  • TonyBrooks
    Senior Member
    • Jun 2009
    • 303

    #1

    454 Library prep for limited sample

    I have a few samples with less than 500ng (some as low as 200ng) of DNA that we would really like to prep for Titanium sequencing.
    Has anyone attempted this before, and if so, did you generate decent libraries that produced decent sequencing?

    Did you modify the protocol in any way?

    Things I'm thinking of;
    a) Covaris shearing as opposed to nebulisation
    b) Gel-based size selection (e-gel) instead of beads
    c) PCR amplification of final library

    Appreciate any pointers...
  • MrGuy
    Member
    • Mar 2009
    • 68

    #2
    Read these:
    We describe a protocol for construction and quantification of libraries for emulsion PCR (emPCR)-based sequencing platforms such as Roche 454 or Ion Torrent PGM. The protocol involves library construction using customized Y adapters, quantification using TaqMan-MGB (minor groove binder) probe-based …

    Comment

    • Zaag
      Senior Member
      • Nov 2009
      • 112

      #3
      I made good libraries with 100 ng starting DNA.

      1) Yes, use the Covaris.
      2) No, use beads.
      3) Only necessary if the final concentration is too low to measure or run on a bio-analyzer.

      Comment

      • TonyBrooks
        Senior Member
        • Jun 2009
        • 303

        #4
        Originally posted by Zaag View Post
        I made good libraries with 100 ng starting DNA.

        1) Yes, use the Covaris.
        2) No, use beads.
        3) Only necessary if the final concentration is too low to measure or run on a bio-analyzer.
        Hi Zaag
        Can you let me know your Covaris settings? Our libraries always seem to be on the large size (800-1000bp). We've tried the suggested settings for both 500bp and 400bp shearing.

        Comment

        • Zaag
          Senior Member
          • Nov 2009
          • 112

          #5
          we use the covaris S2:

          45 seconds
          Intensity 3
          Dutycycle 5%
          Cycleburst 200

          use the tubes from covaris, not eppendorfs or such
          fill up your sample with TE or EB of H2O or whatever till the tube is full (130 ul)

          we routinely have gDNA from blood isolated by a gentra and it works fine for that, however we do see that if DNA is isolated manually the time can differ; settings remain the same;

          hope this helps

          Comment

          • Zaag
            Senior Member
            • Nov 2009
            • 112

            #6
            by the way, as for point 3, maybe performing a few cycles of PCR is not a bad thing because you measure fragments that you can not use for emPCR: the libraries with only 1 adapter

            Comment

            • TonyBrooks
              Senior Member
              • Jun 2009
              • 303

              #7
              I used the same settings but 90 second shearing and 120µL on our S2 with the AFA snap cap tubes. Increased time should generate smaller fragments, surely?
              I wonder why we're getting large libraries.

              We're going to quantify our libraries by qPCR, so no need for enrichment PCR. I never liked the Roche 454 fluorescence assay because, as you say, it doesn't account for single adapter ligation.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                by SEQadmin2



                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                ...
                07-31-2026, 11:01 AM
              • SEQadmin2
                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                by SEQadmin2


                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                The systematic characterization of the human proteome has
                ...
                07-20-2026, 11:48 AM
              • SEQadmin2
                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                by SEQadmin2



                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                ...
                07-09-2026, 11:10 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 07:41 AM
              0 responses
              12 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 08-03-2026, 10:13 AM
              0 responses
              25 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-31-2026, 02:55 AM
              0 responses
              38 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-24-2026, 12:17 PM
              0 responses
              25 views
              0 reactions
              Last Post SEQadmin2  
              Working...