    Concentrating NexteraXT libraries

    Just wondering if anyone has any experience with this. I am still fairly new to preparing NexteraXT libraries. I have had good luck with the bead based normalization and running pools on the MiSeq, but now I want to send some pools to a HiSeq core lab.

    My HiSeq core wants pools to be around 10nM. Illumina recommends pooling the double stranded libraries from the CAN plate and then using that for the HiSeq (vs the single stranded libraries coming out of the bead based normalization). I used a Qubit to quantitate one of my first CAN plates and the concentrations are fairly low, in the 500-3000 pg/uL range. The average size is around 800bp on Bioanalyzer. I can pool them but the pool volume is going to be large and the concentration very low, so I need to concentrate the pools.

    Illumina recommends SpeedVac of the samples without heat, but tech support warned me if I used the buffer in the kit for the elution into the CAN plate (and why wouldn't have I?), then the salts in the buffer will concentrate as well, so they recommend not reducing the volume by >50%. They say in the future I can elute with H20 instead of the buffer in the kit, but that doesn't help me now. Any other column methods, while possible, would run the risk of sample loss. Ethanol precipitation might be an option, but the quantities are so low, I fear being able to see a pellet or achieve success with that method.

    I haven't sat down to do the math yet to figure out my theoretical volumes, but just wondering if anyone else had any experience or advice for running NexteraXT libraries on the HiSeq.

    Best,
    DB
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    AMPure.

    --
    Phillip

    Comment

    • docbio
      Member
      • Jul 2012
      • 26

      #3
      I have AmpureXP beads in house but they are a couple of months out of date - not sure how quickly they degrade but not sure I want to risk it. I have fresh AmpureXP on order.

      If I were to go the AmpureXP route, would you follow the standard PCR cleanup protocol from Beckman with a 1.8x bead ratio, or something more specific to concentrating libraries? My library size ranges from around 200bp to 1kb with a peak around 800bp. I've been looking around but can't find a consensus.

      Thanks for the help.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        0.8x ampure works in our hands to get rid of much of the stuff below 300. Of course this may be somewhat batch-dependent.

        Keep your first supernatant as a fail-safe. If you need to, you can add additional ampure to recover all the DNA.

        --
        Phillip

        Comment

        • docbio
          Member
          • Jul 2012
          • 26

          #5
          My pool volumes wound up being rather large (~500uL) and I didn't have any un-expired AmpureXP on hand, so I decided to call an audible and try an Amicon 0.5ml Ultracel-30K concentrator that we had on hand. It seems to have worked fairly well with around an 80% recovery of my calculated input (too dilute to detect by Qubit). Looks good on the Bioanalyzer, too, although the size range does tend towards the large end in one pool in particular.

          I have heard that the quantitation on the Bioanalyzer is not reliable, compared to the Qubit. In looking at a few individual libraries that I ran on the higher end of the concentration spectrum as well as my pools, it seems the Bioanalyzer calculated concentration is consistently around one third to one half of the Qubit value. Is this consistent with what others see?

          Best,
          docbio

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          43 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          33 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          44 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          45 views
          0 reactions
          Last Post SEQadmin2  