    454 Library prep for limited sample

    I have a few samples with less than 500ng (some as low as 200ng) of DNA that we would really like to prep for Titanium sequencing.
    Has anyone attempted this before, and if so, did you generate decent libraries that produced decent sequencing?

    Did you modify the protocol in any way?

    Things I'm thinking of;
    a) Covaris shearing as opposed to nebulisation
    b) Gel-based size selection (e-gel) instead of beads
    c) PCR amplification of final library

    Appreciate any pointers...
  • MrGuy
    Member
    • Mar 2009
    • 68

    #2
    Read these:
    We describe a protocol for construction and quantification of libraries for emulsion PCR (emPCR)-based sequencing platforms such as Roche 454 or Ion Torrent PGM. The protocol involves library construction using customized Y adapters, quantification using TaqMan-MGB (minor groove binder) probe-based …

    Continuous efforts have been made to improve next-generation sequencing methods for increased robustness and for applications on low amounts of starting material. We applied double-stranded library protocols for the Roche 454 platform to avoid the yield-reducing steps associated with single-stranded …

    Comment

    • Zaag
      Senior Member
      • Nov 2009
      • 112

      #3
      I made good libraries with 100 ng starting DNA.

      1) Yes, use the Covaris.
      2) No, use beads.
      3) Only necessary if the final concentration is too low to measure or run on a bio-analyzer.

      Comment

      • TonyBrooks
        Senior Member
        • Jun 2009
        • 303

        #4
        Originally posted by Zaag View Post
        I made good libraries with 100 ng starting DNA.

        1) Yes, use the Covaris.
        2) No, use beads.
        3) Only necessary if the final concentration is too low to measure or run on a bio-analyzer.
        Hi Zaag
        Can you let me know your Covaris settings? Our libraries always seem to be on the large size (800-1000bp). We've tried the suggested settings for both 500bp and 400bp shearing.

        Comment

        • Zaag
          Senior Member
          • Nov 2009
          • 112

          #5
          we use the covaris S2:

          45 seconds
          Intensity 3
          Dutycycle 5%
          Cycleburst 200

          use the tubes from covaris, not eppendorfs or such
          fill up your sample with TE or EB of H2O or whatever till the tube is full (130 ul)

          we routinely have gDNA from blood isolated by a gentra and it works fine for that, however we do see that if DNA is isolated manually the time can differ; settings remain the same;

          hope this helps

          Comment

          • Zaag
            Senior Member
            • Nov 2009
            • 112

            #6
            by the way, as for point 3, maybe performing a few cycles of PCR is not a bad thing because you measure fragments that you can not use for emPCR: the libraries with only 1 adapter

            Comment

            • TonyBrooks
              Senior Member
              • Jun 2009
              • 303

              #7
              I used the same settings but 90 second shearing and 120µL on our S2 with the AFA snap cap tubes. Increased time should generate smaller fragments, surely?
              I wonder why we're getting large libraries.

              We're going to quantify our libraries by qPCR, so no need for enrichment PCR. I never liked the Roche 454 fluorescence assay because, as you say, it doesn't account for single adapter ligation.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                by SEQadmin2


                I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                Here are nine questions we think about, in roughly the order they matter, before...
                06-18-2026, 07:11 AM
              • SEQadmin2
                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                by SEQadmin2


                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                ...
                06-02-2026, 10:05 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 06-26-2026, 11:10 AM
              0 responses
              12 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-17-2026, 06:09 AM
              0 responses
              46 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-09-2026, 11:58 AM
              0 responses
              106 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-05-2026, 10:09 AM
              0 responses
              125 views
              0 reactions
              Last Post SEQadmin2  
              Working...