  • docphil
    Junior Member
    • Aug 2014
    • 8

    #1

    Pippin Elute directly into MiSeq Library

    Hello, I recently got a PippinHT loaned to me for a demo, and I am using it for final library pool size selection before loading my library onto a MiSEQ. I am wondering if I can use the Elute directly from the Pippin and dilute/denature for MiSEQ prep. I am concerned that the TE buffer in the Pippin will interfere with sequencing. I have contemplated using a Qiagen Clean up spin column after elution from the pippin, but I am concerned about yield loss. Let me know if anyone has tried going directly from pippin elute to Miseq. Thanks!
  • Carcharodon
    Member
    • Jul 2015
    • 40

    #2
    If I may piggy-back on your thread: I'm also wondering whether it's necessary to perform a clean-up step post-Pippin.

    I've performed standard gel-extractions before and sent the product straight to sequencing (Sanger, extraction from 0.15% agarose gel) and it obviated the need for post-PCR ExoSAP clean-up.

    Does the same logic hold here? ^

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      Originally posted by Carcharodon View Post
      If I may piggy-back on your thread: I'm also wondering whether it's necessary to perform a clean-up step post-Pippin.
      Pippin elutate contains 50 mM Tris, 30 mM TAPS, 1-2mM EDTA with pH>8. In my experience some polymerases were able to amplify straight and some failed requiring a buffer change. Clustering may be affected if library concentration is low and it does not get diluted before denaturation step.

      Comment

      • kerplunk412
        Senior Member
        • Jun 2012
        • 119

        #4
        What volume is your Pippin eluate in? You shouldn't lose too much with a column-based purification, and if you can reduce your volume from say 40 to 12 microliters the increase in concentration should more than make up for any loss during cleanup. And considering the time and expense of sequencing versus performing another cleanup I think it makes sense to clean the product before sequencing.
        Last edited by kerplunk412; 08-13-2015, 02:54 PM.

        Comment

        • Carcharodon
          Member
          • Jul 2015
          • 40

          #5
          This all makes sense. Thanks everyone!

          Comment

          • Steven Abbott
            Junior Member
            • Apr 2013
            • 3

            #6
            Pippin to Sequencer

            We have been using Pippin Prep to size select our Small RNA libraries for about 2 years now. We do not perform any additional clean up following elution and have comparable clustering at sequencing. We were concerned initially as well, but testing showed very little variation. Another user said to reduce the elution volume, but that will decrease your efficiency from our experience. We commonly elute small amounts of material, and found that if we reduced the volume, we would lose even more of the sample.

            Comment

            • nucacidhunter
              Jafar Jabbari
              • Jan 2013
              • 1250

              #7
              Hi Steven
              I wonder in your experience how Pippin selected small RNA library compare to gel cut regarding recovery of off target size fragments considering that small RNA fragments cover a narrow range of 10 base around 150 bp.

              Comment

              • Steven Abbott
                Junior Member
                • Apr 2013
                • 3

                #8
                Currently we run a size range of 125 - 160 bp and are getting far better identification of small RNAs than when we were using standard gel isolation. I attribute this to getting less large fragmented RNAs taking up sequencing clusters, leaving mostly our target to take up the reads. It also helped us narrow in our input concentration for sequencing since the library size range is so consistent.

                Comment

                • docphil
                  Junior Member
                  • Aug 2014
                  • 8

                  #9
                  Sorry about the late reply...I have been using a concentrator column on my pippin elute to do a buffer change into EB. I found that using a qiagen spin column resulted in a significant yield loss probably due to pH of the pippin elute. I use an Amicon Ultra 30K 0.5ml column and bring up the pippin elute to 500ul in EB buffer,after the spin I collect the ~30ul of elute. I then quantify, dilute to 2nm, denature and load on the MiSEQ at 5pm. I am going to attempt to load pippin elute directly onto the MiSEQ this friday after reading some of your responses. I will let you all know what happens!

                  **Update**
                  After running my library through the BluePippin I quantified the elute and needed to dilute it about 15X in EB to reach 2nm for MiSEQ denature. I followed the general MiSEQ loading protocol and loaded at 5pm. My cluster density was about 850 and quality was comparable to libraries I did a buffer switch on after elution from the pippin. I would say it is safe to load pippin elute onto the MiSEQ as long as you dilute your library. hope this helps someone. Cheers!
                  Last edited by docphil; 11-11-2015, 01:15 PM. Reason: Update

                  Comment

                  • anna_m
                    Member
                    • Jun 2013
                    • 17

                    #10
                    I agree it is safe to run as it is. We carry out size selection using the blue pippin instrument on our TruSeq nano libraries after amplification and the final library is the eluted sample taken from the elution well of the cartridge. The libraries are quantified as normal (fluorometry, trace, qPCR) and then put forward for sequencing where they are denatured and diluted as normal. Glad it worked out for you.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                      by SEQadmin2




                      NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                      10-30-2026, 06:46 AM
                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-28-2026, 09:57 AM
                    0 responses
                    38 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-28-2026, 09:51 AM
                    0 responses
                    23 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    73 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    48 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...