount to successfully make a library so I am instead trying to do several ChIPs which I will pool together.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by nisha barak View Post
            Hi Iterhune,
            I have discovered that the high amount of DNA indicated by the Qubit was actually the tRNA carrier I was using during phenol-chloroform extraction. When I ran my samples on a gel I just got a concentrated 90 bp band. On the other hand if I used spin columns (and therefore did not use a tRNA carrier) instead of phenol-chloroform extraction, this band disappeared. Now I know that my ChIP had too low a DNA amount to successfully make a library so I am instead trying to do several ChIPs which I will pool together.
            Can I just pause at this point to offer a blanket denunciation on the use of "tRNA carrier" on/with any sample that is to be sequenced? I have sequenced a fair amount of yeast in my day deriving from this source. As a result I am carrying a lot of unsublimated rage over this practice.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by SEQadmin2


              I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.


              Here are nine questions we think about, in roughly the order they matter, before...
              Today, 07:11 AM
            • SEQadmin2
              From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
              by SEQadmin2


              Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


              The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
              ...
              06-02-2026, 10:05 AM
            • SEQadmin2
              Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
              by SEQadmin2


              With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


              Introduction

              Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
              05-22-2026, 06:42 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Yesterday, 06:09 AM
            0 responses
            16 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-09-2026, 11:58 AM
            0 responses
            36 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-05-2026, 10:09 AM
            0 responses
            42 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-04-2026, 08:59 AM
            0 responses
            49 views
            0 reactions
            Last Post SEQadmin2  
            Working...