    I have 2 small questions concerning the 16s sequencing protocol for MiSEq:

    1) Some samples have very low concentration so we are considering loading a 2pM library instead of the more standard 4pM. Anyone having bad (low) clustering doing this or no problem when using the v3 MisEq kit (2 X 300 bp) ?

    2) There are various older threads discussing the PHIX % to be added to the library. The protocol says 5 % (v3 kits again), but the rep says 10-20 %. Any comments on this ?

    thanks

    Greg
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Anyone having bad (low) clustering doing this or no problem when using the v3 MisEq kit (2 X 300 bp) ?
    We saw inconsistent cluster density. Illumina recommends concentration by SpeedyVac. Other options would be concentrating them using column or AMPure beads by reducing elution volume.

    The protocol says 5 % (v3 kits again), but the rep says 10-20 %. Any comments on this ?
    Illumina's current official recommendation with the latest software is 5% PhiX spike-in, but 10% spike-in increases the likelihood of successful run.

    Comment

    • BioGenomics
      Member
      • Apr 2009
      • 25

      #3
      Thanks for this nucacidhunter. So is 4 pM the optimal (clustering) conc for amplicon Seq ?

      cheers

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        Illumina’s guide for clustering libraries on MiSeq can be find by google searching “Preparing Libraries for Sequencing on the MiSeq”. 4 pM is the concentration for denaturation step. Final pM concentration for clustering varies among labs as the quantification systems and practices differ. For amplicons less input improves PF% and also quality scores which usually is achieved with cluster densities below 1k. Illumina also recommends incubation of HT1 diluted denatured libraries at 96°C for 2 min before loading to improve clustering GC rich fragments

        Comment

        • BioGenomics
          Member
          • Apr 2009
          • 25

          #5
          I had another question concerning low cleaned amplicon concentration before indexing PCR. Can the Nextera indexing PCR cycle (8 cycles) be increased (to 10-12) without any issues to increase the concentration before final bead clean-up ? Also, what's the contamination likelihood using a SpeedVac ?

          cheers
          Last edited by BioGenomics; 07-30-2014, 12:50 AM.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            If concentration of amplicon PCR after clean-up is 200 pg/ul, using 5 ul in index PCR will result in over 200 ng amplicon which after clean-up will be at concentration of 4ng/ul. This amount will be equivalent to 12 nM (if you are using Illumina validate region primers). If you are not obtaining these figures, the problem might be in input DNA. Increasing cycle number may cause biases in amplification of various species DNA present in sample. In any case the metagenomics profile is always one view of community and there is no common agreement on it. If increasing amplicon concentration is the aim, it can be achieved easily by decreasing elution volume from beads.

            I do not know the likelihood of cross contamination in SpeedyVac (I think pretty low), but it should not matter because the amplicons already have been indexed.

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by nucacidhunter View Post
              Illumina’s guide for clustering libraries on MiSeq can be find by google searching “Preparing Libraries for Sequencing on the MiSeq”. 4 pM is the concentration for denaturation step.
              You mean 4nM.

              --
              Phillip

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                Yes, 4 nM. Thanks for correction.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by nucacidhunter View Post
                  We saw inconsistent cluster density. Illumina recommends concentration by SpeedyVac. Other options would be concentrating them using column or AMPure beads by reducing elution volume.
                  Yes.
                  The real issue here is probably the concentration of NaOH in the sample loaded into the cassette. If it is too high, then it would interfere with binding of the amplicons to the flowcell.

                  I wish Illumina would bear down and create a denaturation procedure that doesn't require a subsequent 50X dilution of the sample with buffer prior to loading.

                  --
                  Phillip

                  Comment

                  • nucacidhunter
                    Jafar Jabbari
                    • Jan 2013
                    • 1250

                    #10
                    I wish Illumina would bear down and create a denaturation procedure that doesn't require a subsequent 50X dilution of the sample with buffer prior to loading.
                    It seems that Illumina looking into different methods for denaturing libraries for clustering. For HiSeq X system, library is mixed with three solutions (PCX1-3) and RSB (without an initial denaturation) and loaded into cBot. Libraries for this system must be denatured in cBot perhaps by heating (if someone have the related cBot recipe can check this). Also, llumina’s current protocol for clustering and sequencing on NextSeq 500 includes a naturalisation with Tris-Cl pH7.0 after NaOH denaturation step.

                    Comment

                    • BioGenomics
                      Member
                      • Apr 2009
                      • 25

                      #11
                      Hi nucacidhunter et al, anybody has experience with using a 0.5 nM 16s library concentration (instead of 2nM) on the MiSeq ? The NextSeq500 has this protocol (updated October 2014, so more recent) and we thought to add the 200mM TRis buffer Ph 7.0 too to hydrolyse the higher NaOH concentration needed initially. Thanks

                      Comment

                      • acockburn
                        Member
                        • Aug 2013
                        • 10

                        #12
                        For subnanomolar libraries we use the protocol found in the supplemental section (Protocol 12) in the paper below. Its a fantastic protocol and I have used it for libraries with 100 pM (diluted to 4 pM for the final load) and had successful runs.



                        Nature Methods 5, 1005 - 1010 (2008)
                        Published online: 25 November 2008 | doi:10.1038/nmeth.1270

                        A large genome center's improvements to the Illumina sequencing system



                        Supplementary info;
                        Supplementary Protocol 12: Modified hybridisation buffers

                        Comment

                        • RCJK
                          Senior Member
                          • May 2009
                          • 156

                          #13
                          We use the NextSeq protocol with the 200mM Tris pH 7 all of the time for our MiSeq runs.

                          Comment

                          • rpm27
                            Junior Member
                            • Sep 2014
                            • 3

                            #14
                            Hi,
                            I am wondering for subnanomolar concentration library, what was the final volume you (acockburn) loaded on the sequencing machine? The miseq protocol says 600ul, but curious what others have tried loading. Thank you.

                            Comment

                            • acockburn
                              Member
                              • Aug 2013
                              • 10

                              #15
                              I usually make 700 uL and load 600 uL. I have never tried less.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              38 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              23 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              73 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...