Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • sdavis
    Member
    • Jan 2010
    • 14

    #1

    Insert size with Agilent SureSelect

    We have noticed that our Agilent SureSelect libraries have a very small insert size after hybrid selection. Prior to selection, we see BioAnalyzer modes of 200-220 bp (with adapter length subtracted, so this is insert size). After hybridization and selection, we see 130-140 bp. On the sequencer, we are seeing similar very low numbers. Have other people seen this effect? Is there a way to work around it? We are planning to increase the input size to the hybridization, obviously, but we have not heard much about this issue and so are wondering if this is something specific to us or if this is a more general observation. The short insert size has some considerable implications for data quality and subsequent variant calling, so we would certainly like to get an insert size that is more reasonable.

    Thanks,
    Sean
  • GW_OK
    Senior Member
    • Sep 2009
    • 411

    #2
    I believe the sureselect baits are ~120bp in length, so they're probably preferentially binding the smaller fragments of your DNA. I know their protocol calls for shearing to ~150bp with an expected shift to 200-250 after ligation and amplification (prior to hybridization). Have you considered the possibility of concatamerization of your DNA if you shear to larger lengths?

    Comment

    • sdavis
      Member
      • Jan 2010
      • 14

      #3
      Insert size with Agilent SureSelect

      Thanks for the thoughts. We are using covaris to do the fragmentation and end up with 200-220 BP (without adapters) insert size prior to hybridization and end up loosing 60-80 bp for the mode by going through the hybridization process. If we end up running 2x80 or 2x100 bp runs, then we end up with many of the paired reads actually overlapping one another. While not a disaster, it isn't cost effective and makes later variant calling more difficult (because many of the bases are not independently read, being present in both the first and second reads). I'm not sure that concatemorizing would be our first choice, as it would introduce chimeras not present in the sample, if I understand you correctly.

      Sean

      Comment

      • GW_OK
        Senior Member
        • Sep 2009
        • 411

        #4
        Concatamerizing is no one's first choice, lol. I'm saying long fragments of DNA not covered by either baits or the adapter blocking oligos might hybridize, leading to off target sequences. While I can't say it would happen for sure, I'm not sure if I want to find out. Anyway, are you stuck with doing 2x80 or 2x100? I'm doing 2x75 on 150 bp fragments and it seems to be working out ok. Also, are you using the sureselect settings for the covaris? I've always had consistent peaks around 150 using those.

        Comment

        • sdavis
          Member
          • Jan 2010
          • 14

          #5
          I looked at the agilent kit a bit more. The size that the protocol targets is about 150bp, which we are a little under. However, even at 150bp for an average, there is a very large percentage of pairs that are going to overlap each other for a 2x75bp run. For the same reason that folks recommend removing duplicates, the portion of overlapping sequence should probably be trimmed so that a read is not "double-counted". So, in essence, this ends up reducing the effective read length and the extent of this reduction depends on the actual read length and how that relates to the insert size distribution. In practice, I think we are going to go ahead with a much larger insert size test (~300bp) and see what that gives us.

          Thanks for the help on this, GW_OK.

          Sean

          Comment

          • GW_OK
            Senior Member
            • Sep 2009
            • 411

            #6
            I'd be very interested to know how your longer insert sizes work once you've tried them out.

            Comment

            • krobison
              Senior Member
              • Nov 2007
              • 734

              #7
              Note: I've posted a question around the issue of how variant callers handle overlapping paired end reads to the bioinformatics forum. It's a good question & I wanted to make sure the right audience sees it.

              Comment

              • scotthappe
                Junior Member
                • May 2010
                • 1

                #8
                New Agilent All-Exon protocol

                To address this issue, we (Agilent) have released a new protocol for All-Exon capture that shifts the insert size, using a combination of Herculase II high-fidelity polymerase and SPRI-bead purification throughout. It may be found at:

                http://www.chem.agilent.com/en-US/Se...spx?whid=60197

                G3362-90001
                SureSelect Target Enrichment System for Illumina Paired-End Sequencing Library - Human All Exon and Human All Exon Plus Protocol
                v.2.0.1

                When following this protocol, the library size after capture (including adapters) is 325-350 bp (= insert size of ~230 bp).

                Best regards,
                Scott

                Comment

                • GW_OK
                  Senior Member
                  • Sep 2009
                  • 411

                  #9
                  Hmmm. Is this compatible with the exome kits I've already got in my freezer (I'm assuming yes)? Also, the new protocol linked says after shearing you should peak at 190, not 230.

                  Edit: I suppose it technically doesn't say base size at all after shearing, but the included BioA trace peak is at 190.

                  Edit Edit: It does say 150 to 200 at the ed of the shearing step but it doesn't say at the size verification step.

                  Edit^3: You didn't change the covaris settings at all (?!)

                  Edit^4: Agilent tech support says the Covaris settings listed in table 8 of the new protocol do not reflect the new insert size changes and they will fix it.
                  Last edited by GW_OK; 05-10-2010, 10:51 AM.

                  Comment

                  • GW_OK
                    Senior Member
                    • Sep 2009
                    • 411

                    #10
                    OK so Owen Hardy from Sureselect Support says they're not really changing Covaris settings and are merely widening the acceptable range of fragments from a peak at 150 to a peak anywhere between 150-200.

                    This has gotten me thinking, though. When my lab upgrades to the HiSeq this insert size issues is going to become very important as the HiSeq default run length is 100bp paired end or 200bp single end (you can do shorter but that seems to be a bit of a hassle, what with removing reagents and such). Should I switch to single end reads for my exome captures then?

                    Comment

                    • upenn_ngs
                      Member
                      • Sep 2009
                      • 70

                      #11
                      After capture the size-range put on the sequencer might be 150-500bp. There will be overlap on a small portion paired reads, but the increased throughput will compensate.

                      Comment

                      • GW_OK
                        Senior Member
                        • Sep 2009
                        • 411

                        #12
                        True, but if you did long single end reads there wouldn't be any overlap at all. And if the majority of your reads are ~150-200bp (which they ought to be, I think) you can read the whole thing in one go without having to do paired ends.

                        Why do paired end reads at all (for exome at least) if you can sequence the whole (or quite a bit of the) captured piece of DNA? You don't really need the assembly benefits of paired ends, especially since you're assembling to reference. And it's much cheaper in reagents...

                        Comment

                        • upenn_ngs
                          Member
                          • Sep 2009
                          • 70

                          #13
                          For us, it looks like 5 gigabases is the minimum sequence to get >80% exome coverage at 20x. Right now 2x100bp reads are the quickest route. As the read length and cluster density increase I can see where single-end reads might become advantageous.

                          Comment

                          Latest Articles

                          Collapse

                          • SEQadmin2
                            16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                            by SEQadmin2




                            NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                            10-30-2026, 06:46 AM
                          • SEQadmin2
                            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                            by SEQadmin2



                            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                            ...
                            10-14-2026, 09:55 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by SEQadmin2, 10-28-2026, 09:57 AM
                          0 responses
                          40 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 10-28-2026, 09:51 AM
                          0 responses
                          24 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 10-22-2026, 11:13 AM
                          0 responses
                          73 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 10-22-2026, 10:48 AM
                          0 responses
                          48 views
                          0 reactions
                          Last Post SEQadmin2  
                          Working...